Hello Jijing,
1. In short - it does matter.
Vials' material also generates local magnetic field, however, it usually is disregarded under assumption that Gd-related field is much stronger. Also, as a rule these materials do not have signal at conventionally used echo times which further hinders mapping of their susceptibility.
Thus, to minimize role of the containing material one should use thin-walled vials made of low-susceptibility material (something like a falcon tube should work).
2. Yes, you're correct - values in QSM images are relative. Furthermore, such notions as "para-" and "diamagnetism" are also normally defined in literature relatively to water.
It is just a matter of setting a zero level - you still can convert to absolute susceptibilities by adding known shift (e.g., -9 ppm in case you measure susceptibilities relative to water.)
What was the issue with the phantom you made?
Regards,
--
Alexey V. Dimov