Damage to X-Ray Tube of CT Scan , under EEI policy

748 views
Skip to first unread message

Vinod Soorma

unread,
Mar 6, 2017, 9:06:21 AM3/6/17
to insurance...@googlegroups.com
Dear All,

As per EEI policy, Special Exclusion to Section 1, The company is not liable for 

"loss of or damage to bulbs, valves, tubes... etc."

Kindly confirm, whether or not X-Ray Tube falls under the definition of 'valve'.

Thanks,

Vinod Soorma
 09810088013 Delhi

Piyush Siinghal

unread,
Mar 6, 2017, 11:52:35 PM3/6/17
to insurance_surveyors
X Ray tubes are special purpose valves but the policy defines the life of the X-ray tube. No standard deletion is done, rather depnding upon the exposures, the claim is paid on the remaining life.

--
You received this message because you are subscribed to the Google Groups "Insurance surveyors" group.
To unsubscribe from this group and stop receiving emails from it, send an email to insurance_surveyors+unsub...@googlegroups.com.
To post to this group, send email to insurance_surveyors@googlegroups.com.
Visit this group at https://groups.google.com/group/insurance_surveyors.
For more options, visit https://groups.google.com/d/optout.



--
B/Rgds

Piyush Siinghal
0-9811047785

VINOD SOORMA

unread,
Mar 7, 2017, 8:24:44 AM3/7/17
to insurance...@googlegroups.com
Thanks Dear Piyush

From: Piyush Siinghal
Sent: ‎07-‎03-‎2017 10:22 AM
To: insurance_surveyors
Subject: Re: {Surveyors:28940} Damage to X-Ray Tube of CT Scan , under EEIpolicy

To unsubscribe from this group and stop receiving emails from it, send an email to insurance_surve...@googlegroups.com.
To post to this group, send email to insurance...@googlegroups.com.

vinod saboo

unread,
Mar 7, 2017, 8:24:46 AM3/7/17
to Insurance surveyors
Dear All,
It is covered by endorsement & paid as part of limited life.Percentage to be paid is also mentioned.
Rgds.

Piyush Siinghal

unread,
Mar 7, 2017, 10:46:08 PM3/7/17
to insurance_surveyors, soorma...@gmail.com
Here is the circular which can be found at  https://iib.gov.in/IIB/tac/circulars/ecir2001.htm


Engg/Gen-16/2001/26                                                                13th June, 2001

 

                        Re: Electronic Equipment Insurance :

                              A) Excess for Personal Computers

                              B) Insurance of VSAT Eqpt.

                              C) Insurance of Medical Eqpt.-Additional “Endorsement

 

            Tariff Advisory Committee decided to introduce the following amendments  in “Electronic Equipment Insurance”Tariff..

 

A)            “Personal Computers”

            It has been decided to revise the excess for “Personal Computers” to 5% of claim amount                       subject to a minimum of Rs. 2500/-.

 

B)        Insurance of VSAT equipment

            It has been decided to introduce the following rates and terms for V-SAT Equipment

 

            Rate                                 -  Rs.1.00% for all parts of VSAT Eqpt.

 

            Excess for AOG claims  -  10% of claim amount subject to a minimum  of  Rs.10,000/-

            Excess for other claims   -  As per EEI tariff

 

C)        Medical Equipment using X-ray tubes

           As regards depreciation for  Medical  Equipment using X-ray tubes , it has been decided to adopt the following ‘Endorsement’s wordings’ for which are given in  the “annexure”

 

            i) Cover of x-ray valves & tubes

           ii) Special condition Concerning Computer Tecnographs

          iii) Warranty for Lightning and over-voltage Protection Divices

           iv) Warranty concerning Air-conditioning Plant.

 

          The decision is effective from 13th June, 2001.

 

Encl: “annexure”                                                                                        

 

           

 

            ANNEXURE  TO  ITEM - 8(c) of 192 & Cir. No. 26 of 13/06/2001

_______________________________________________________________

Cover of Valves and Tubes

 

It is agreed and understood that otherwise subject to the terms, exclusions, provisions and conditions contained in the Policy or endorsed thereon, this insurance shall be extended to include loss of or damage to valves and tubes. Indemnification shall be limited to the actual value of such items (cf 1 - 7) immediately prior to the occurrence of the loss or damage, including ordinary freight, erection costs and custom duties and dues, if any.

 

1.  Actual values of

 

1.1.  stationary anode X-ray tubes in single-tank setup and rotating anode X-ray tubes without exposure counters for diagnostic equipment

1.2.  surface and close-range radio-therapy X-ray tubes and valves

1.3.  video amplifier tubes

 

     Age

     (months)

Actual value in % of new replacement value

     Less than   18

100

     Less than   20

90

     Less than   23

80

     Less than   26

70

     Less than   30

60

     Less than   34

50

     Less than   40

40

     Less than   46

30

     Less than   52

20

     Less than   60

10

    More than   60

0

 

2.  Actual values of valves for diagnostic equipment

     Age

     (months)

Actual value in % of  new replacement value

     Less than   33

100

     Less than   36

90

     Less than   39

80

     Less than   42

70

     Less than   45

60

     Less than   48

50

     Less than   51

40

     Less than   54

30

     Less than   57

20

     Less than   60

10

    More than   60

0

3.  Actual value of rotating anode X-ray tubes with lead-sealed exposure counters         for diagnostic equipment

     Number of  exposures

Actual value in % of new replacement value

     Less than    10,000

100

     Less than    12,000

90

     Less than    14,000

80

     Less than    16,000

70

     Less than    19,000

60

     Less than    22,000

50

     Less than    26,000

40

     Less than    30,000

30

     Less than    35,000

20

     Less than    40,000

10

    More than    40,000

0

 

4.  Actual values of deep therapy X-ray tubes and valves

    Period of operation (hours)OR Age (months)

(whichever results in the lower actual value)

   Actual value in % of      new replacement value

Period of operation (hours)

Age (months)

 

 

     Less than  400

     Less than  18

100

     Less than  500

     Less than  22

90

     Less than  600

     Less than  26

80

     Less than  700

     Less than  30

70

     Less than  800

     Less than  35

60

     Less than  900

     Less than  40

50

     Less than 1000

     Less than  45

40

     Less than 1100

     Less than  50

30

     Less than 1200

     Less than  55

20

     Less than 1300

     Less than  60

10

    More than 1300

    More than  60

0

                       

5.  Actual values of X-ray tubes and valves for material testing equipment

    Period of operation (hours) OR  Age(months)

   (whichever results in the lower actual value)

 

   Actual value in % of      new replacement value

    Period of operation or’

    (hours)

     Age

     (months)

 

     Less than  300

     Less than  6

100

     Less than  380

     Less than  8

90

     Less than  460

     Less than  10

80

     Less than  540

     Less than  12

70

     Less than  620

     Less than  14

60

     Less than  700

     Less than  16

50

     Less than  780

     Less than  18

40

     Less than  860

     Less than  20

30

    More than  860

    More than  20

20

 

 

 

6.  Actual values of picture and pick-up tubes for TV equipment

 

After 12 months’ use, the actual values of picture and pick-up tubes shall be reduced by 3% per month down to a minimum of 20% of the new replacement values.

 

7.  Actual values of other types of tubes and valves

 

For other types of tubes and valves the actual values on the date of an occurrence shall be determined on the basis of data furnished by the supplier.

 

________________________________________________________________

 

Special condition concerning Computer Tomographs

 

It is agreed and understood that otherwise subject to the terms, exclusions, provisions and conditions contained in the Policy or endorsed thereon, the Insurers shall not be liable for any demage consisting in the failure of individual construction elements or components, unless it can be proved that such damage has been caused by an external event acting on the system or by a fire generated within the system.

 

In contrast to the indemnity scales of Endorsement on “Cover for valves and tubes” incorporated in medical equipment, the following scales shall apply to the tubes indicated below built into computer tomographs:

 

1.  X-ray tubes

 

with high-voltage time meter (stationary-anode tubes):

(operating hours up to) 

with exposure counter (rotating-anode tubes):     (No. of exposures up to) 

           indemnity:                                                                    ( % )

400

10,000

100

440

11,000

90

480

12,000

80

520

13,000

70

600

15,000

60

720

18,000

50

840

21,000

40

960

24,000

30

1,080

27,000

20

1,200

30,000

10

 

 

 

2.  Tubes for voltage stabilization and regulation

 

  Period of use

  (months)

                      Indemnity                        %

36

100

39

90

41

80

44

70

47

60

49

50

52

40

55

30

57

20

60

10

 

______________________________________________________________________________

Warranty for Lightning and Overvoltage Protection Devices

 

It is agreed and understood that otherwise subject to the terms, exclusions, provisions and conditions contained in the Policy or endorsed thereon, the Insurers shall only indemnify the Insured in respect of loss of or damage to electronic equipment or data media or increased cost of working as a result of lightning or overvoltage if the electronic equipment is fitted with lightning and overvoltage protection devices and alarm system and these have been installed and maintained in accordance with the recommendations of the manufacturers of the electronic equipment and the lightning and overvoltage protection devices.

 

This means that the lightning and overvoltage protection devices and alarm system

 

--- are regularly serviced by qualified personnel of the manufacturer or supplier,

 

--- are kept under supervision by trained personnel,

 

--- are provided with an automatic switch-off device complying with the latest requirements for electronic equipment and the manufactur’s recommendations.


On Tue, Mar 7, 2017 at 12:16 PM, VINOD SOORMA <soorma...@gmail.com> wrote:
Thanks Dear Piyush

From: Piyush Siinghal
Sent: ‎07-‎03-‎2017 10:22 AM
To: insurance_surveyors
Subject: Re: {Surveyors:28940} Damage to X-Ray Tube of CT Scan , under EEIpolicy

X Ray tubes are special purpose valves but the policy defines the life of the X-ray tube. No standard deletion is done, rather depnding upon the exposures, the claim is paid on the remaining life.
On Mon, Mar 6, 2017 at 1:45 PM, Vinod Soorma <soorma...@gmail.com> wrote:
Dear All,

As per EEI policy, Special Exclusion to Section 1, The company is not liable for 

"loss of or damage to bulbs, valves, tubes... etc."

Kindly confirm, whether or not X-Ray Tube falls under the definition of 'valve'.

Thanks,

Vinod Soorma
 09810088013 Delhi

--
You received this message because you are subscribed to the Google Groups "Insurance surveyors" group.
To unsubscribe from this group and stop receiving emails from it, send an email to insurance_surveyors+unsubscribe...@googlegroups.com.

To post to this group, send email to insurance_surveyors@googlegroups.com.
Visit this group at https://groups.google.com/group/insurance_surveyors.
For more options, visit https://groups.google.com/d/optout.



--
B/Rgds

Piyush Siinghal
0-9811047785

--
You received this message because you are subscribed to the Google Groups "Insurance surveyors" group.
To unsubscribe from this group and stop receiving emails from it, send an email to insurance_surveyors+unsub...@googlegroups.com.
To post to this group, send email to insurance_surveyors@googlegroups.com.
Visit this group at https://groups.google.com/group/insurance_surveyors.
For more options, visit https://groups.google.com/d/optout.

--
You received this message because you are subscribed to the Google Groups "Insurance surveyors" group.
To unsubscribe from this group and stop receiving emails from it, send an email to insurance_surveyors+unsub...@googlegroups.com.
To post to this group, send email to insurance_surveyors@googlegroups.com.
Visit this group at https://groups.google.com/group/insurance_surveyors.
For more options, visit https://groups.google.com/d/optout.

SANJAY SURPURIYA

unread,
Mar 8, 2017, 1:04:49 AM3/8/17
to Insurance surveyors, insurance...@googlegroups.com, insurance-su...@googlegroups.com, insurance...@googlegroups.com, iiisla...@googlegroups.com, iiisl...@googlegroups.com
Dear All
            This  circular  is  almost  16 year old. At  that  time most  of  the  C.T. Scan machines were  single slice  machines. Due  to improvement  in technology now  new  machines  available  are up to 320 slices. 64  Slice  C.T. Scanners  are very common now. By  this  circular  the  value of  the  X ray tube after  40000 exposures  is  zero. If  we  consider  the average use  of  the  machine (  32 - 64 slice  machine) for  any  good  organisation  40000 exposures are  completed in 6 - 8 months  time. If  we  go by this circular  X ray tube  be  replaced  every  4-6 months. Is  it  the  case ??  I  have  seen a 64 slice C.T. Scanner  X Ray tube which has  completed 9 Lakh exposures  and  is  still   working. When this is pointed out   facts  are  side lined  by giving standard  reply-  always  there  are  exceptions. 

            Now  a  days C.T. Scanner ( 32 - 64 slice m/c)   X ray tube  working  for more  than 250000 exposures  is  very  common. Many  of  us  must  have  handled  the  claims  for  X  ray tubes  which have  completed  more  than 250000  exposures. So is it not an injustice  with  the  insured  when  we  consider  the life  of  X ray  tube  as 40000 exposures ( as  per  the  circular). If  the  data  is  collected  all over  India-  it can be  very well proved  that this  16 year old circular  is outdated  and  needs  revision along  with other  engineering  circulars. Due to improvement in technology certainly  present X ray tube last longer. Let us first find out  among  us. I request all  my  friends to  share  the  recent data for  the number  of  exposures X ray tube has lasted (for  the  claim handled  by  them). 

 Regards
SANJAY SURPURIYA.
PUNE. 
            
      
To unsubscribe from this group and stop receiving emails from it, send an email to insurance_surveyors+unsub...@googlegroups.com.
To post to this group, send email to insurance_surveyors@googlegroups.com.
Visit this group at https://groups.google.com/group/insurance_surveyors.
For more options, visit https://groups.google.com/d/optout.



--
B/Rgds

Piyush Siinghal
0-9811047785

--
You received this message because you are subscribed to the Google Groups "Insurance surveyors" group.
To unsubscribe from this group and stop receiving emails from it, send an email to insurance_surveyors+unsub...@googlegroups.com.
To post to this group, send email to insurance_surveyors@googlegroups.com.
Visit this group at https://groups.google.com/group/insurance_surveyors.
For more options, visit https://groups.google.com/d/optout.

--
You received this message because you are subscribed to the Google Groups "Insurance surveyors" group.
To unsubscribe from this group and stop receiving emails from it, send an email to insurance_surveyors+unsub...@googlegroups.com.
To post to this group, send email to insurance_surveyors@googlegroups.com.
Visit this group at https://groups.google.com/group/insurance_surveyors.
For more options, visit https://groups.google.com/d/optout.

raghavan kodoth

unread,
Mar 8, 2017, 1:45:52 AM3/8/17
to iiisla...@googlegroups.com, Insurance surveyors, insurance-su...@googlegroups.com, insurance...@googlegroups.com, iiisl...@googlegroups.com

In the present generation CTs, the exposure is counted in mA-Sec, some times only in Sec. There is no counter which will give number of exposures. In earlier CTs, the exposure time and x-ray current were much higher, hence, 40,000 scans of those times maybe equivalent to 2,00,000 or more slices in present generation CTs., in respect of x-ray tube exploitation. Over the years, the exposure time and current have been brought down considerably for each exposure, to reduce effect of radiation on patients. To get the exact picture you need to see the patient scan register in which the exposure time and current set for each exposure will be written. Collect the data for a whole year to know, how much it adds to per year (average). 

I have done such an exercise in the Malabar Cancer Centre, Kannur, Kerala, a couple of years ago, which revealed some interesting figures. 

I agree that we cannot go by the circular, since there is no exposure counter anyway. The software will show the total exposure time done so far for the x-ray tube. There is also a gantry rotation counter reading available in some of the machines.

K.P.Raghavan


On Wednesday, 8 March 2017 11:57 AM, 'chirag shah' via Indian Independent Insurance Surveyor And Loss Assessors Bengal <iiisla...@googlegroups.com> wrote:


Dear All,

I met one repairer cum dealer of CT scanner machine and discussed about average life span of X-ray tube. He informed that he has experience of 20 years in this field. Average life span of X-ray tube of CT scan machine is 4 lacs exposures. It may vary from Company to Company.

with regards,

CHIRAG SHAH

Mob. : 98252 69184



On Wednesday, 8 March 2017 11:34 AM, SANJAY SURPURIYA <sanjays...@gmail.com> wrote:


Dear All
            This  circular  is  almost  16 year old. At  that  time most  of  the  C.T. Scan machines were  single slice  machines. Due  to improvement  in technology now  new  machines  available  are up to 320 slices. 64  Slice  C.T. Scanners  are very common now. By  this  circular  the  value of  the  X ray tube after  40000 exposures  is  zero. If  we  consider  the average use  of  the  machine (  32 - 64 slice  machine) for  any  good  organisation  40000 exposures are  completed in 6 - 8 months  time. If  we  go by this circular  X ray tube  be  replaced  every  4-6 months. Is  it  the  case ??  I  have  seen a 64 slice C.T. Scanner  X Ray tube which has  completed 9 Lakh exposures  and  is  still   working. When this is pointed out   facts  are  side lined  by giving standard  reply-  always  there  are  exceptions. 

            Now  a  days C.T. Scanner ( 32 - 64 slice m/c)   X ray tube  working  for more  than 250000 exposures  is  very  common. Many  of  us  must  have  handled  the  claims  for  X  ray tubes  which have  completed  more  than 250000  exposures. So is it not an injustice  with  the  insured  when  we  consider  the life  of  X ray  tube  as 40000 exposures ( as  per  the  circular). If  the  data  is  collected  all over  India-  it can be  very well proved  that this  16 year old circular  is outdated  and  needs  revision along  with other  engineering  circulars. Due to improvement in technology certainly  present X ray tube last longer. Let us first find out  among  us. I request all  my  friends to  share  the  recent data for  the number  of  exposures X ray tube has lasted (for  the  claim handled  by  them). 

 Regards
SANJAY SURPURIYA.
PUNE. 
            
      
       

On Wednesday, 8 March 2017 09:16:08 UTC+5:30, Piyush Singhal wrote:
Here is the circular which can be found at  https://iib.gov.in/IIB/tac/ circulars/ecir2001.htm


Engg/Gen-16/2001/26                                                                  13th June, 2001
 
                        Re: El ectronic Equipment Insurance :
--
You received this message because you are subscribed to the Google Groups "Indian Independent Insurance Surveyor And Loss Assessors Bengal" group.
To unsubscribe from this group and stop receiving emails from it, send an email to iiisla-benga...@googlegroups.com.

For more options, visit https://groups.google.com/d/optout.


--
You received this message because you are subscribed to the Google Groups "Indian Independent Insurance Surveyor And Loss Assessors Bengal" group.
To unsubscribe from this group and stop receiving emails from it, send an email to iiisla-benga...@googlegroups.com.

raghavan kodoth

unread,
Mar 8, 2017, 1:46:32 AM3/8/17
to iiisla...@googlegroups.com, Insurance surveyors, insurance-su...@googlegroups.com, insurance...@googlegroups.com, iiisl...@googlegroups.com
In the present generation CTs, the exposure is counted in mA-Sec, some times only in Sec. There is no counter which will give number of exposures. In earlier CTs, the exposure time and x-ray current were much higher, hence, 40,000 scans of those times maybe equivalent to 2,00,000 or more slices in present generation CTs., in respect of x-ray tube exploitation. Over the years, the exposure time and current have been brought down considerably for each exposure, to reduce effect of radiation on patients. To get the exact picture you need to see the patient scan register in which the exposure time and current set for each exposure will be written. Collect the data for a whole year to know, how much it adds to per year (average). 

I have done such an exercise in the Malabar Cancer Centre, Kannur, Kerala, a couple of years ago, which revealed some interesting figures. 

I agree that we cannot go by the circular, since there is no exposure counter anyway. The software will show the total exposure time done so far for the x-ray tube. There is also a gantry rotation counter reading available in some of the machines.

K.P.Raghavan


On Wednesday, 8 March 2017 11:35 AM, SANJAY SURPURIYA <sanjays...@gmail.com> wrote:


Dear All
            This  circular  is  almost  16 year old. At  that  time most  of  the  C.T. Scan machines were  single slice  machines. Due  to improvement  in technology now  new  machines  available  are up to 320 slices. 64  Slice  C.T. Scanners  are very common now. By  this  circular  the  value of  the  X ray tube after  40000 exposures  is  zero. If  we  consider  the average use  of  the  machine (  32 - 64 slice  machine) for  any  good  organisation  40000 exposures are  completed in 6 - 8 months  time. If  we  go by this circular  X ray tube  be  replaced  every  4-6 months. Is  it  the  case ??  I  have  seen a 64 slice C.T. Scanner  X Ray tube which has  completed 9 Lakh exposures  and  is  still   working. When this is pointed out   facts  are  side lined  by giving standard  reply-  always  there  are  exceptions. 

            Now  a  days C.T. Scanner ( 32 - 64 slice m/c)   X ray tube  working  for more  than 250000 exposures  is  very  common. Many  of  us  must  have  handled  the  claims  for  X  ray tubes  which have  completed  more  than 250000  exposures. So is it not an injustice  with  the  insured  when  we  consider  the life  of  X ray  tube  as 40000 exposures ( as  per  the  circular). If  the  data  is  collected  all over  India-  it can be  very well proved  that this  16 year old circular  is outdated  and  needs  revision along  with other  engineering  circulars. Due to improvement in technology certainly  present X ray tube last longer. Let us first find out  among  us. I request all  my  friends to  share  the  recent data for  the number  of  exposures X ray tube has lasted (for  the  claim handled  by  them). 

 Regards
SANJAY SURPURIYA.
PUNE. 
            
      
       

On Wednesday, 8 March 2017 09:16:08 UTC+5:30, Piyush Singhal wrote:
Here is the circular which can be found at  https://iib.gov.in/IIB/tac/ circulars/ecir2001.htm


Engg/Gen-16/2001/26                                                                  13th June, 2001
 
                        Re: El ectronic Equipment Insurance :
--
You received this message because you are subscribed to the Google Groups "Indian Independent Insurance Surveyor And Loss Assessors Bengal" group.
To unsubscribe from this group and stop receiving emails from it, send an email to iiisla-benga...@googlegroups.com.
Reply all
Reply to author
Forward
0 new messages