One thing you can do is pay attention to how your leads are routed from
the 1260 to the SOFC. First off, ALWAYS use a 4-point connection to
minimize the effects of the leads. Secondly, ensure that the leads to
cathode and anode are run parallel to each other, and from the same side
into the testing chamber. We're able to get acceptable resolution on
cells with similar resistance to yours. I assume you are testing cells
with larger area, correct? What temperature are you operating at?
A
good resource is Barsoukov and J. Ross Macdonald's book "Impedance
Spectroscopy: Theory, experiment and applications". Fig 4.1.23 on p.
233 describes what I'm talking about, and the book holds a wealth of
other information you will find valuable.