Jim1984
unread,Jul 17, 2008, 4:10:10 PM7/17/08You do not have permission to delete messages in this group
Either email addresses are anonymous for this group or you need the view member email addresses permission to view the original message
to
Sappster,
Thanks for the reply. I would say that the voltage range I will probably see lies between +/- 2 mV, so a somewhat small resolution (~ 1 uV) would be ideal. These thin-foil type gages rely on a material with known thermal conductivity that separates the two copper leads. The voltage/temperature gradient comes from this separation and not from the junction of two dissimilar metals like thermocouples. One lead is simply positioned at a "hotter" location relative to the other lead. I do not expect the temperature of the leads to exceed 70 °C at the module inputs, so the ambient temperature range is not of a huge concern. The reasons I am looking into replacing some of the TC-120's are because 1) they have only 8 channels whereas the AI-112's have 16 channels and these would allow me to measure every gage I have, 2) I have been having problems with getting unexpected negative readings from the gages, which may be some physical phenomenon or something with the TC-120 modules (I have the revision B modules), and 3) I am not certain if the cold-junction compensation on the TC-120 modules might be interfering with my mV readings. Thanks again for the reply.