Hi Mehrzad,Thank you for your email. Here are some suggestions I have. The noise at high frequency could be inductance from the electrodes as a result of passage of a potential. You can model that using a inductor in your circuit. Also, make sure that your electrodes are not touching and there is enough distance between them. This could be a reason for the noise as well. Is there a reason why your amplitude is 40 mV (rms) and not 10 mV (rms)?Eugene Engmann
On Tue, Apr 14, 2020 at 5:18 AM Mehrzad Soleimany <mehrzad....@gmail.com> wrote:
--Dear Friends,Sometimes in EIS measurements, at high frequencies, a noise-like curve shows up that it is a repeatable phenomenon. To what could it be related?I attached some of the EIS results. The tests were taken at different temperatures and in the air atmosphere from a solid oxide proton-conducting electrolyte, and the Ag paste was used as electrodes material. The applied voltage amplitude was 40 mV, and the equipment was a Solartron 1260 in combination with a Solartron1286.Thank you in advance for your attention
You received this message because you are subscribed to the Google Groups "Impedance Spectroscopy" group.
To unsubscribe from this group and stop receiving emails from it, send an email to impedance-spectroscopy+unsub...@googlegroups.com.
To unsubscribe from this group and stop receiving emails from it, send an email to impedance-spectro...@googlegroups.com.
To view this discussion on the web visit https://groups.google.com/d/msgid/impedance-spectroscopy/9ad56374-2d75-4465-a1e3-eb03442d14c4%40googlegroups.com.
--
You received this message because you are subscribed to the Google Groups "Impedance Spectroscopy" group.
To unsubscribe from this group and stop receiving emails from it, send an email to impedance-spectro...@googlegroups.com.
To view this discussion on the web visit https://groups.google.com/d/msgid/impedance-spectroscopy/d588cba8-e5f4-4775-bc38-a2fb0f2a23dd%40googlegroups.com.
To view this discussion on the web visit https://groups.google.com/d/msgid/impedance-spectroscopy/CAE_U5d75E4yeooSdGerQDcyHfw%2BWAiAe%3DCrONji1u5_b9_q3dw%40mail.gmail.com.
Hi Mehrzad,Most of the literature out there use 10 mV (rms) (it is small enough, preventing any damage to the surface of the electrode. Rather than take out the noise, see if you can fit it using an inductor. The reason is because the noise is part of the scan. Ask you professor to see if you can try a 10 unless there is a reason to stick with 40.Eugene
On Wed, Apr 15, 2020 at 7:46 AM Mehrzad Soleimany <mehrzad....@gmail.com> wrote:
Dear Eugene,I appreciate your kind answer. I completely agree with you.Actually, no. 40 mV was the amplitude my professor said that I have to choose. Are you asking it because 40 mV could lead to nonlinearity in comparison with 10 mV?And do you think there would be any problem if I model only the clean part and remove the noise-like curve at high frequency?
On Tuesday, April 14, 2020 at 8:09:29 PM UTC+4:30, Eugene Engmann wrote:
Hi Mehrzad,Thank you for your email. Here are some suggestions I have. The noise at high frequency could be inductance from the electrodes as a result of passage of a potential. You can model that using a inductor in your circuit. Also, make sure that your electrodes are not touching and there is enough distance between them. This could be a reason for the noise as well. Is there a reason why your amplitude is 40 mV (rms) and not 10 mV (rms)?Eugene Engmann
On Tue, Apr 14, 2020 at 5:18 AM Mehrzad Soleimany <mehrzad....@gmail.com> wrote:
--Dear Friends,Sometimes in EIS measurements, at high frequencies, a noise-like curve shows up that it is a repeatable phenomenon. To what could it be related?I attached some of the EIS results. The tests were taken at different temperatures and in the air atmosphere from a solid oxide proton-conducting electrolyte, and the Ag paste was used as electrodes material. The applied voltage amplitude was 40 mV, and the equipment was a Solartron 1260 in combination with a Solartron1286.Thank you in advance for your attention
You received this message because you are subscribed to the Google Groups "Impedance Spectroscopy" group.
To unsubscribe from this group and stop receiving emails from it, send an email to impedance-spectroscopy+unsub...@googlegroups.com.
To view this discussion on the web visit https://groups.google.com/d/msgid/impedance-spectroscopy/9ad56374-2d75-4465-a1e3-eb03442d14c4%40googlegroups.com.
--
You received this message because you are subscribed to the Google Groups "Impedance Spectroscopy" group.
To unsubscribe from this group and stop receiving emails from it, send an email to impedance-spectroscopy+unsub...@googlegroups.com.
Are you sure it is not related to the bandwidth of the 1286?. It looks to me you are trying too high a frequency start for your equipment combo.Greadey
On Wed, 15 Apr 2020, 16:38 Eugene Engmann, <englis...@gmail.com> wrote:
Hi Mehrzad,Most of the literature out there use 10 mV (rms) (it is small enough, preventing any damage to the surface of the electrode. Rather than take out the noise, see if you can fit it using an inductor. The reason is because the noise is part of the scan. Ask you professor to see if you can try a 10 unless there is a reason to stick with 40.Eugene
On Wed, Apr 15, 2020 at 7:46 AM Mehrzad Soleimany <mehrzad....@gmail.com> wrote:
Dear Eugene,I appreciate your kind answer. I completely agree with you.Actually, no. 40 mV was the amplitude my professor said that I have to choose. Are you asking it because 40 mV could lead to nonlinearity in comparison with 10 mV?And do you think there would be any problem if I model only the clean part and remove the noise-like curve at high frequency?
On Tuesday, April 14, 2020 at 8:09:29 PM UTC+4:30, Eugene Engmann wrote:
Hi Mehrzad,Thank you for your email. Here are some suggestions I have. The noise at high frequency could be inductance from the electrodes as a result of passage of a potential. You can model that using a inductor in your circuit. Also, make sure that your electrodes are not touching and there is enough distance between them. This could be a reason for the noise as well. Is there a reason why your amplitude is 40 mV (rms) and not 10 mV (rms)?Eugene Engmann
On Tue, Apr 14, 2020 at 5:18 AM Mehrzad Soleimany <mehrzad....@gmail.com> wrote:
--Dear Friends,Sometimes in EIS measurements, at high frequencies, a noise-like curve shows up that it is a repeatable phenomenon. To what could it be related?I attached some of the EIS results. The tests were taken at different temperatures and in the air atmosphere from a solid oxide proton-conducting electrolyte, and the Ag paste was used as electrodes material. The applied voltage amplitude was 40 mV, and the equipment was a Solartron 1260 in combination with a Solartron1286.Thank you in advance for your attention
You received this message because you are subscribed to the Google Groups "Impedance Spectroscopy" group.
To unsubscribe from this group and stop receiving emails from it, send an email to impedance-spectroscopy+unsub...@googlegroups.com.
To view this discussion on the web visit https://groups.google.com/d/msgid/impedance-spectroscopy/9ad56374-2d75-4465-a1e3-eb03442d14c4%40googlegroups.com.
--
You received this message because you are subscribed to the Google Groups "Impedance Spectroscopy" group.
To unsubscribe from this group and stop receiving emails from it, send an email to impedance-spectroscopy+unsub...@googlegroups.com.
To view this discussion on the web visit https://groups.google.com/d/msgid/impedance-spectroscopy/d588cba8-e5f4-4775-bc38-a2fb0f2a23dd%40googlegroups.com.
--
You received this message because you are subscribed to the Google Groups "Impedance Spectroscopy" group.
To unsubscribe from this group and stop receiving emails from it, send an email to impedance-spectroscopy+unsub...@googlegroups.com.
--
You received this message because you are subscribed to the Google Groups "Impedance Spectroscopy" group.
To unsubscribe from this group and stop receiving emails from it, send an email to impedance-spectro...@googlegroups.com.
To view this discussion on the web visit https://groups.google.com/d/msgid/impedance-spectroscopy/a3c3a521-62e0-4ad9-8283-25f7e1d19805%40googlegroups.com.