9234 not seen by M&A explorer.

0 views
Skip to first unread message

PatMcHargue

unread,
Apr 9, 2008, 4:40:11 PM4/9/08
to
Folks;
 
I have an NI 9233 that I have connected to my laptop with a USB-9162 carrier.  M&A explorer sees it just fine.  The NiDAQmx version is: 8.5.0f5
 
When I swap the NI 9233 for my new NI 9234...  nothing.  M&A shows that the 9233 is no longer connected, but it does not show the 9234 as being connected.  My laptop doesn't ask to install any drivers either.  It's just isn't there.
 
Now I looked online to see what version of MiDAQmx I need to use with the 9234, and I seem to have the right version.
 
Any ideas about what I can/should do to get this to work?  Is the best solution just to chuck it back in th box, and get an RMA from NI?
 
TIA;
 
Patrick
 

PatMcHargue

unread,
Apr 10, 2008, 12:40:12 PM4/10/08
to
Mr. Johnson:
 
>
>National Instruments has not completed all the testing for the USB-9234, and thus you do not find it as an orderable
>part number on our website.  However, support for the 9234 in the USB sleve (9152) was added in 8.6 of DAQmx. 
>Our current shipping version of DAQmx is 8.71 which also contains support for USB-9234. 
>
 
Thank you.  I hope that you're referring to eh USB-9162 carrier. 
 
I'll dig up the NiDAQmx >= 8.6, and see what happens
 
Pat
 

preston johnson

unread,
Apr 11, 2008, 4:10:12 PM4/11/08
to
Sorry for the typo, yes I meant the 9162 carrier. 
Preston

PatMcHargue

unread,
Apr 14, 2008, 12:40:07 PM4/14/08
to
Hot dog!  NiDAQmx v8.7.1 did the trick, and the 9234 is now visible in M&A explorer.  Thanks for the help.
 
Pat
 