I don't understand the addressing on the SC-2345 Carrier

0 views
Skip to first unread message

Fat Man

unread,
Mar 9, 2006, 1:41:39 PM3/9/06
to
I am using an SCC-TC02 thermocouple module with an SC-2345 signal carrier.&nbsp; The SCC-TC02 is plugged into Analog input 2/10 on the carrier.&nbsp; I wish to monitor the ampflified temperature signal coming from this thermocouple module with an oscope.&nbsp; I should be able to do this at the terminal block on the carrier.&nbsp; I am looking at the pinout on this terminal block (located in <a href="http://www.ni.com/pdf/manuals/371064d.pdf" target="_blank">http://www.ni.com/pdf/manuals/371064d.pdf</a>&nbsp; page 2-2) I see nothing that coincides with analog input 2 or 10.&nbsp; I do not undestand how this addressing works.&nbsp; Which screws on the terminal do&nbsp;I need to connect with?&nbsp;
&nbsp;
Thanks,
&nbsp;
&nbsp;
&nbsp;
&nbsp;

Laura F.

unread,
Mar 13, 2006, 2:44:42 PM3/13/06
to
Hello,
The terminals on the SC-2345 carrier are used to access digital and PFI lines that can be used for digital I/O and triggering.&nbsp; Unfortunately, you cannot access the analog input lines with those terminals.&nbsp;
I do not know of a way to access the voltage amplified by the SCC-TC02.&nbsp; What is the reason that you want to do this?&nbsp; If you want to read the voltage in software, you can do that from this module, instead of reading the temperature value.&nbsp; Would this help?&nbsp; If you want to try this, you can use Measurement and Automation Explorer (MAX) to create a voltage task or you can run a shipping example that reads voltages.&nbsp; In either method, you would specify the module as your physical channel, for example SCC1Mod1/ai0.
Hope this helps,
Laura