Hey All,
I've been looking at the longslit reduction routines and there seems to
be an error with the way it is calculating the variance. I looked into
it since the longslit data i was reducing didn't look to have the
appropriate signal to noise.
The regular image data is fine but the variance is off. Instead of
computing
[ (A+B)*Gain + RN**2 ] / itime**2 ,
as the variance (as is done in the regular multi-slit mask reductions)
it computes
[ (A+B) + (RN/Gain)**2] * (Gain / itime) ,
where RN takes into account the MCDS mode for the number of reads.
On another note, as set up the longslit reductions produces a image of
the variance while the multi-slit mask reductions give images the
standard deviation instead. Should the longslit routines be modified to
give the same kind of outputs as the regular mask reductions?