For GR&R the rough rule is 1-10% great, 11-20 okay, 21-30 okay with
approval, etc.
We can calculate the above values, but do not know what the results mean.
Any books available that discuss this?
Thanks,
George
If the percentages are calculated as a percent of Tolerance then they mean that
the amount of variation caused by the act of measuring with that gage is the
calculated percent o the tolerance. So if you have a Tolerance of +/- .005"
with a 20 % gage R&R
that means that 20% of your tolerance is already used up by measurement induced
variation leaving only 80% of the tolerance available for all other sources of
variation including parts. In this case 80% of +/- .005 is +/- .004 !
If the percentages are calculated as a percent of total variation, rather than
tolerance, they are the percent of all variation of your process which is due
to measuring. You would need to know that if you want to reduce the process
variation.
regards,
Dirk Jansen
George Kasdorf <kas...@compuserve.com> schreef in artikel
<655iui$5...@examiner.concentric.net>...
D .Jansen wrote in message <01bcfb7d$4e0a1a60$9b1b56c3@zwc00241>...
>Part of the QS-9000 is a manual called Measurement System Analysis.
>In this book you will find a lot of usefull information regarding this
>subject
>
>regards,
Yes it certainly does have a lot of useful information. BUT is does not list
what "acceptable" levels are, or if it does I haven't found the correct
page.
So will continue to look for a reference work on this.
Rejb15kc wrote in message <19971124013...@ladder02.news.aol.com>...
>>Can anyone help out with what "acceptance" levels are for gage Stability,
>Linearity and Bias studies?
>>
>For Bias we use 1/4 of the smallest increment the gage can measure
>For Linearity, we require the slope be less than .05
>For Stability, the Automotive industry standard is that the stability, when
>plotted on an X-bar & R chart, mus never go out of control, If it does then
the
>gage is not stable.
>
We compared these to some of our results and everything looked good. So feel
this will work... but I still think I need to have some evidence of where we
derived these values from, so will continue to hunt down a book on the
subject.
Later,
George
The Stability requirement I quote is directly from AIAG book"Measurement system
Analysis". The Bias and Linearity requirements are agreements between us and
the Auto manufacturers. AS for Gage R&R the AIAG book specifically says 10% is
acceptable, up to 30% is acceptable under certain conditions.