Hi,
You can calibrate your Digital thickness gauge for C.S. Vessel thickness gauging on Carbon steel material Step Wedge preferably having nearest thicknesses to your job thickness.
They are available in range of 1-10 mm (1 mm step), or 5 -25 ( 5mm step).
For Twin crystal probe do the calibration on two different thicknesses.
For cladding material if cladding material of S.S. on C.S. pipe with low thickness ( below 10 mm) then there will be not much difference in measured thickness if you calibrate on C.S. block.
But if your pipe thickness is more than 15mm then you have to make a block of C.S. to Clad material having thickness same as your pipe.
If you are having variable thicknesses then you can make this blocks in 5mm steps.
you could not calculate the thickness by formula for unknown metal thickness.
Regards,
Nilesh Pathare.
ASNT NDT LEVEL III