Is there any way to setup NUS for 2D experiments with software topspin 3.0?
3D NUS for HNCACB, HNCO, HNCA, 15N-separated NOESY and HNHA run very well
and the reconstructed data with mddnmr2.0 show ~2-time S/N improvement
compared with traditional data. But the 3D NUS 13C-separated NOESY is very
bad and the spectrum shows much smaller number of cross peaks compared
traditional NOESY. Did anyone run the 13C-separated NOESY experiment? How
are your data? Any suggestion?
Many thanks,
Youlin Xia
University of Minnesota
www.umn.edu/nmr
--
You received this message because you are subscribed to the Google Groups "mddnmr" group.
To post to this group, send email to mdd...@googlegroups.com rom your registered email address.
To unsubscribe from this group, send email to mddnmr-un...@googlegroups.com
For more options, visit this group at http://groups.google.com/group/mddnmr
I acquired a 3D 13C-separated NOESY (NOESYHSQCETGPSI3D) with traditional method. The spectrum looks good. The parameters are d1=1, ns=8, d8=0.2, TD=(70, 160, 2k).
Then I acquired three corresponding NUS spectra with sparse rate=30% & J=(0,0,0), and T2=(0.02, 0.02, 1.0), (0.05, 0.10, 1.0), or (0.02, 0.05, 1.0). I used mddnmr2.0 to process the data with parameters METHOD=’MDD’, SRSIZE=0.1, NCOMP=25 to 600, NITER=50 to 100, OVLP=3 MDD_NOISE=0.7, lambda=0.01, CT_SP=’nnn’, CEX=’nnn’.
To process other 3D data including 15N-separated NOESY, above mddnmr parameters seem ok. But I’ll try to reduce MDD_NOISE as Maxim suggested to test.
I acquired a 3D 13C-separated NOESY (NOESYHSQCETGPSI3D) with traditional method. The spectrum looks good. The parameters are d1=1, ns=8, d8=0.2, TD=(70, 160, 2k).
Then I acquired three corresponding NUS spectra with sparse rate=30% & J=(0,0,0), and T2=(0.02, 0.02, 1.0), (0.05, 0.10, 1.0), or (0.02, 0.05, 1.0). I used mddnmr2.0 to process the data with parameters METHOD=’MDD’, SRSIZE=0.1, NCOMP=25 to 600,
The number of iterations may be be too little and is good only for the first look at the spectrum. For the final calculations, try 200-500. It will lengthen the calculations, of cause.NITER=50 to 100,
Dear Dr. Orekhov,
It’s very nice to have the chance to discuss mddnmr with you. The enclosed is a table to compare S/N for traditional and NUS HNCACB and HNCA. I randomly chose three peaks. And I changed parameters NCOMP and NITER. It seems to me NCOMP=600 and NITER=100 are good, which gives me more than 2-time S/N improvement compared with traditional way. Increasing NCOMP can significantly decrease noise level. Please note that the factor 1.825 is from sqrt(15/4.5). This is because the traditional HNCACB took 15 hours and NUS HNCACB only took 4.5 hours, and the S/N of NUS spectrum should be multiplied by 1.825 if it took 15 hours instead of 4.5 hours.
What else parameters should I change? I may play with the parameter MDD_NOISE.