We have in our workshop many GE Dash 3000/4000/5000 monitors that reboots in the middle of monitoring. They all have the same problem. Monitor displays for about 2-3 min. and reboot, or the monitor reboots almost right after it finishes starting up. It keeps doing it. All Dash’s that failed, all did similar.
Common error messages: - Error message "SERVICE MONITOR, Possible Data Corruption"
- Power on self-test fails every time at the "RAM test FAILED at ADR 30xxxxxx"
- Missing ECG waveform.
- Error message "Not for patient use"
- SPO2 intermittently picking up the patients signal. This is an issue with the Dash 3000/4000/5000
- Error message "NIBP Hardware Malfunction"
These errors occurs normally when there is a problem in the DAS (Data Acquisition Assembly). You will need to replace it. (GE approx. value: $4,000.00).
We've replaced a lot of DAS Modules for half GE prices. These Dash's have been also upgraded to v6.7 with Oximax or Masimo SpO2 technology.