We are Client Server 5.67 pp21. We have this instrument in Bacti since 2014.
We are having some technical difficulties getting it connected to Meditech.
Does anyone have this instrument interfaced who could tell us if the machine is acting as the host or if
Meditech is acting as the host ( vs Client) ?
Our network connects to a computer, and from the computer there is a cable that goes to the instrument.
Any help anyone could provide would be much appreciated.
Gale Y. Freeman, MIS Systems Analyst
Samaritan Medical Center MIS dept,
830 Washington St, Watertown, NY 13601
315-785-5751, gfre...@shsny.com
Please consider the environment before printing this email. This message and any attachments may contain information that is protected by law as privileged and confidential, and is transmitted for the sole use of the intented recipient. If you are not the intended recipient, you are hereby notified that any use, dissemination, copying or retention of the e-mail or the information contained herein is strictly prohibited. If you received this e-mail in error, please immeditately notify the sender by e-mial and permantently delete this e-mail.
How am I doing? Fee free to email my manager, Joan Varsics, jvar...@shsny.com , for any suggestions or comments.

We are 5.66 PP8 and have this instrument. It has been awhile since I set it up but I would be glad to dig thru documentation and look thru my dictionaries if you want to give me a call. Tuesdays and Wednesdays are best for me. As far as the next two days, mornings are better.
Michelle Windle
LIS/Clinical Coordinator
Saline Memorial Hospital
(501) 776-6157 phone
(501) 776-6158 fax
Disclaimer
The information transmitted via this e-mail is intended only for the person or entity to which it is addressed and may contain confidential and/or proprietary information. Any use, review, retransmission, dissemination or other use of, or pursuing of any action in reliance upon this information by persons or entities other than the intended recipient is strictly prohibited. If you are the recipient of this e-mail transmission in error, please reply to the sender and delete the material from any computer. Thank you.
We are Meditech 6.08
Our GX Cepheid is used for MRSA,CTNG, etc. We set this up under the LAB instruments NOT in Microbiology.
We have this interfaced since 2014, no issues so far.
Gene X acts as the server.
I have this in my Notes (cause I get confused when is meditech considered a server or host):
You can tell by checking the device dictionary. If there is no network name/ip address then Meditech is the server. If there is a network name/ip address, then it could be the analyzer as the server or you are connecting through a termserver.
Thank you,
Leah Dungo,CLS,ASCP,AMT
INFORMATION SYSTEM SPECIALIST-LAB
P:(830)278-6251 ext. 1143
Uvalde Memorial Hospital ·
1025 Garner Field Road
· Uvalde, TX 78801
·
www.umhtx.org
Promoting the healing of those we serve by providing compassionate, high-quality healthcare.
This message, together with any attachments, is intended only for the use of the individual or entity to which it is addressed and may contain information that is confidential and prohibited from disclosure. If you are not the intended recipient, you are hereby notified that any dissemination, or copying of this message, or any attachment, is strictly prohibited. Please notify the original sender immediately by telephone or by return e-mail and delete this message along with any attachments, from your computer. Thank you. This e-mail is Property of UMH.
Thanks
What do you have for these settings on the instrument…
1. Check Enable Host Communication box
2. Change Host ID to read MEDITECH ?
3. Protocol : - is it ASTM or HL7
4. Run Host as is it Server or Client ?
From: Leah Dungo [mailto:l.d...@umhtx.org]
Sent: Monday, May 08, 2017 1:12 PM
To: Windle, Michelle; Freeman, Gale; meditech-l
Answers below…
Thank you,
Leah Dungo,CLS,ASCP,AMT
INFORMATION SYSTEM SPECIALIST-LAB
P:(830)278-6251 ext. 1143
Uvalde Memorial Hospital ·
1025 Garner Field Road
· Uvalde, TX 78801
·
www.umhtx.org
Promoting the healing of those we serve by providing compassionate, high-quality healthcare.
From: Freeman, Gale [mailto:GFRE...@shsny.com]
Sent: Tuesday, May 09, 2017 8:08 AM
To: Leah Dungo; Windle, Michelle; meditech-l
Subject: RE: Bacteriology instrument GenExpert Cepheid
What do you have for these settings on the instrument…
1. Check Enable Host Communication box [Leah Dungo] yes
2. Change Host ID to read MEDITECH ?[Leah Dungo] yes
3. Protocol : - is it ASTM or HL7[Leah Dungo] ASTM
4. Run Host as is it Server or Client ?[Leah Dungo] CLIENT
My answers are the same as Leah’s.
Michelle Windle
x6157
Thank you everyone who tried to help us ! We did get it working in test. …
Ours is Client server and we are looking at Meditech as the host.
Device dictionary is setup is
Type is SOCKET
Usage is LAB ANALYZER
Network name is blank
Port number we used is 56565
Listening time out is 60
Then on the instrument side…
Instrument side…
Enable host communication
HOST ID = MEDITECH
Protocol = ASTM
Run Host as = Server
Server IP Address is our test MEDITECH background server that runs lab jobs.
For live it will be our live lab background server.
Port # is 56565
We did not use any special cabling.
I hope this helps someone going through same issues…
Gale