Hi,
I have been trying to use Lipidxplorer for the Precurson Ion and neutral loss scan data obtained from a Waters triple quadrupole instrument. I am new to Lipixplorer and have tried to read whatever is available online on Lipidxplorer. While importing data I get the following debugging message.

Importing Pos_184
Polarity or sample C:\Users\Prem Gupta\Documents\Lipidxplorer\Pos_184: 1
Importing Pos_184A
Polarity or sample C:\Users\Prem Gupta\Documents\Lipidxplorer\Pos_184A: 1
0.24 sec. for reading the spectra
0.00 sec. for calibrating the spectra
Aligning MS spectra linear
> Nb. of MS peaks (after alg.)..          0

Aligning MS/MS spectra linear

0.00 sec. for aligning the spectra

Save output to C:\Users\Prem Gupta\Documents\Lipidxplorer\Lipidxplorer.sc.
0.25 sec. for the whole import process




What could be the issue? Any help will be highly appreciated.
Thanks,
Prem Gupta.

Prem Gupta

unread,
Sep 8, 2017, 12:01:03 AM9/8/17
to LipidXplorer discussion group
Hi,
I also wanted to add that I am checking MS Only as my spectra is in the form of parent ion masses and intensities. Is it possible to know generic triple quadrupole settings to begin? Then I can tweak as per my requirement.
Thanks,
Prem.

Dr. Ronny Herzog

unread,
Sep 8, 2017, 4:02:49 AM9/8/17
to lipidxplorer-d...@googlegroups.com

Dear Prem Gupta,

It looks like your import settings don't fit. Did you checkout this paper? There are some sentences about settings for quadrupol instruments.

I attached the settings file which we used for this paper. It contains a setting for a low resolution instrument (HCD_7500_7500).

However, I do not think that the resolution setting is the problem, but the time range and/or mass range setting. Check if this is correct.

Cheers,
Ronny


On 08.09.2017 06:01, Prem Gupta wrote:
Hi,
I also wanted to add that I am checking MS Only as my spectra is in the form of parent ion masses and intensities. Is it possible to know generic triple quadrupole settings to begin? Then I can tweak as per my requirement.
Thanks,
Prem.

On Thursday, September 7, 2017 at 8:37:52 PM UTC-7, Prem Gupta wrote:
Hi,
I have been trying to use Lipidxplorer for the Precurson Ion and neutral loss scan data obtained from a Waters triple quadrupole instrument. I am new to Lipixplorer and have tried to read whatever is available online on Lipidxplorer. While importing data I get the following debugging message.

Importing Pos_184
Polarity or sample C:\Users\Prem Gupta\Documents\Lipidxplorer\Pos_184: 1
Importing Pos_184A
Polarity or sample C:\Users\Prem Gupta\Documents\Lipidxplorer\Pos_184A: 1
0.24 sec. for reading the spectra
0.00 sec. for calibrating the spectra
Aligning MS spectra linear
> Nb. of MS peaks (after alg.)..          0

Aligning MS/MS spectra linear

0.00 sec. for aligning the spectra

Save output to C:\Users\Prem Gupta\Documents\Lipidxplorer\Lipidxplorer.sc.
0.25 sec. for the whole import process




What could be the issue? Any help will be highly appreciated.
Thanks,
Prem Gupta.

--
You received this message because you are subscribed to the Google Groups "LipidXplorer discussion group" group.
To unsubscribe from this group and stop receiving emails from it, send an email to lipidxplorer-discuss...@googlegroups.com.
For more options, visit https://groups.google.com/d/optout.

--
Dr. Ronny Herzog
Software Development and IT-Infrastructure

Lipotype GmbH
Tatzberg 47
01307 Dresden
Germany

Tel.: +49 351 796 5343
Fax: +49 351 796 5349
Email: her...@lipotype.com
Web: www.lipotype.com

Follow us on LinkedIn and subscribe to our newsletter to receive the latest news about Lipotype and lipidomics.

VAT ID: DE282974828
County Court Dresden HRB 31124
Managing Director: Prof. Dr. Kai Simons

This message is intended for the addressee only as it contains private and confidential information. The contents are not to be disclosed to anyone other than the addressee. Unauthorised recipients are requested to comply with the above and to inform the sender immediately of any errors in transmission.

lpdxImportSettings_tutorial.ini

Prem Gupta

unread,
Sep 11, 2017, 9:53:59 AM9/11/17
to LipidXplorer discussion group
Thanks a lot Ronny for suggestions. I will look into these settings.
Prem.
To unsubscribe from this group and stop receiving emails from it, send an email to lipidxplorer-discussion-group+unsub...@googlegroups.com.

For more options, visit https://groups.google.com/d/optout.

Prem Gupta

unread,
Sep 11, 2017, 10:53:10 AM9/11/17
to LipidXplorer discussion group
Ronny,
Why do we use time setting if the .csv file does not contain any information about time? It contains just m/z and intensity. I am attaching two spectra. Is this a correct format? Tthis is a precursor ion scan m/z 184 for PC and LPC.
Thanks,
Prem Gupta.
Thanks,
Prem.

On Friday, September 8, 2017 at 1:02:49 AM UTC-7, Dr. Ronny Herzog wrote:
To unsubscribe from this group and stop receiving emails from it, send an email to lipidxplorer-discussion-group+unsub...@googlegroups.com.

For more options, visit https://groups.google.com/d/optout.
184.csv
184B.txt

Dr. Ronny Herzog

unread,
Sep 13, 2017, 5:24:36 AM9/13/17
to lipidxplorer-d...@googlegroups.com

Dear Prem,

LipidXplorer does not support numbers formatted as a combination of exponents. You have to make sure that e.g. 5.12E+03 is written 5120.

Cheers,
Ronny
To unsubscribe from this group and stop receiving emails from it, send an email to lipidxplorer-discuss...@googlegroups.com.

For more options, visit https://groups.google.com/d/optout.