Just an update on the above issue:
In applying the methods described in the Tank Lab paper (pages 48-50), we were surprised to see that sometimes a timestamp was advanced in time relative to the corresponding TTL. Upon further investigation, we found out that TTL pulses were also occasionally dropped (29 out of a total of 276080 TTLs in our ~2.5 hr test recording at 30 Hz). It's not a lot of dropped TTLs, but it could throw off time syncing if not accounted for. One possible reason is that the Neuralynx DAQ we have (Digital Lynx SX) is made to receive TTLs that are 5V, and the Miniscope DAQ outputs 3.3V pulses. On the other hand, the vast majority of the Miniscope DAQ’s 3.3V TTLs are recognized by the Neuralynx DAQ, so who knows. Since the timing of them are normally so regular, it was easy to identify where they were missing and insert guessed timestamps, but I thought it would be good to report this possible issue.
As a separate question, can the Miniscope DAQ hardware safely receive 5V TTLs to trigger the start of miniscope recording?
Thanks,
Eric
--