GE Dash 3000/4000/5000 monitors issues

20 views
Skip to first unread message

William Skaff

unread,
Oct 4, 2022, 5:49:12 AM10/4/22
to MedsVise Forum
We have in our workshop many GE Dash 3000/4000/5000 monitors that reboots in the middle of monitoring. They all have the same problem. Monitor displays for about 2-3 min. and reboot, or the monitor reboots almost right after it finishes starting up. It keeps doing it. All Dash’s that failed, all did similar.

Common error messages: 
  • Error message "SERVICE MONITOR, Possible Data Corruption"
  • Power on self-test fails every time at the "RAM test FAILED at ADR 30xxxxxx"
  • Missing ECG waveform.
  • Error message "Not for patient use"
  • SPO2 intermittently picking up the patients signal. This is an issue with the Dash 3000/4000/5000
  • Error message "NIBP Hardware Malfunction"
These errors occurs normally when there is a problem in the DAS (Data Acquisition Assembly). You will need to replace it. (GE approx. value: $4,000.00). 
 
We've replaced a lot of DAS Modules for half GE prices. These Dash's have been also upgraded to v6.7 with Oximax or Masimo SpO2 technology.
Click  here to buy this item.