IMAGE UID in TID 1004 & 10003

37 views
Skip to first unread message

yaron

unread,
Apr 4, 2012, 3:49:50 AM4/4/12
to medical-imaging-radiation-dose-informatics
Hi All.
During the preparation to the bern connectathon I am validating my REM
SR with DoseUtility.
I got 2 errors that both does not exists when validating with DVTK &
dcmtk:
Error: 1.7.9.4: /CONTAINER (113701,DCM,"X-Ray Radiation Dose Report")/
CONTAINER (113702,DCM,"Accumulated X-Ray Dose Data")/CODE
(113876,DCM,"Device Role in Procedure")/UIDREF (121012,DCM,"Device
Observer UID"): Content Item not in template

Error: 1.8.9: /CONTAINER (113701,DCM,"X-Ray Radiation Dose Report")/
CONTAINER (113706,DCM,"Irradiation Event X-Ray Data")/IMAGE: Content
Item not in template

both items are mandatory items (the first for TID 1004 and the second
for TID 10003).
Is there a problem with the validation tool?


David Clunie

unread,
Apr 4, 2012, 7:59:31 AM4/4/12
to medical-imaging-radia...@googlegroups.com, yar...@cmt.co.il
Hi Yaron

Email me the file please, and I will take a look at it.

The problem with the IMAGE content item may be that you do not
have the require concept name for the IMAGE perhaps; it is
required to be (113795,DCM,�Acquired Image�), otherwise it
doesn't match the template.

The Device Observer UID error is probably spurious; this content
item was added in CP 1065 and I may have missed adding it to the
validator. I will check and let you know when I have a new version
of the validator if this is indeed the problem.

David