Both the RelativeXrayExposure and the new ExposureIndex tags do NOT
provide
information relevant to the radiation exposure of the patient and
associated organ dose.
They are used to monitoring image quality. Low values imply that the
image will
have high noise. High values imply low noise and suggest that a lower
technique could be used.
We post acceptable range for our technologists. If a radiograph is
below the bottom limit,
they are asked to repeat it. If it is above, they are told to
definitely not repeat it, but
to monitor subsequent studies to see if their techniques should be
systematically reduced.
We are just now setting up to extract these from DICOM tags and manage
them
within a dB as a few other sites have done. The legacy values are
still useful if
reports are compared against the high/low limits for the manufacturer.
Moreover, the legacy units can also be converted to standardized units
since
the manufacturers have been open about the relations between their
indices and the
air kerma in the calibration beam.
Entrance exposure (ESAK) estimation still needs kVp, ma, sec (ma-S)
information
that is generally only available from integrated system like a DR
room.
Even with the technique, a good estimate still needs output
calibration data.
> >> The Standard definition of **RelativeXRayExposure states it is the