Hi,

We are planning a pipeline for processing SMS DSI data for a multi-vendor TBI project starting from raw data from each platform. For the Philips we've decided to start with LIST/DATA rather than SIN/LAB/RAW since the steps between are better done using the Philips algorithms. 

Do you know of any effort to do this - no sense reinventing if it's already done.

Thanks,

Joe

Research Associate
Division of Magnetic Resonance Research
Department of Radiology and Radiological Science
Johns Hopkins University School of Medicine

Michael Hansen

unread,
Feb 26, 2016, 12:47:35 PM2/26/16
to jsgi...@gmail.com, ISMRMRD
I don't know of a LIST/DATA -> ISMRMRD converter. I have limited experience with that format, but I think it should be very easy to write the converter since the data is already sort of preprocessed a bit as you point out. If you write a converter, I am happy to help in terms of how to get started and some other general guidance. You can contact me directly if you would like to have a chat on the phone about it. I would also encourage you to share the converted if you make one. 

--
You received this message because you are subscribed to the Google Groups "ISMRMRD" group.
To unsubscribe from this group and stop receiving emails from it, send an email to ismrmrd+u...@googlegroups.com.
To post to this group, send email to ism...@googlegroups.com.
To view this discussion on the web visit https://groups.google.com/d/msgid/ismrmrd/54fed6ee-c5f1-4e58-b752-91e239e8565a%40googlegroups.com.
For more options, visit https://groups.google.com/d/optout.

Joseph S. Gillen

unread,
Mar 24, 2016, 12:21:41 PM3/24/16
to Michael Hansen, ISMRMRD
I've made progress on integrating code for LIST/DATA conversion into the existing philips_to_ismrmrd code used for SIN/LAB/RAW format. I have some questions:

Raw data in the LIST/DATA file contains multiple data "types":
STD - standard acq
REJ - rejected standard data - arrythmia rejection - doesn't seem to have a place. If it appears, I suppose it can be skipped and not copied into the h5 file but we’re not using any cardiac gating. 
PHX - phase correction - I am setting the ISMRMRD_ACQ_IS_PHASECORR_DATA flag. 
FRX - frequency spectrum correction - there does not appear to be an appropriate flag - what to use? New flag?
NOI - setting the ISMRMRD_ACQ_IS_NOISE_MEASUREMENT flag
NAV - setting the ISMRMRD_ACQ_IS_NAVIGATION_DATA flag
DNA - dynamic navigator - same as NAV??

In the SMS acquisitions I'm using for test, I see only NOI, FRX, PHX and STD.
NOI - one long acq for each coil
FRX - 2 acqs for each coil for each slice - one with sign +1, one sign -1
PHX - 2 acqs for each coil for each slice at first PE - one with sign +1, one sign -1
the FRX and PHX acqs are 1/2 size of STD acqs - appears they are not oversampled

Philips encodes ky and kz using signed integers with center k-space at zero and negative and positive values. The kspace_encoding_step_1 and _2 of the ISMRMRD_EncodingCounters struct are unsigned integers. So I am shifting the values up by the minimum.

It appears that we’ll really need multiple .xsl files to map dimensions like fMRI dynamics or DTI/DSI vector or ASL types to the repetitions dimension of the schema. Is this how it was envisioned to work?

Thanks,