Although it is generally desirable to avoid multiple expressions for the same concept (i.e. “Non-invasive ventilation has been selected”) it is not the end of the world as far as interoperability goes, and nobody has expressed major concerns about it (as of yesterday).
Vendor | VendorMode | ISOMode | SelAdjuncts | TC | CC | LC | NIV | SIGH | TRG | FGF | Notes | NoteDetails |
GE | NIV | CSV-PS <ACAP> (NIV) | TRG | ○ | ○ | ○ | ■ | ○ | ● | ○ |
|
|
ISO 19923 §3.1.15
NIV
non-invasive ventilation
positive-pressure ventilation without the use of an invasive airway device
Note 1 to entry: The connection to the patient is typically by means of a specially designed face or nasal mask.
Note 2 to entry: The provision of NIV does not, in itself, require dedicated ventilation-modes although some might be more suited to its use, but it does typically require certain compensating measures, particularly those relating to the possibility of increased and variable leakage. These can include, added, extended or deactivated compensations, modified alarms limits, the deactivation of some alarms and the modification of inflation initiation and termination criteria.
Note 3 to entry: On ventilators intended for NIV only, these compensations are classified as a permanently active NIV adjunct. On ventilators where the compensations made to suit NIV are selectable as an option, although also adjunctive in their actions, these have been typically classified as an NIV ventilator operational mode.
RTM Webex Meeting Information
Meeting number (access code): 145 134 2502
Meeting password: Devices1
Occurs every Wednesday effective Wednesday, September 2, 2020 from 11:00 AM to 12:00 PM, (UTC-05:00) Central Time (US & Canada) 11:00 am | (UTC-05:00) Central Time (US & Canada) | 1 hr
Join meeting
Tap to join from a mobile device (attendees only) 1-866-469-3239,,1451342502## Call-in toll-free number (US/Canada) +1-650-429-3300,,1451342502## Call-in number (US/Canada)
Join by phone 1-866-469-3239 Call-in toll-free number (US/Canada) 1-650-429-3300 Call-in number (US/Canada) Toll-free calling restrictions
Join from a video system or application Dial 14513...@ihe.webex.com You can also dial 173.243.2.68 and enter your meeting number.
Join using Microsoft Lync or Microsoft Skype for Business Dial 1451342...@lync.webex.com
On Jun 22, 2021, at 4:51 PM, Paul Schluter <pssch...@gmail.com> wrote:
RTM Webex Meeting Information
Meeting number (access code): 145 134 2502
Table 1: This table lists vendor modes for Draeger, Getinge and GE, along with mappings to the ISO 19223 mode string and IEEE 11073 REFID. This table will be included in IEEE P11073-10101b and can be extended with contributions from other vendors.
Selectable Adjuncts that are permanently associated with the VendorMode are pre-coordinated into the ISOMode string and the IEEE 11073 REFID. Selectable Adjuncts that are conveyed as Boolean flags are listed in the SelAdjuncts column. Receivers shall construct the logical-OR of ISOMode and SelAdjuncts to determine the overall Selectable Adjunct state and, if needed, to construct the overall canonical ISO Mode string.
The { TC, CC, LC, NIV, SIGH, TRG, FGF } columns provide a Boolean flag bit view of the Selectable Adjuncts for each VendorMode that are strictly derived from ISOMode (indicated by ⦿) and SelAdjuncts (indicated by ●).
The Notes and NoteDetails columns list the notes that are associated with each mode. The original vendor note reference numbers have been consolidated as Notes starting at 10, reserving Notes 1 - 9 for future vendor-independent Notes. For traceability, the original vendor note numbers are listed in the vNote column.
Column |
Description/Notes |
Vendor |
Vendor [and optional Model suffix, e.g. GE-R860] |
VendorMode |
Vendor mode |
VendorModeShortDescription |
Vendor mode short description (not shown, need Draeger info) |
ISOMode |
ISO 19223 string, including SelAdjuncts and consolidated Note numbers associated with mode |
REFID |
IEEE 11073 REFID |
CODE10 |
CODE10 (final PART, CODE10 and CF_CODE10 assigned on July 23 for RTMMSv2) |
SelAdjuncts |
Selectable Adjuncts conveyed as Boolean flags |
TC |
Tube compensation (output only) |
CC |
Compliance compensation (output only) |
LC |
Leakage compensation (output only) |
NIV |
Non Invasive Ventilation (output only) |
SIGH |
SIGH mode (Draeger) (output only) |
TRG |
Trigger compensation (GE) (output only) |
FGF |
Fresh gas flow (GE) (output only) |
vNotes |
Original Vendor Note reference numbers (will not appear in standard) |
Notes |
(0..*) Note References (white-space separated consolidated Note numbers) |
NoteDetails |
(0..*) Note Details (⬩ separated strings) |
Table 2: This table lists the ISOMode and equivalent IEEE 11073 REFID, and includes Selectable Adjuncts and consolidated Note numbers permanently associated with the mode. This table will be included as normative content in IEEE P11073-10101b and can be extended.
Column |
Description/Notes |
ISOMode |
ISO 19223 string, including SelAdjuncts and consolidated Note numbers associated with mode |
REFID |
IEEE 11073 REFID |
CODE10 |
CODE10 (final PART, CODE10 and CF_CODE10 assigned on July 23 for RTMMSv2) |
Notes |
(0..*) Note References (white-space separated consolidated Note numbers) |
NoteDetails |
(0..*) Note Details (⬩ separated strings) |
Table 3: This table lists the consolidated Note number and details associated with the Note. This table will be included as normative content in IEEE P11073-10101b and can be extended.
Column |
Description/Notes |
Note |
Note Reference number (consolidated) |
NoteDetails |
Note Detail |
On Jul 14, 2021, at 9:13 AM, <cha...@interopehealth.com> <cha...@InteropeHealth.com> wrote:Paul,I will unfortunately be in flight between the USA and France at the time of our t-con.I have one input for the group:When looking at all of those ventilation modes and how they map to the various vendor modes labelling, I see great analysis and progress. However, this does not connect with the most important goal which is to express the enumerated values for the ventilation mode in a standardized way.In IEEE 11073, there is a practical approach with a “32 bit map”, where one bit represents each elementary mode. That way combining mode, when applicable can be easily represented. By assigning REFID to each mode is fine, but assigning one bit in that 32 bit array is equally important. The current list in IEEE 11073 2019 draft is:vent-mode-spont(0),
vent-mode-cpap(1),
vent-mode-bipap(2),
vent-mode-ippw(3),
vent-mode-cmv(4),
vent-mode-irv(5),
vent-mode-imv(6),
vent-mode-simv(7),
vent-mode-insp-assist(8),
vent-mode-press-release(9),
vent-mode-psv(10),
vent-mode-mmv(11),
vent-mode-prop-assist(12),
vent-mode-hfv(13),
vent-mode-hfjv(14),
vent-mode-hfo(15),
vent-mode-peep(31)In your table I see some of those modes, but some are never mentioned, others are not listed.Am I getting concern for a non-issue ?If the group can address this point, this would be great.Now leaving for the airport.CharlesDe : ihepc...@googlegroups.com <ihepc...@googlegroups.com> De la part de Paul Schluter
Envoyé : mercredi 14 juillet 2021 09:56
À : ihe-p...@googlegroups.com; ihe-p...@googlegroups.com; PCD Technical Committee GG <ihepc...@googlegroups.com>; ihe-pcd-...@googlegroups.com; IHE-PCD-DPI-WG <ihe-p...@googlegroups.com>; device...@lists.hl7.org
Cc : Monroe Pattillo <monroe_...@bellsouth.net> (monroe_...@bellsouth.net) <monroe_...@bellsouth.net>; Mathieu Roullet <mrou...@enovacom.fr>; Stefan Schlichting <stefan.sc...@ornet.org>; Dalibor Pokrajac <Dalibor....@guardrfid.com>; STEVEN DAIN <sd...@rogers.com>; Jan Wittenber <jan.wi...@gmail.com>; Walsh John L.M.D. <john....@mgh.harvard.edu>; Abhyankar, Swapna <sabh...@regenstrief.org>; Paul Sherman <paulrsh...@gmail.com>; Garguilo, John J. <john.g...@nist.gov>; Crouzier, Nicolas <nicolas....@nist.gov>; Todd Cooper <toddco...@gmail.com>; John Rhoads <johnr...@johnrhoads.net>; Chris Courville <ccou...@epic.com>; michael.faughn <michael...@prometheuscomputing.com>; Tom.Ko...@bbraun.com; Norman Jones <nsjo...@btinternet.com>; Hassing, Kai <kai.h...@philips.com>; Klotz, Tobias <Tobias...@draeger.com>; Ken Fuchs <kenj....@gmail.com>; Gisch, Simon <Simon...@draeger.com>; Abell, Josh <Josh....@draeger.com>; Staudenmaier, Greg (Stod) <Greg.Sta...@va.gov>; Tracy Rausch <tr...@docboxinc.com>; Daidi Zhong <daidi...@ieee.org>; Michael J. Kirwan <mki...@dsheet.com>; Lisa Perry <l.p...@ieee.org>; Sabo, Perry (Carefusion, non-GE) <perry...@carefusion.com>; di...@moberg.com; Marks, Kenneth E <KEMa...@madisoncollege.edu>; Mike Krajnak <michael...@med.ge.com>; Neal Seidl <Neal....@med.ge.com>; Comunale, Carter <ccom...@bernoullihealth.com>; Szente, Andras <asz...@bernoullihealth.com>; Iglesias, Brian <bigl...@bernoullihealth.com>; Hadi Rahemi <rah...@circulationconcepts.com>; shimp...@marshfieldresearch.org; Andrew Norton <andrew...@tiscali.co.uk>; Tietz, Michael <Michae...@draeger.com>; Beier, Manfred <Manfre...@draeger.com>; Per Johansson <per.jo...@getinge.com>; Anders Häggström <anders.h...@getinge.com>; Mikael Petrini <mikael....@getinge.com>; Jeff Rinda <jeffre...@gmail.com>; Nichols, Steven (GE Healthcare) <Steven....@ge.com>; Ramachandran, Karthik R (GE Healthcare) <r.karthikr...@med.ge.com>; Brauer, Kimberly (GE Healthcare) <Kimberl...@ge.com>; snich...@gmail.com; Paul Schluter <pssch...@gmail.com>
Objet : [ihepcdtech:3158] Please join our RTM tcon today, Wednesday, July 14th, starting at 11:00 AM CDT - consolidated ventilator mode lists for Draeger, Getinge and GE
--
You received this message because you are subscribed to the Google Groups "IHE PCD Technical Committee" group.
To unsubscribe from this group and stop receiving emails from it, send an email to ihepcdtech+...@googlegroups.com.
To view this discussion on the web visit https://groups.google.com/d/msgid/ihepcdtech/4C8BCCD1-B517-474A-8A17-CFED5F5D4B10%40gmail.com.
(1) only be pre-coordinated into the main ISOMode string,(2) only be Boolean flag in SelAdjuncts, or(3) represented either way, using the logical-OR of ISOMode and SelAdjuncts to determine the effective { TC, CC, LC, NIV, SIGH, TRG, FGF } Selectable Adjuncts.
Begin forwarded message:
From: Steven Dain <sd...@rogers.com>
Subject: Re: [IHE PCD ACM:1469] Please join our RTM tcon today, Wednesday, July 14th, starting at 11:00 AM CDT - consolidated ventilator mode lists for Draeger, Getinge and GE
Date: July 20, 2021 at 9:37:14 PM CDT
To: Paul Schluter <pssch...@gmail.com>, Norman Jones <n...@normansjones.com>, Paul Dixon <paul....@vyaire.com>, Anders Häggström <anders.h...@getinge.com>, Manfred Beier <manfre...@draeger.com>
Cc: "Tietz, Michael" <Michae...@draeger.com>, "Andersen, Bjoern" <bjoern....@draeger.com>, "Fuchs, Ken" <Ken....@draeger.com>, Paul Schluter <pssch...@gmail.com>
Reply-To: Steven Dain <sd...@rogers.com>
I'm ok if everyone indicates niv in the selAdjuncts instead of making it mandatory in the ISOMode. We need one consistent place to identify that niv is being used. Risk management and safety considerations dictate that we must be prescriptive in this use case due to the serious hazards if niv and invasive ventilation get mixed up, particularly as many of the niv devices are managed remotely.
I do have a meeting at the same time but I'll try to join your meaning once I see the agenda for the other.
Best regards,
Steven
Please excuse the typosOn Tue, Jul 20, 2021 at 15:49, Paul Schluter
<pssch...@gmail.com> wrote:
Steven and colleagues,
Although I agree with you that three sources of NIV information {ISOMode, SelAdjuncts and Operational Mode} may be one too many, we also need to avoid being overly prescriptive.
In my opinion, removing or ignoring “NIV” in Operational Mode to obtain ventilation mode is a sensible idea; we should only use ISOMode and SelAdjuncts to convey overall ventilator mode. This pair of observations (or settings) is very similar to how existing ventilation mode is sent today.
Using ISOMode and SelAdjuncts, and combining the Selectable Adjunct information using a logical-OR operation, will remain “as is” - it provides an additional degree of flexibility for vendors to express their ventilation modes using ISO 19223 in a manner consistent with their existing design conventions, philosophies and documentation.
That said, of all the Selectable Adjuncts we have considered so far, requiring that (NIV) be expressed in the main systematic ISOMode (rather than as a SelAdjuncts Boolean flag) would also be a sensible idea. In most of the systems I have reviewed, a non-trivial number of ventilator characteristics are associated with (NIV) operation, whereas most of the other Selectable Adjuncts have far fewer settings associated with them.
The only downside when including (NIV) in the main systematic ISOMode is that it increases the number of permutations. Although this is not an issue for Getinge and GE, it increases the number of distinct ISOModes (and IEEEModes and associated REFIDs and numeric codes) from 35 to 62 distinct modes, out of a total of 136 modes if all Selectable Adjuncts were pre-coordinated in the ISOMode and IEEEMode strings.
I have attached an Excel file that illustrates this, starting with the current total of 35 distinct Draeger ISO/IEEE Modes when all Selectable Adjuncts are conveyed as Boolean flags in SelAdjunct (columns vISOMode and vSelAdjuncts) which increases to 62 distinct Draeger ISO/IEEE Modes if (NIV) is conveyed by the main systematic ISOMode string (columns nvISOMode and nvSelAdjuncts). [This uses the original Draeger vNote numbers.]
This represents a nearly 80% increase in the number of ISO/IEEEModes required to support the current list of Draeger modes. Although this is not a big deal of computers and programmers, it is a lot more information for review and consumption, especially with the numerous duplicated base concepts.
We can discuss this during tomorrow’s (7/21) RTM tcon starting at 11:00 AM CDT. I believe it is essential that representatives from Draeger attend as well as Steven Dain attend, so if tomorrow at 11:00 CDT AM doesn’t work out, an alternative time and teleconference service will need to be arranged or we will just have to continue this discussion by email.
<<ISO-Filter.3j.2021-07-20T12.Draeger-NIV-trimmed.xls>>On Jul 19, 2021, at 9:38 AM, Steven Dain <sd...@rogers.com> wrote:
Is NIV a function of a ventilator that provides ancillary actions to a ventilator mode? (19223 3.11.4)
ancillary=providing necessary support to the primary activities or operation of an organization, institution, industry, or system.
or is NIV the primary activity?
Sorry for the profound philosophical question on a Monday
My concern is
"The upshot of all of this is that the indication of NIV can be done several ways, in accordance with ISO 19223 while also supporting some vendor preferences. The only downside to supporting multiple representations is that determination of whether NIV is in use or not requires that the receiving application must examine three sources of this information: ISOMode, SelAdjuncts and Operational Mode for the presence of ‘NIV’ in any.[SD1] "
I don't like this downside. We need to make a decision and choose one of " ISOMode, SelAdjuncts or Operational Mode". I think will need to be practical here rather than philosophical. I'll leave it up to the manufacturers to choose the one. Personally I like the (NIV) in the ISOMode string and would include it in the Draeger ISOModes, like you did for the Getinge and GE.
I'll request Nihon Koden, Medtronic, Thornhill Research and Vyaire to complete their tables once we have settled on this issue. Paul Dixon will have time now that he has finished ISO 4135
Please find my attached document for a few other, I think non-controversial edits.
Thanks for this Paul S! Well done.
Steven Dain
1. The <ISOMode> string and associated IEEE 11073 <REFID> only include Selectable Adjuncts that are considered ‘inherent’ to the associated mode by the vendor, and <SelAdjuncts> includes all Selectable Adjuncts, both ‘inherent’ to the mode and conveyed as Boolean flags.2. A <VendorModeShortDescription> is provided for each distinct ISOMode, and in some cases, with combinations of Selectable Adjuncts expressed as Boolean flags.3. Row order is specified by the vendor. This includes rows having the same ISOMode but different combinations of Selectable Adjuncts expressed as Boolean flags. This allows the modes to be listed in a sensible clinical and tutorial order and will simplify review and updates later on.4. At the present time, we will not require that vendors provide ISOMode mappings for the { ApneaVentilation, BackupVentilation, or FailSafeVentilation } states conveyed by MDC_VENT_OPERATIONAL_STATE. This information has been difficult to obtain and it is often not conveyed with the clarity of the OperationalMode, which has been the primary focus of the ISO 19223 and IEEE 11073 efforts.
On Jul 14, 2021, at 8:55 AM, Paul Schluter <pssch...@gmail.com> wrote: August 7 updates shown with red font
Greetings!Please join our RTM tcon this Wednesday, July 14th, starting at 11:00 AM CDT.I am very happy to say that we have resolved the remaining issues regarding the use of Notes for representing <ACAP*>. The XSLT has been updated to create the penultimate set of tables that will define the IEEE P11073-10101b Group6 ventilator mode terms based on the ISO 19223 ventilator mode syntax and vocabulary.
The material is based on the substantial contributions by Draeger, Getinge and GE Healthcare and has undergone significant review by the vendors and the leadership of the ISO 19223 standard, especially by Drs. Norman Jones and Steven Dain, MD. I would also like to thank Anders Häggström of Getinge regarding how <ACAP*> can be represented. I plan to provide an overview of this effort during today's RTM call. This work demonstrates that "we can do the impossible” given enough time, talent and energy. I would also like to thank Thomas Kreuger and Michi Tietz (Draeger), Anders Häggström (Getinge) and Kim Brauer and Mike Krajnak (GE) for their contributions regarding <VendorModeShortDescription>.
Here’s a synopsis of the primary content in the attached set of three HTML tables:
Table 1: This table lists vendor modes for Draeger, Getinge and GE, along with mappings to the ISO 19223 mode string and IEEE 11073 REFID. This table will be included in IEEE P11073-10101b and can be extended with contributions from other vendors.
Selectable Adjuncts that are permanently associated with the VendorMode are pre-coordinated into the ISOMode string and the IEEE 11073 REFID. All Selectable Adjuncts, both ‘inherent’ to the mode and that are conveyed as Boolean flags are listed in the SelAdjuncts column.
The { TC, CC, LC, NIV, SIGH, TRG, FGF } columns provide a Boolean flag bit view of the Selectable Adjuncts for each VendorMode that are pre-coordinated in the ISOMode (indicated by ⦿) or sent as Boolean flags (indicated by ●).
The Notes and NoteDetails columns list the notes that are associated with each mode. The original vendor note reference numbers have been consolidated as Notes starting at 10, reserving Notes 1 - 9 for future vendor-independent Notes. For traceability, the
originalprevious vendor note numbers are listed in the vNote column.
Column
Description/Notes
Vendor
Vendor [and optional Model suffix, e.g. GE-R860]
VendorMode
Vendor mode
VendorModeShortDescription
Vendor mode short description
ISOMode
ISO 19223 string, including ‘inherent’ Selectable Adjuncts and consolidated Note numbers
REFID
IEEE 11073 REFID
CODE10
CODE10 (final PART, CODE10 and CF_CODE10 assigned at end of August for RTMMSv2)
SelAdjuncts
All Selectable Adjuncts, both ‘inherent’ to the mode and conveyed as Boolean flags
TC
Tube compensation (output only)
CC
Compliance compensation (output only)
LC
Leakage compensation (output only)
NIV
Non Invasive Ventilation (output only)
SIGH
SIGH mode (Draeger) (output only)
TRG
Trigger compensation (GE) (output only)
FGF
Fresh gas flow (GE) (output only)
vNotes
Original Vendor Note reference numbers (will not appear in standard)
Notes
(0..*) Note References (white-space separated consolidated Note numbers)
NoteDetails
(0..*) Note Details (⬩ separated strings)
Table 2: This table lists the ISOMode and equivalent IEEE 11073 REFID, and includes Selectable Adjuncts and consolidated Note numbers permanently associated with the mode. This table will be included as normative content in IEEE P11073-10101b and can be extended.
Column
Description/Notes
ISOMode
ISO 19223 string, including ‘inherent’ Selectable Adjuncts and consolidated Note numbers
REFID
IEEE 11073 REFID
CODE10
CODE10 (final PART, CODE10 and CF_CODE10 assigned at end of August for RTMMSv2)
Notes
(0..*) Note References (white-space separated consolidated Note numbers)
NoteDetails
(0..*) Note Details (⬩ separated strings)
Table 3: This table lists the consolidated Note number and details associated with the Note. This table will be included as normative content in IEEE P11073-10101b and can be extended.
Column
Description/Notes
Note
Note Reference number (consolidated)
NoteDetails
Note Detail
In addition to the three tables, most of the material in the <<VentModeISO.5x…docx>> series of write-ups and other supporting material will be adapted and included in IEEE P11073-10101b to provide a largely stand-alone treatment of this set of terms, along with references to key sections of the ISO 19223 standard.Due to summer vacations and other schedule conflicts, some of you may not be able to attend today’s meeting at 11:00 AM CDT. I will be happy to schedule an additional meeting at a different date and time.Thanks and regards,Paul Schluteremail: pssch...@gmail.comcell: (414) 702-2026
RTM Webex Meeting Information
Meeting number (access code): 145 134 2502
On Aug 7, 2021, at 12:38 PM, Paul Schluter <pssch...@gmail.com> wrote:Greetings!I am happy to say that we now have <VendorModeShortDescriptions> for all three vendors and these have been incorporated into the attached HTML tables for the 10101b Group6 ventilator modes. The tables also incorporate other changes and additions since the previous ventilator mode listing that I sent to you on July 14. The most notable changes include:1. The <ISOMode> string and associated IEEE 11073 <REFID> only include Selectable Adjuncts that are considered ‘inherent’ to the associated mode by the vendor, and <SelAdjuncts> includes all Selectable Adjuncts, both ‘inherent’ to the mode and conveyed as Boolean flags.2. A <VendorModeShortDescription> is provided for each distinct ISOMode, and in some cases, with combinations of Selectable Adjuncts expressed as Boolean flags.3. Row order is specified by the vendor. This includes rows having the same ISOMode but different combinations of Selectable Adjuncts expressed as Boolean flags. This allows the modes to be listed in a sensible clinical and tutorial order and will simplify review and updates later on.4. At the present time, we will not require that vendors provide ISOMode mappings for the { ApneaVentilation, BackupVentilation, or FailSafeVentilation } states conveyed by MDC_VENT_OPERATIONAL_STATE. This information has been difficult to obtain and it is often not conveyed with the clarity of the OperationalMode, which has been the primary focus of the ISO 19223 and IEEE 11073 efforts.Although <VendorModeShortDescription> will be useful for searching and providing additional information about the ventilation modes for each vendor, it may be difficult to include this highly vendor-specific information in the published IEEE 11073 and ISO/IEEE 11073 standards later on, due to the reluctance of SDOs to include vendor-specific information in the standard. That said, the Vendor and VendorModes will be included in the published standards and we certainly plan to include <VendorModeShortDescription> in RTMMSv2, where it would be searchable.Please take this opportunity to review this material over the next few weeks. After I return from vacation, we could consider having an RTM tcon on Wednesday, August 18th, or possibly later, depending on everyone’s schedules. We should be in a good position to assign final numeric codes by the end of August, depending on any feedback we receive. In the meantime, exchanging email comments will work just fine.Thanks for all your help, and best regards,Paul Schlutercell: (414) 702-2026<<ISO-Filter.2d.x3m.2021-08-05T21.html>>
On Aug 4, 2021, at 9:11 AM, Paul Schluter <pssch...@gmail.com> wrote:
Greetings!The RTM tcons for today, Wednesday, August 4th, and next Wednesday, August 11th, are cancelled. I have not received a final confirmation of the <VendorModeShortDescriptions> from one vendor and I will be on vacation next week in Door County.With so many people out on vacation this month, it makes it nearly impossible to complete multiple internal reviews, especially on something complex like ventilator mode.
Best regards, and I will see you in a few weeks!
On Aug 18, 2021, at 8:42 AM, <cha...@interopehealth.com> <cha...@InteropeHealth.com> wrote:
Paul,Thanks so much for the great work and making you available in the middle of the summer.Four questions :
- I do not see the REFID and CF_CODE10 specified for the MDC_VENT_MODE_ISO_19223 and for the MDC_VENT_MODE_SEL_ADJUNCTS_ISO_19223. Where is it in the attachments ? The final codes have not been assigned since we have an open item (should FGF be a Selectable Adjunct or not) and possibly others to resolve, probably after next week’s 8/25 RTM tcon due to schedule/vacation conflicts.
- IEEE 11073 seems incomplete in specifying the way the 32 BitMap was encoded. For interoperability it is critical to be explicit between IHE DEC Profile and IEEE 11073. How will the Selected Adjunct bitmap be encoded ? As a series of 7 digits with a 0 or 1 value ? As an integer value between 0 and 255 corresponding to the 7 bits out of a byte (extensible) ? Other ? Should we plan for the addition of other adjuncts in the future? Can private bit be privately added ? We could assign ASN.1 style bit names and numbers, in addition to the short tokens ‘TC', ‘CC', ‘LC', ‘NIV', ‘SIGH', ‘TRG', ‘FGF’ that we already have. The vast majority if IHE PCD DEC HL7 V2 messages use the short token representation, and this information is captured in the hRTM “harmonized Rosetta” along with any synonyms, e.g. tokens, ASN.1-bits, OIDs.
- Several vendor specific ventilation modes include values such as “standby” or “neonatal/pediatric/adult”. Are those included in the newly defined REFIDs for ISOModes ? I do not see them. ’Standby’ is included as a special mode, along with several other special modes, in the PEGjs [5e] in the write-up I sent you earlier. I believe we have a generic device "neonatal/pediatric/adult” selection, and we do have a generic sensor classification for "neonatal/pediatric/adult”. Currently there are no plans to include it in the ISO 19223 and IEEE 11073 vent mode REFIDs unless that is everyone’s preference.
- Will the vendor notes be included in the standard ? It seems necessary as the vendor note numbers appear in the REFID. Yes, they will, in an extensible table similar to HTML Table 3 in the listing.
Thanks so much.Charles
Envoyé : mercredi 18 août 2021 14:54
À : ihe-p...@googlegroups.com; ihe-p...@googlegroups.com; PCD Technical Committee GG <ihepc...@googlegroups.com>; ihe-pcd-...@googlegroups.com; IHE-PCD-DPI-WG <ihe-p...@googlegroups.com>; device...@lists.hl7.org
Cc : Monroe Pattillo <monroe_...@bellsouth.net> (monroe_...@bellsouth.net) <monroe_...@bellsouth.net>; Mathieu Roullet <mrou...@enovacom.fr>; Stefan Schlichting <stefan.sc...@ornet.org>; Dalibor Pokrajac <Dalibor....@guardrfid.com>; STEVEN DAIN <sd...@rogers.com>; Jan Wittenber <jan.wi...@gmail.com>; Walsh John L.M.D. <john....@mgh.harvard.edu>; Abhyankar, Swapna <sabh...@regenstrief.org>; Paul Sherman <paulrsh...@gmail.com>; Garguilo, John J. <john.g...@nist.gov>; Crouzier, Nicolas <nicolas....@nist.gov>; Todd Cooper <toddco...@gmail.com>; John Rhoads <johnr...@johnrhoads.net>; Chris Courville <ccou...@epic.com>; michael.faughn <michael...@prometheuscomputing.com>; Tom.Ko...@bbraun.com; Norman Jones <nsjo...@btinternet.com>; Hassing, Kai <kai.h...@philips.com>; Klotz, Tobias <Tobias...@draeger.com>; Ken Fuchs <kenj....@gmail.com>; Gisch, Simon <Simon...@draeger.com>; Abell, Josh <Josh....@draeger.com>; Staudenmaier, Greg (Stod) <Greg.Sta...@va.gov>; Tracy Rausch <tr...@docboxinc.com>; Daidi Zhong <daidi...@ieee.org>; Michael J. Kirwan <mki...@dsheet.com>; Lisa Perry <l.p...@ieee.org>; Sabo, Perry (Carefusion, non-GE) <perry...@carefusion.com>; di...@moberg.com; Marks, Kenneth E <KEMa...@madisoncollege.edu>; Mike Krajnak <michael...@med.ge.com>; Neal Seidl <Neal....@med.ge.com>; Comunale, Carter <ccom...@bernoullihealth.com>; Szente, Andras <asz...@bernoullihealth.com>; Iglesias, Brian <bigl...@bernoullihealth.com>; Hadi Rahemi <rah...@circulationconcepts.com>; shimp...@marshfieldresearch.org; Andrew Norton <andrew...@tiscali.co.uk>; Tietz, Michael <Michae...@draeger.com>; Beier, Manfred <Manfre...@draeger.com>; Per Johansson <per.jo...@getinge.com>; Anders Häggström <anders.h...@getinge.com>; Mikael Petrini <mikael....@getinge.com>; Jeff Rinda <jeffre...@gmail.com>; Nichols, Steven (GE Healthcare) <Steven....@ge.com>; Ramachandran, Karthik R (GE Healthcare) <r.karthikr...@med.ge.com>; Brauer, Kimberly (GE Healthcare) <Kimberl...@ge.com>; snich...@gmail.com; Peeters, Wouter <w.pe...@philips.com>; Paul Schluter <pssch...@gmail.com>
Objet : [ihepcdtech:3165] Please join our RTM tcon today, Wednesday, August 18 - Q&A and general discussion regarding ventilator mode and related topics
--
You received this message because you are subscribed to the Google Groups "IHE PCD Technical Committee" group.
To unsubscribe from this group and stop receiving emails from it, send an email to ihepcdtech+...@googlegroups.com.
To view this discussion on the web visit https://groups.google.com/d/msgid/ihepcdtech/E3491561-7E57-4055-8275-C7BC9A21E259%40gmail.com.
Going forward, all “compensation” adjuncts will uniformly end with “C”, maintaining the convention established by the original ISO 19223 ’TC’, ‘CC’ and ‘LC’ tokens. [This change affects some of the <ISOModes> and <SelAdjuncts> for Getinge and GE.]
Since we are a reasonably certain that most, if not all, contemporary anesthesia machines either ‘compensate’ for FGF or, in the case of several Draeger models, use fresh gas decoupling, we should make the compensation or diversion of fresh gas flow the default expectation, and a Selectable Adjunct would not be shown in the common case. If we later encounter an anesthesia machine that does not compensate nor eliminate this error, then the negative ’noFGFC’ would be used as the Selectable Adjunct to indicate this.This modification only affects some of the <ISOModes> and <SelAdjuncts> for GE The FGF and (FGF) indications for the GE-CS750 have been removed, and all three vendors (Draeger, Getinge and GE) are now consistent in this regard since they compensate for or otherwise divert the fresh gas flow so that the delivered volume is correct. [Pending confirmation by one vendor.] [I have attached a more detailed explanation below.]
On Aug 24, 2021, at 11:07 AM, Paul Schluter <pssch...@gmail.com> wrote: Re: FGF compensation - defining the negative of a Selectable Adjunct
Steven, Norman and John,I would like to propose another method for indicating the absence or negative of a Selectable Adjunct, using the ‘FGF compensation’ as the primary example.Before we proceed, let’s agree to append ‘C’ for ‘compensation’, thus changing ‘FGF’ to ‘FGFC’ or possibly to ‘FGC’.Since we are a reasonably certain that most, if not all, contemporary anesthesia machines either ‘compensate’ for FGF or, in the case of several Draeger models, use fresh gas decoupling, we should make the compensation or diversion of fresh gas flow the default expectation, and a Selectable Adjunct would not be shown. If we later encounter an anesthesia machine that does not compensate nor eliminate this error, then the negative ’noFGFC’ (or the shorter ’noFGC’) would be used as the Selectable Adjunct to indicate this.This provides several benefits:1. It eliminates the need to indicate FGFC (or shorter FGC) by the three vendors (Draeger, Getinge, GE) since they all provide FGF compensation or diversion.2. If we need ’noFGFC’ (or the shorter ’noFGC’) later on, we can deploy it without affecting backwards compatibility. It will be defined and supported in the PEGjs and XSLT but would not be displayed as an Selectable Adjuncts column until it is actually needed.3. Defining ‘noFGFC’ is more vendor-neutral, since the Draeger FGDV solves this issue directly without the need for a compensation calculation.Note that adopting the negative ’noFGFC’ (or the shorter ’noFGC’) precludes the use of the positive ‘FGFC’ (or the shorter ‘FGC’) - you can’t have both the negative and positive representations since the absence of the token would be ambiguous. If we adopt the negative ’noFGFC’ (or the shorter ’noFGC’) convention, then the presence of the ‘noFGFC' token indicates 'no compensation' whereas the absence of the ‘noFGFC' token indicates that 'the additional FGF is compensated for or diverted by the machine'. This allows the Selectable Adjuncts to be encoded as a sequence of Booleans, indicated by the presence (true()) or absence (false()) of the token, and is supported by “classic” IEEE 11073, HL7 V2 and V3, FHIR, and other encodings.Best regards,
PS. Steve, what have your found out from the ISO or other anesthesia standards regarding this topic?
On Aug 25, 2021, at 8:45 AM, Paul Schluter <pssch...@gmail.com> wrote: Re: Please join our RTM tcon today, Wednesday, August 25 - penultimate list of ISO 19223 and IEEE 11073 ventilator modes for Draeger, Getinge and GE
On Aug 25, 2021, at 8:45 AM, Paul Schluter <pssch...@gmail.com> wrote:
Greetings!Please join our RTM tcon this Wednesday, August 25th, starting at 11:00 AM CDT. (this meeting was cancelled due to IEEE EMBS Public Forum last week)
Charles,Please read ISO 19223 lung ventilator vocabulary and semantics or look on the ISO online browsing platform. It is all explained well there.Best regards,
Steven Dain
Please excuse the typos
On Wed, Sep 1, 2021 at 17:11, cha...@InteropeHealth.com<cha...@InteropeHealth.com> wrote:Paul,
Great Progress. One suggestion.
In working on the ventilation modes, the descriptive text by Draeger is quite precise and clarifies each one of the major modes.
A similar level of documentation is missing for Gettinge and GE, where the vendor mode acronyms are simply spelled out.
Such a stronger documentation would really strengthen the standards. Has this been discussed ?
Charles
De : ihepc...@googlegroups.com <ihepc...@googlegroups.com> De la part de Paul Schluter
Envoyé : mercredi 1 septembre 2021 16:27
À : ihe-p...@googlegroups.com; ihe-p...@googlegroups.com; PCD Technical Committee GG <ihepc...@googlegroups.com>; ihe-pcd-...@googlegroups.com; IHE-PCD-DPI-WG <ihe-p...@googlegroups.com>; device...@lists.hl7.org
Cc : Monroe Pattillo <monroe_...@bellsouth.net> (monroe_...@bellsouth.net) <monroe_...@bellsouth.net>; Mathieu Roullet <mrou...@enovacom.fr>; Stefan Schlichting <stefan.sc...@ornet.org>; Dalibor Pokrajac <Dalibor....@guardrfid.com>; STEVEN DAIN <sd...@rogers.com>; Jan Wittenber <jan.wi...@gmail.com>; Walsh John L.M.D. <john....@mgh.harvard.edu>; Abhyankar, Swapna <sabh...@regenstrief.org>; Paul Sherman <paulrsh...@gmail.com>; Garguilo, John J. <john.g...@nist.gov>; Crouzier, Nicolas <nicolas....@nist.gov>; Todd Cooper <toddco...@gmail.com>; John Rhoads <johnr...@johnrhoads.net>; Chris Courville <ccou...@epic.com>; michael.faughn <michael...@prometheuscomputing.com>; Tom.Ko...@bbraun.com; Norman Jones <nsjo...@btinternet.com>; Hassing, Kai <kai.h...@philips.com>; Klotz, Tobias <Tobias...@draeger.com>; Ken Fuchs <kenj....@gmail.com>; Gisch, Simon <Simon...@draeger.com>; Abell, Josh <Josh....@draeger.com>; Staudenmaier, Greg (Stod) <Greg.Sta...@va.gov>; Tracy Rausch <tr...@docboxinc.com>; Daidi Zhong <daidi...@ieee.org>; Michael J. Kirwan <mki...@dsheet.com>; Lisa Perry <l.p...@ieee.org>; Sabo, Perry (Carefusion, non-GE) <perry...@carefusion.com>; di...@moberg.com; Marks, Kenneth E <KEMa...@madisoncollege.edu>; Mike Krajnak <michael...@med.ge.com>; Neal Seidl <Neal....@med.ge.com>; Comunale, Carter <ccom...@bernoullihealth.com>; Szente, Andras <asz...@bernoullihealth.com>; Iglesias, Brian <bigl...@bernoullihealth.com>; Hadi Rahemi <rah...@circulationconcepts.com>; shimp...@marshfieldresearch.org; Andrew Norton <andrew...@tiscali.co.uk>; Tietz, Michael <Michae...@draeger.com>; Beier, Manfred <Manfre...@draeger.com>; Per Johansson <per.jo...@getinge.com>; Anders Häggström <anders.h...@getinge.com>; Mikael Petrini <mikael....@getinge.com>; Jeff Rinda <jeffre...@gmail.com>; Nichols, Steven (GE Healthcare) <Steven....@ge.com>; Ramachandran, Karthik R (GE Healthcare) <r.karthikr...@med.ge.com>; Brauer, Kimberly (GE Healthcare) <Kimberl...@ge.com>; snich...@gmail.com; Peeters, Wouter <w.pe...@philips.com>; Paul Schluter <pssch...@gmail.com>
Objet : [ihepcdtech:3172] The RTM tcon is cancelled for today, Wednesday, September 1st. NEWSFLASH: Provisional numeric codes for the 10101b Group6 ISO 19223 and IEEE 11073 ventilator modes have now been assigned!
--
You received this message because you are subscribed to the Google Groups "IHE PCD Technical Committee" group.
To unsubscribe from this group and stop receiving emails from it, send an email to ihepcdtech+...@googlegroups.com.
To view this discussion on the web visit https://groups.google.com/d/msgid/ihepcdtech/D6586ECC-371C-4F60-ABE3-23D25FF571B2%40gmail.com.
The columns for <AllTogetherGroup6.x4a.2021-08-27T14.isoVentMode-final.html/xls> are summarized below:
Column | Description/Notes |
REFID | IEEE 11073 REFID |
CommonTerm | IEEE Common Term |
Acronym | IEEE Acronym |
Description | IEEE Description/Definition |
AlertType | (not used) |
Disc | Discriminator Group |
PartCode | PART :: CODE10 |
CF_CODE10 | 32-bit context-free code = (PART * 216) + CODE10 |
gp | 10101b Group number (for internal bookkeeping only) |
UOM_MDC | (not used) |
UOM_UCUM | (not used) |
Enum_Values | Enumeration or bit-flag values |
Enum_Descriptions | Enumeration or bit-flag descriptions |
PEGjs | (not used – PEGjs for ISOMode will be documented in a separate section) |
The columns are described below for the three tables that comprise <ISO-Filter.2e.x3n.2021-08-27T10.isoVentModes-final.html/.xls>:
Table 1: This table lists vendor modes for Draeger, Getinge and GE, along with mappings to the ISO 19223 mode string and IEEE 11073 REFID. This table will be included in IEEE P11073-10101b and can be extended.
Selectable Adjuncts that are permanently associated with the VendorMode are pre-coordinated into the ISOMode string and the IEEE 11073 REFID. All Selectable Adjuncts, both ‘inherent’ to the mode and those that are conveyed as Boolean flags are listed in the SelAdjuncts column.
The { TC, CC, LC, NIV, SIGH, TRGC, noFGFC } columns provide a Boolean flag bit view of the Selectable Adjuncts for each VendorMode that are conveyed by ISOMode (indicated by ⦿) and SelAdjuncts (indicated by ●).
The Notes and NoteDetails columns list the notes that are associated with each mode. The Notes reference numbers are assigned independently of vendor. [For traceability, the original vendor note numbers are listed in the vNote column.]
Column | Description/Notes |
Vendor | Vendor [and optional Model suffix, e.g. GE-R860] |
VendorMode | Vendor mode |
VendorModeShortDescription | Vendor mode short description |
ISOMode | ISO 19223 string, including Selectable Adjuncts and consolidated Note numbers associated with the mode |
REFID | IEEE 11073 REFID (strictly derived from ISO 19223 string) |
Disc | Discriminator Group = ‘1’ |
PART | Partition = ‘7’ |
CODE10 | CODE10 |
CF_CODE10 | 32-bit context-free code = (PART * 216) + CODE10 |
SelAdjuncts | (0..*) Selectable Adjuncts conveyed as Boolean flags |
TC | Tube compensation |
CC | Compliance compensation |
LC | Leakage compensation |
NIV | Non Invasive Ventilation |
SIGH | SIGH mode |
TRGC | Trigger compensation |
noFGFC | The additional delivered flow (or tidal volume) due to fresh gas flow is not compensated (anesthesia) |
vNotes | Original Vendor Note reference numbers (will not be published in standard) |
Notes | (0..*) Note References (white-space separated consolidated Note numbers) |
NoteDetails | (0..*) Note Details (⬩ separated strings) |
vmg | Vendor Mode Group = ‘1’ for initial set of Draeger, Getinge and GE modes (essential) |
gp | 10101b Group = ‘6’ (for internal bookkeeping only) |
Table 2: This table lists the ISOMode and equivalent IEEE 11073 REFID and numeric codes and the consolidated Note information associate with the mode. This table will be included as normative content in IEEE P11073-10101b and can be extended.
Column | Description/Notes |
ISOMode | ISO 19223 string, including Selectable Adjuncts and consolidated Note numbers associated with the mode |
REFID | IEEE 11073 REFID |
Disc | Discriminator Group = ‘1’ |
PART | Partition = ‘7’ |
CODE10 | CODE10 |
CF_CODE10 | 32-bit context-free code = (PART * 216) + CODE10 |
Notes | (0..*) Note References (white-space separated consolidated Note numbers) |
NoteDetails | (0..*) Note Details (⬩ separated strings) |
Table 3: This table lists the consolidated Note number and details associated with the Note. This table will be included as normative content in IEEE P11073-10101b and can be extended.
Column | Description/Notes |
Note | Note Reference number (consolidated) |
NoteDetails | Note Detail |
Both HTML file use <br style="mso-data-placement:same-cell;”> to prevent Excel row-splitting when multiple <br> are present in the same table cell for the UOM_MDC, UOM_UCUM and new <PEGjs> column. Excel row-splitting is now permitted for Enum_Values and Enum_Descriptions due to the increased size of the Enum_Descriptions for several of the terms. I have also provided the Excel version (created by opening the HTML with Excel) for comparison.
Monday 9/20: HL7 Plenary (no IEEE or DEV meetings)Tuesday 9/21: IEEE 11073 PoCD WG (IEEE 11073 Nomenclature Update starts at 10:45 AM EDT, after the break)Wednesday 9/22: HL7 Devices (DEV) WGThursday 9/23: HL7 Devices (DEV) WG and HL7 Mobile HealthFriday 9/24: Devices on FHIR and Gemini SDC/SDPi Work Session
Group4 and Group5: Getinge ventilation and anesthesia termsGroup6: ISO 19223 and IEEE 11073 ventilation modes and related control variablesGroup7: IEEE P11073-10701 SDC terms and miscellaneous units-of-measure
“Subject (source) ,…, Condition/State/Action ,…, {optional qualifiers} ,…, {optional recommendations}
would have the following Systematic Name:“The blower temperature is excessively high; checking required”
ErrorEvent | Blower, temperature, excessively high, CheckingRequired | FunctionalDisturbance | Device, Ventilator
The bold font highlighting of the comma-separated Condition/State/Action field is triggered by the presence of certain key words and phrases plus a small number of additional rules. Comma-separated fields to the left of the Condition/State/Action field are detailed source identifiers and are compared with the fourth component “Device, Ventilator” as well as the proposed <source> identifiers expressed as REFIDs and numeric codes conveyed by the <src-evt> source-event identifier pairs.
On Mar 2, 2022, at 8:49 AM, Paul Schluter <pssch...@gmail.com> wrote:
Greetings!
The RTM tcon for today, Wednesday, March 2, 2022, is cancelled.
We had a very successful review of the new infusion pump terms two days ago (2/28) and have updated the tabular pump containment tree that contains the terms along with their co-constraints and containment context.The discussion notes for the pump terms, updated Excel file and “distilled” tabular containment tree will be sent out to the Pump WG later today. If no other issues are identified, the 15+ new terms will be assigned final “provisional” numeric codes after Monday noon next week.
On Mar 30, 2022, at 8:46 AM, Paul Schluter <pssch...@gmail.com> wrote:
Greetings!
The RTM tcon for today, Wednesday, March 30th, 2022, is cancelled.
I do not have any new material to discuss today, and other topics are being discussed with individual groups. Needless to say, work continues on the first ballot draft of IEEE P11073-10101b.
I am very happy to say that the REFIDs, Systematic Names, Descriptions have been largely finalized for the 10101b Group2 and Group8 infusion pump terms (except for the possible addition of several finer-grained enumerated values being considered by one vendor). Pre-provisional numeric codes have been assigned for the new Group8 pump terms and were submitted to Michael Faughn for review and acceptance as “provisional” terms in RTMMSv2.
I am currently consolidating all of the IEEE P11073-10101b terms by partition number and then by device, similar to how the tables in the base IEEE 11073-10101 nomenclature standard are arranged.
Individual units-of-measure can be listed with each term-row that requires them. Both UOM_MDC and UOM_UCUM will be listed in a side-by-side manner that defines an unambiguous mapping between the two.Named _UOM_GROUP identifiers can be also be specified that reference _UOM_GROUP identifiers in a compact summary list of all UOM_MDC and UOM_UCUM known to be used in any shipping device, product or system.I have attached an example of an Annex A table with the additional UOM_MDC and UOM_UCUM information and the current compact summary listing of all the units.
This information would be specified in a separate Annex. Although many of the enumerated values are expressed using a single token, others may prefer to use OID values with numeric codes. These can be captured together in a single normative (but extendable) mapping table dedicated to that purpose, with the following columns:_ENUM_GROUP membershipReferenceID and Part::Code (if an OID value has been defined)Token(s) and optionally ASN.1 flag-bit(s)Enum_Description (applies to all item in the same row)The _ENUM_GROUPS will use identifiers (e.g. pumps) that have already been defined. We can also use _MDC_my_refid as the enumeration group identifier for specific term-row having the ReferenceID MDC_my_refid. The MDC_my_refid will appear in the Annex A Description/Definition and will be trivial to extract from the Definition (since there is no room for this otherwise). [For export/upload to RTMMSv2, the _ENUM_GROUPS, Enum_Values and Enum_Descriptions in additional columns, as before.]
This lists all the 10101b Part=2 SCADA terms as either (a) subtables that are appended to existing tables in 10101-2019 or (b) as new tables. The table format is identical to previous IEEE 11073-10101* standards except for the addition of UOM_MDC and UOM_UCUM units-of-measure. [A compact summary list of UOM_MDC and UOM_UCUM units and _UOM_GROUPS will also be included as a new Annex, based on the table <CheckMDCUnits.4m.x2j.2022-04-09T12.show_X_aliases.html> that I sent to you two weeks ago. This table will not replace the customary main table listing the new units-of-measure for 10101b.] The additional UOM_MDC and UOM_UCUM information will fit in the IEEE-SA “large table” format using 8 pt Arial Narrow font, permitted by both the IEEE-SA and ISO.Annex A terms that have an enumerated value set specify the _ENUM_GROUPS with the text “EnumGroup = _MDC_HDIALY_MACH_MODE_OF_OPERATION” in the Description/Definition. The _ENUM_GROUP identifier must be globally unique across the entire standard. Historical _ENUM_GROUPS identifiers are used wherever possible; if one has not been defined, then the REFID for the parent term is used, prefaced by an underscore.
This is the table of contents from 10101-2019, and this was used to obtain the complete table titles that will be required for the 10101b amendment.
This is a listing of the { Reference ID, Disc, Part::Code and CF_CODE10 } that we adopted in 10101a and 10101-2019, derived from the same list of files used to produce Annex A.
_ENUM_GROUPS lists the name of the enumeration groupEnum_Values lists the enumerated tokens that would appear in a text-encoded message, such as HL7 V2.Enum_Descriptions list a brief description of the enumerated valueThis table will be incorporated in the draft standard with Excel subrows for the Enum_Values, Enum_Descriptions plus additional columns that are added in the future (e.g. ASN.1 mnemonic with a bit-position designator, OID representation using an IEEE 11073 REFID, etc.). The central idea is that this table represents the mapping between the various on-the-wire representations of the enumerated values, all sharing the common Enum_Description. We can (and will) define additional requirements and/or guidelines, e.g. all Enum_Values are scoped to the containing parent _ENUM_GROUP identifier, whereas we may strive for globally unique ASN.1 identifiers, and shall have globally unique OID-type identifiers within their defining vocabulary, etc.).
I think the question we really need to ask here is if the mappings between MDC and UCUM hold true for every instance where a metric identifies unit co-constraints. The answer is almost always “yes” but can be “no”. There are instances where the mappings are scoped just to the unit in question. Paul and I had this discussion some months ago and it actually led me to extend the model behind RTMMS to accommodate just such a situation. The mapping that really triggered this was the metric term MDC_SAT_O2_REGIONAL_DELTA_BASELINE_RATIO 2::19640 MMM(0) which was listed with the MDC units MDC_DIM_PERCENT and the UCUM units {Relative}%. Not every usage of MDC_DIM_PERCENT is a relative percent so the mapping could not be universal and had to be scoped just to that particular metric term. FWIW, I argued that a new MDC term should be created for relative percent. “Percent” and “relative percent” are not synonymous concepts.
We didn’t really address the question of whether or not this was a bidirectional mapping between UCUM and MDC units either. When I designed RTMMSv2, there was no direct mapping between metric terms and UCUM units. It was assumed (by me) that mappings between MDC units and UCUM units were universal. As such, all UCUM units associated with metric terms were derived from the association between the metric term and it’s MDC unit co-constraints. So, I wasn’t (and really still am not) concerned with a bidirectional mapping. Is there a use case for mapping from UCUM Units back to MDC Units??
Just running some rough scripts on RTMMSv2 here. This is fairly raw and you’ll notice that there are synonymous UCUM Units that get duplicated and also MDC units that are synonymous as well. You’ll have to parse through those.
Mapping from MDC Units to UCUM Units…These are probably all OK. I’m pretty sure all of the UCUM units here are synonyms. This does not take into account the few instances where there is a scoped mapping as described above. The mappings that are “per metric scoped” are in the next block of results.
MDC_DIM_PARTS_PER_10_TO_MINUS_9 --> 10*-9, 10^-9
MDC_DIM_PARTS_PER_10_TO_MINUS_12 --> 10*-12, 10^-12
MDC_DIM_X_L_PER_BREATH --> L/{breath}, L, l/{breath}, l
MDC_DIM_MILLI_L_PER_BREATH --> mL/{breath}, mL, ml/{breath}, ml
MDC_DIM_PER_MIN --> {count}/min, 1/min, /min
MDC_DIM_BEAT_PER_MIN --> {beat}/min, {beats}/min, 1/min, /min
MDC_DIM_PULS_PER_MIN --> {pulse}/min, {pulses}/min, 1/min, /min
MDC_DIM_RESP_PER_MIN --> {resp}/min, {breath}/min, {breaths}/min, 1/min, /min
MDC_DIM_X_EQUIV_PER_CM_CUBE --> mol/cm3, eq/cm3
MDC_DIM_CM_H2O_PER_L_PER_SEC --> cm[H2O].s/L, cm[H2O]/(L/s), cm[H2O].s/l, cm[H2O]/(l/s)
MDC_DIM_DYNE_SEC_PER_M_SQ_PER_CM_5 --> dyn.s/cm5/m2, dyn.s.cm-5.m-2
MDC_DIM_X_L_PER_BEAT --> L/{beat}, L, l/{beat}, l
MDC_DIM_BEAT_PER_MIN_PER_X_L --> {beat}/min/L, {beats}/min/L, 1/min/L, /min/L, {beat}/min/l, {beats}/min/l, 1/min/l, /min/l
MDC_DIM_X_LUMEN_PER_M_SQ --> lm/m2, lx
MDC_DIM_MILLI_L_PER_BEAT --> mL/{beat}, mL, ml/{beat}, ml
MDC_DIM_X_ROTATIONS_PER_MIN --> 360.deg/min, 2.[pi].rad/min
MDC_DIM_PER_X_OHM --> 1/Ohm, /Ohm, S
MDC_DIM_DYNE_SEC_M_SQ_PER_CM_5 --> dyn.s.m2/cm5, dyn.s.cm-5.m2
MDC_DIM_X_JOULES_PER_BREATH --> J/{breath}, J
MDC_DIM_MMHG_SEC_PER_ML --> [PRU], mm[Hg].s/mL, mm[Hg].s/ml
MDC_DIM_MMHG_MIN_PER_L --> [wood'U], mm[Hg].min/L, mm[Hg].min/l
MDC_DIM_SQUARE_BREATHS_PER_MIN_PER_L --> {breaths-squared}/min/L, /min/L, {breaths-squared}/min/l, /min/l
MDC_DIM_O2_SAT_PERCENT_SEC --> {sat}.%.s, %.s
MDC_DIM_X_M_PER_VOLT --> mm/mV, m/V
MDC_DIM_X_BAR_SEC_PER_L --> bar.s/L, bar/(L/s), bar.s/l, bar/(l/s)
MDC_DIM_X_L_PER_BEAT_PER_M2 --> L/{beat}/m2, L/m2, l/{beat}/m2, l/m2
MDC_DIM_NANO_VOLT_SEC --> uV.ms, nV.s
MDC_DIM_PER_KILO_OHM --> 1/kOhm, /kOhm, mS
MDC_DIM_MILLI_BAR_SEC_PER_L --> mbar.s/L, mbar/(L/s), mbar.s/l, mbar/(l/s)
MDC_DIM_MILLI_L_PER_BEAT_PER_M2 --> mL/{beat}/m2, mL/m2, ml/{beat}/m2, ml/m2
MDC_DIM_CM_H2O_SEC_PER_BREATH --> cm[H2O].s/{breath}, cm[H2O].s
MDC_DIM_MILLI_BAR_SEC_PER_BREATH --> mbar.s/{breath}, mbar.s
MDC_DIM_HECTO_PASCAL_SEC_PER_L --> hPa.s/L, hPa/(L/s), hPa.s/l, hPa/(l/s)
MDC_DIM_BREATHS_PER_MIN_PER_MILLI_L --> {breaths}/min/mL, {breath}/min/mL, /min/mL, {breaths}/min/ml, {breath}/min/ml, /min/ml
MDC_DIM_BREATHS_PER_MIN_PER_X_L --> {breaths}/min/L, {breath}/min/L, /min/L, {breaths}/min/l, {breath}/min/l, /min/l
MDC_DIM_X_BAR_SEC_PER_BREATH --> bar.s/{breath}, bar.s
MDC_DIM_MICRO_ABSORBANCE --> {microabsorbance}, {unitless}
MDC_DIM_STEP --> {step}, {steps}, {unitless}
MDC_DIM_STEP_PER_MIN --> {step}/min, {steps}/min, 1/min, /min, {unitless}/min
MDC_DIM_TICK --> {tick}, {ticks}
MDC_DIM_X_EVT_PER_HR --> {event}/h, {events}/h, 1/h, /h
MDC_DIM_X_WATT_HR --> W.h, Wh
MDC_DIM_PARTS_PER_10_TO_15 --> 10*-15, 10^-15
MDC_DIM_X_LUMEN --> lm, cd.sr
MDC_DIM_PARTS_PER_10_TO_MINUS_15 --> 10*-15, 10^-15
MDC_DIM_PARTS_PER_10_TO_9 --> 10*-9, 10^-9, [ppb]
MDC_DIM_PARTS_PER_10_TO_12 --> 10*-12, 10^-12, [pptr]
MDC_DIM_PARTS_PER_10_TO_18 --> 10*-18, 10^-18
MDC_DIM_X_ROTATIONS --> 360.deg, 2.[pi].rad
MDC_DIM_INHG --> [in_i'Hg], [in_i’Hg]
MDC_DIM_KILO_WATT_HR --> kW.h, kWh
MDC_DIM_O2_SAT_PERCENT_MIN --> {sat}.%.min, %.min
MDC_DIM_KILO_ROTATIONS --> 360000.deg, 2000.[pi].rad
MDC_DIM_DECI_HZ --> dHz, Hz/10
MDC_DIM_VIAL --> {Vial}, {Vials}
All metric terms with scoped mappings. Ignore the number in square brackets. The metric mapped to relative percent is highlighted.
Term[7873] Provisional (10101b) MDC_VOL_FLUID_CONTAINER_START 2::26922 1(0)
Universal mapping: MDC_DIM_MILLI_L --> mL, ml
Scoped mapping: MDC_DIM_MILLI_L -- {liquid}mL
Term[7890] Provisional (10101b) MDC_VOL_REMAIN_CONTAINER 2::26923 1(0)
Universal mapping: MDC_DIM_MILLI_L --> mL, ml
Scoped mapping: MDC_DIM_MILLI_L -- {liquid}mL
Term[7893] Provisional (10101b) MDC_VOL_FLUID_DELIV_TOTAL 2::26921 1(0)
Universal mapping: MDC_DIM_MILLI_L --> mL, ml
Scoped mapping: MDC_DIM_MILLI_L -- {liquid}mL
Term[7945] Provisional (10101b) MDC_VOL_FLUID_DELIV_SEGMENT 2::26920 1(0)
Universal mapping: MDC_DIM_MILLI_L --> mL, ml
Scoped mapping: MDC_DIM_MILLI_L -- {liquid}mL
Term[40723] Provisional (10101b) MDC_SAT_O2_REGIONAL_DELTA_BASELINE_RATIO 2::19640 MMM(0)
Universal mapping: MDC_DIM_PERCENT --> %
Scoped mapping: MDC_DIM_PERCENT -- {Relative}%
Term[40946] Provisional (10101b) MDC_VOL_FLUID_DELIV_PUMP_TOTAL 2::27008 MMM(0)
Universal mapping: MDC_DIM_MILLI_L --> mL, ml
Scoped mapping: MDC_DIM_MILLI_L -- {liquid}mL
Term[40950] Provisional (10101b) MDC_VOL_FLUID_DELIV_METHOD_TOTAL 2::27012 MMM(0)
Universal mapping: MDC_DIM_MILLI_L --> mL, ml
Scoped mapping: MDC_DIM_MILLI_L -- {liquid}mL
Term[40954] Provisional (10101b) MDC_VOL_FLUID_DELIV_FOR_PERIOD 2::27016 MMM(0)
Universal mapping: MDC_DIM_MILLI_L --> mL, ml
Scoped mapping: MDC_DIM_MILLI_L -- {liquid}mL
Term[40968] Provisional (10101b) MDC_VOL_FLUID_DELIV_FOR_DOSE 2::27032 MMM(0)
Universal mapping: MDC_DIM_MILLI_L --> mL, ml
Scoped mapping: MDC_DIM_MILLI_L -- {liquid}mL
Mappings from UCUM to MDC. Like I said, I’m not sure we really care about these. The one in green is definitely not 1:1. Also, do these all look correct? E.g., the one’s in red?
{unknown} --> MDC_DIM_NOS, MDC_DIM_MULT, MDC_DIM_DIV, MDCX_DIM_PSIG
{unitless} --> MDC_DIM_DIMLESS, MDC_DIM_MICRO_ABSORBANCE, MDC_DIM_STEP
1 --> MDC_DIM_DIMLESS, MDC_DIM_RBC, MDC_DIM_BEAT, MDC_DIM_BREATH, MDC_DIM_CELL, MDC_DIM_COUGH, MDC_DIM_SIGH, MDC_DIM_BOOLEAN, MDC_DIM_INR, MDC_DIM_MICRO_ABSORBANCE, MDC_DIM_STEP, MDC_DIM_TICK
dB --> MDC_DIM_DECIBEL, MDC_DIM_DECI_BEL
dB[V] --> MDC_DIM_DECIBEL_X_VOLT, MDC_DIM_DECIBEL_X_V
dB[10.nV] --> MDC_DIM_DECIBEL_10_NANO_VOLT, MDC_DIM_DECIBEL_DECA_NV
[ppth] --> MDC_DIM_PARTS_PER_10_TO_MINUS_3, MDC_DIM_PARTS_PER_THOUSAND, MDC_TST_UNIT, MDC_DIM_PARTS_PER_10_TO_3
[ppm] --> MDC_DIM_PARTS_PER_10_TO_MINUS_6, MDC_DIM_PARTS_PER_MILLION, MDC_DIM_PARTS_PER_10_TO_6
10*-9 --> MDC_DIM_PARTS_PER_10_TO_MINUS_9, MDC_DIM_PARTS_PER_10_TO_9
10^-9 --> MDC_DIM_PARTS_PER_10_TO_MINUS_9, MDC_DIM_PARTS_PER_10_TO_9
10*-12 --> MDC_DIM_PARTS_PER_10_TO_MINUS_12, MDC_DIM_PARTS_PER_10_TO_12
10^-12 --> MDC_DIM_PARTS_PER_10_TO_MINUS_12, MDC_DIM_PARTS_PER_10_TO_12
10*-15 --> MDC_DIM_PARTS_PER_10_TO_15, MDC_DIM_PARTS_PER_10_TO_MINUS_15
10^-15 --> MDC_DIM_PARTS_PER_10_TO_15, MDC_DIM_PARTS_PER_10_TO_MINUS_15
L/m2 --> MDC_DIM_X_L_PER_M_SQ, MDC_DIM_X_L_PER_BEAT_PER_M2
mL/m2 --> MDC_DIM_MILLI_L_PER_M_SQ, MDC_DIM_MILLI_L_PER_BEAT_PER_M2
L --> MDC_DIM_X_L, MDC_DIM_X_L_PER_BREATH, MDC_DIM_X_L_PER_BEAT, MDC_DIM_L
mL --> MDC_DIM_MILLI_L, MDC_DIM_MILLI_L_PER_BREATH, MDC_DIM_MILLI_L_PER_BEAT, DIM_MILLI_L
/min --> MDC_DIM_PER_MIN, MDC_DIM_BEAT_PER_MIN, MDC_DIM_PULS_PER_MIN, MDC_DIM_RESP_PER_MIN, MDC_DIM_STEP_PER_MIN
1/min --> MDC_DIM_PER_MIN, MDC_DIM_BEAT_PER_MIN, MDC_DIM_PULS_PER_MIN, MDC_DIM_RESP_PER_MIN, MDC_DIM_STEP_PER_MIN
/h --> MDC_DIM_PER_HR, MDC_DIM_X_EVT_PER_HR
/min/L --> MDC_DIM_BEAT_PER_MIN_PER_X_L, MDC_DIM_SQUARE_BREATHS_PER_MIN_PER_L, MDC_DIM_BREATHS_PER_MIN_PER_X_L
mL/min/m2 --> MDC_DIM_MILLI_L_PER_MIN_PER_M_SQ, MDC_DIM_MILLI_L_PER_MIN_PER_M2
J --> MDC_DIM_X_JOULES, MDC_DIM_X_JOULES_PER_BREATH
dyn.s/cm5 --> MDC_DIM_DYNE_SEC_PER_CM_5, MDC_DIM_DYNE_SEC_PER_CM5, MDC_DIM_X_DYNE_SEC_PER_CM5
Pa/L --> MDC_DIM_PA_PER_X_L, MDC_DIM_X_PASCAL_PER_L
T --> MDC_DIM_X_TESLA, MDC_Test_UNIT
mol/cm3 --> MDC_DIM_X_MOLE_PER_CM_CUBE, MDC_DIM_X_EQUIV_PER_CM_CUBE
km/h --> MDC_DIM_KILO_M_PER_SEC, MDC_DIM_KILO_M_PER_HR
{unitless} --> MDC_DIM_TOD, MDC_DIM_DATE
l/m2 --> MDC_DIM_X_L_PER_M_SQ, MDC_DIM_X_L_PER_BEAT_PER_M2
ml/m2 --> MDC_DIM_MILLI_L_PER_M_SQ, MDC_DIM_MILLI_L_PER_BEAT_PER_M2
l --> MDC_DIM_X_L, MDC_DIM_X_L_PER_BREATH, MDC_DIM_X_L_PER_BEAT
ml --> MDC_DIM_MILLI_L, MDC_DIM_MILLI_L_PER_BREATH, MDC_DIM_MILLI_L_PER_BEAT
/min/l --> MDC_DIM_BEAT_PER_MIN_PER_X_L, MDC_DIM_SQUARE_BREATHS_PER_MIN_PER_L, MDC_DIM_BREATHS_PER_MIN_PER_X_L
Pa/l --> MDC_DIM_PA_PER_X_L, MDC_DIM_X_PASCAL_PER_L
-Michael
From:
Daniel Rubery <Daniel...@freseniusmedicalcare.com>
Date: Thursday, May 5, 2022 at 5:39 AM
To: ihe-pcd-...@googlegroups.com <ihe-pcd-...@googlegroups.com>, Malcolm Clarke <mclar...@gmail.com>, Paul Schluter <pssch...@gmail.com>, ihe-p...@googlegroups.com <ihe-p...@googlegroups.com>, ihe-p...@googlegroups.com <ihe-p...@googlegroups.com>,
IHE DEV Tooling and Testing <ihe-dev...@googlegroups.com>, PCD Technical Committee GG <ihepc...@googlegroups.com>, device...@lists.hl7.org <device...@lists.hl7.org>, Steven Dain <sd...@rogers.com>
Cc: Monroe Pattillo <monroe_...@bellsouth.net>, Mathieu Roullet <mrou...@enovacom.fr>, Stefan Schlichting <stefan.sc...@ornet.org>, Dalibor Pokrajac <Dalibor....@guardrfid.com>, Walsh John L.M.D. <john....@mgh.harvard.edu>, Paul Sherman
<paulrsh...@gmail.com>, Garguilo, John J. (Fed) <john.g...@nist.gov>, Crouzier, Nicolas (Assoc) <nicolas....@nist.gov>, Todd Cooper <toddco...@gmail.com>, John Rhoads <johnr...@johnrhoads.net>, Chris Courville <ccou...@epic.com>, Faughn,
Michael R. (Fed) <michael...@nist.gov>, Tom.Ko...@bbraun.com <Tom.Ko...@bbraun.com>, Norman Jones <nsjo...@btinternet.com>, Hassing, Kai <kai.h...@philips.com>, Klotz, Tobias <Tobias...@draeger.com>, Ken Fuchs <kenj....@gmail.com>, Gisch,
Simon <Simon...@draeger.com>, Abell, Josh <Josh....@draeger.com>, Staudenmaier, Greg (Stod) <Greg.Sta...@va.gov>, Tracy Rausch <tr...@docboxinc.com>, Daidi Zhong <daidi...@ieee.org>, Michael J. Kirwan <mki...@dsheet.com>, Lisa Perry <l.p...@ieee.org>,
Sabo, Perry (Carefusion, non-GE) <perry...@carefusion.com>, di...@moberg.com <di...@moberg.com>, Marks, Kenneth E <KEMa...@madisoncollege.edu>, Mike Krajnak <michael...@med.ge.com>, Neal Seidl <Neal....@med.ge.com>, Comunale, Carter <ccom...@bernoullihealth.com>,
Szente, Andras <asz...@bernoullihealth.com>, Iglesias, Brian <bigl...@bernoullihealth.com>, Hadi Rahemi <rah...@circulationconcepts.com>, shimp...@marshfieldresearch.org <shimp...@marshfieldresearch.org>, Andrew Norton <andrew...@tiscali.co.uk>,
Tietz, Michael <Michae...@draeger.com>, Beier, Manfred <Manfre...@draeger.com>, Per Johansson <per.jo...@getinge.com>, Anders Häggström <anders.h...@getinge.com>, Jeff Rinda <jeffre...@gmail.com>, Nichols, Steven (GE Healthcare) <Steven....@ge.com>,
Ramachandran, Karthik R (GE Healthcare) <r.karthikr...@med.ge.com>, Brauer, Kimberly (GE Healthcare) <Kimberl...@ge.com>, Peeters, Wouter <w.pe...@philips.com>, Philipsen, Jan-Alrik <Jan-Alrik...@draeger.com>, Paul Schluter <psch...@center4mi.org>
Subject: Re: [EXTERNAL EMAIL] Re: [ihe-pcd-rtm-vent:1459] Please join our RTM tcon tomorrow, Wednesday, May 4th, at 11:00 AM CDT - additional information to be included in IEEE P11073-10101b (follow-up)
Malcom made the following suggestion a couple of emails back:
I am assuming there is a 1:1 mapping, and therefore information on mapping could be included, but it is not clear why a list of UCUM measures should be given for every SCADA term, a mapping table would "save much space" as many terms use common nomenclature.
During the meeting yesterday, it was said that there is not always a 1:1 mapping. Can some one provide an example of when there is not a 1:1 mapping between an IEEE UOM and a UCUM UOM?
Thanks,
Dan
From: Steven Dain <sd...@rogers.com>
Sent: Wednesday, May 4, 2022 11:55 AM
To: ihe-pcd-...@googlegroups.com <ihe-pcd-...@googlegroups.com>; Malcolm Clarke <mclar...@gmail.com>; Paul Schluter <pssch...@gmail.com>; ihe-p...@googlegroups.com <ihe-p...@googlegroups.com>; ihe-p...@googlegroups.com <ihe-p...@googlegroups.com>;
IHE DEV Tooling and Testing <ihe-dev...@googlegroups.com>; PCD Technical Committee GG <ihepc...@googlegroups.com>; device...@lists.hl7.org <device...@lists.hl7.org>
Cc: Monroe Pattillo <monroe_...@bellsouth.net>; Mathieu Roullet <mrou...@enovacom.fr>; Stefan Schlichting <stefan.sc...@ornet.org>; Dalibor Pokrajac <Dalibor....@guardrfid.com>; Walsh John L.M.D. <john....@mgh.harvard.edu>; Paul Sherman
<paulrsh...@gmail.com>; Garguilo, John J. <john.g...@nist.gov>; Crouzier, Nicolas <nicolas....@nist.gov>; Todd Cooper <toddco...@gmail.com>; John Rhoads <johnr...@johnrhoads.net>; Chris Courville <ccou...@epic.com>; Faughn, Michael R.
(Fed) <michael...@nist.gov>; Tom.Ko...@bbraun.com <Tom.Ko...@bbraun.com>; Norman Jones <nsjo...@btinternet.com>; Hassing, Kai <kai.h...@philips.com>; Klotz, Tobias <Tobias...@draeger.com>; Ken Fuchs <kenj....@gmail.com>; Gisch, Simon
<Simon...@draeger.com>; Abell, Josh <Josh....@draeger.com>; Staudenmaier, Greg (Stod) <Greg.Sta...@va.gov>; Tracy Rausch <tr...@docboxinc.com>; Daidi Zhong <daidi...@ieee.org>; Michael J. Kirwan <mki...@dsheet.com>; Lisa Perry <l.p...@ieee.org>;
Sabo, Perry (Carefusion, non-GE) <perry...@carefusion.com>; di...@moberg.com <di...@moberg.com>; Marks, Kenneth E <KEMa...@madisoncollege.edu>; Mike Krajnak <michael...@med.ge.com>; Neal Seidl <Neal....@med.ge.com>; Comunale, Carter <ccom...@bernoullihealth.com>;
Szente, Andras <asz...@bernoullihealth.com>; Iglesias, Brian <bigl...@bernoullihealth.com>; Hadi Rahemi <rah...@circulationconcepts.com>; shimp...@marshfieldresearch.org <shimp...@marshfieldresearch.org>; Andrew Norton <andrew...@tiscali.co.uk>;
Tietz, Michael <Michae...@draeger.com>; Beier, Manfred <Manfre...@draeger.com>; Per Johansson <per.jo...@getinge.com>; Anders Häggström <anders.h...@getinge.com>; Jeff Rinda <jeffre...@gmail.com>; Nichols, Steven (GE Healthcare) <Steven....@ge.com>;
Ramachandran, Karthik R (GE Healthcare) <r.karthikr...@med.ge.com>; Brauer, Kimberly (GE Healthcare) <Kimberl...@ge.com>; Peeters, Wouter <w.pe...@philips.com>; Philipsen, Jan-Alrik <Jan-Alrik...@draeger.com>; Daniel Rubery <Daniel...@freseniusmedicalcare.com>;
Paul Schluter <psch...@center4mi.org>
Subject: [EXTERNAL EMAIL] Re: [ihe-pcd-rtm-vent:1459] Please join our RTM tcon tomorrow, Wednesday, May 4th, at 11:00 AM CDT - additional information to be included in IEEE P11073-10101b (follow-up)
NOT an FMCNA email - External emailHi Paul I think Malcolm suggestion in his last paragraph is reasonable and will save lots of space reducing the cost of standard
Best regards,
Steven
Please excuse the typos
On Wed, May 4, 2022 at 11:39, Malcolm Clarke
<mclar...@gmail.com> wrote:
Dear Paul
My take.
In PHD we have always included the list of units that may (should) be used with a metric (normally SCADA), assuming this is informative, ie it is not exhaustive nor mandatory, and could be extended in the future. Likewise we have listed the enumerations that may/should be used in a metric.
However the case for including UCUM in an IEEE 11073 document is not so clear, as we do not encourage its use in place of MDC units in any IEEE based protocol, and indeed for some applications we require and use only 11073 numeric codes.
I am assuming there is a 1:1 mapping, and therefore information on mapping could be included, but it is not clear why a list of UCUM measures should be given for every SCADA term, a mapping table would "save much space" as many terms use common nomenclature.
Regards
Malcolm
--
You received this message because you are subscribed to the Google Groups "IHE PCD Rosetta Term Mapping - Ventilator Task Group" group.
To unsubscribe from this group and stop receiving emails from it, send an email to ihe-pcd-rtm-ve...@googlegroups.com.
To view this discussion on the web visit https://groups.google.com/d/msgid/ihe-pcd-rtm-vent/BE8B0D97-88CE-4A7A-B648-99B96C629D31%40gmail.com.
--
You received this message because you are subscribed to the Google Groups "IHE PCD Rosetta Term Mapping - Ventilator Task Group" group.
To unsubscribe from this group and stop receiving emails from it, send an email to ihe-pcd-rtm-ve...@googlegroups.com.
To view this discussion on the web visit https://groups.google.com/d/msgid/ihe-pcd-rtm-vent/BE8B0D97-88CE-4A7A-B648-99B96C629D31%40gmail.com.
--
You received this message because you are subscribed to the Google Groups "IHE PCD Rosetta Term Mapping - Ventilator Task Group" group.
To unsubscribe from this group and stop receiving emails from it, send an email to ihe-pcd-rtm-ve...@googlegroups.com.
To view this discussion on the web visit https://groups.google.com/d/msgid/ihe-pcd-rtm-vent/BE8B0D97-88CE-4A7A-B648-99B96C629D31%40gmail.com.
--
You received this message because you are subscribed to the Google Groups "IHE PCD Rosetta Term Mapping - Ventilator Task Group" group.
To unsubscribe from this group and stop receiving emails from it, send an email to ihe-pcd-rtm-ve...@googlegroups.com.
To view this discussion on the web visit https://groups.google.com/d/msgid/ihe-pcd-rtm-vent/BE8B0D97-88CE-4A7A-B648-99B96C629D31%40gmail.com.--
You received this message because you are subscribed to the Google Groups "IHE PCD Rosetta Term Mapping - Ventilator Task Group" group.
To unsubscribe from this group and stop receiving emails from it, send an email to ihe-pcd-rtm-ve...@googlegroups.com.
To view this discussion on the web visit https://groups.google.com/d/msgid/ihe-pcd-rtm-vent/46440b79-02f7-8b6c-0982-3894f4698e75%40gmail.com.
CF_UCODE10 |
UCODE10 |
UOM_MDC |
Unit_of_Measure |
UOM_UCUM |
_UOM_GROUPS |
262688 |
544 |
MDC_DIM_PERCENT |
10-2 (percent) |
% |
_UOM_CONC_GAS |
268384 |
6240 |
MDC_DIM_VOL_PERCENT |
volume percent |
%{vol} % |
_UOM_CONC_GAS |
263392 |
1248 |
MDC_DIM_PCT_PCV |
percent of packed cell volume |
%{pcv} % |
|
274688 |
12544 |
MDC_DIM_RELATIVE_PERCENT |
relative percent |
%{relative} {relative}.% % |
|
On May 5, 2022, at 6:37 AM, Daniel Rubery <Daniel...@freseniusmedicalcare.com> wrote:
Malcom made the following suggestion a couple of emails back:I am assuming there is a 1:1 mapping, and therefore information on mapping could be included, but it is not clear why a list of UCUM measures should be given for every SCADA term, a mapping table would "save much space" as many terms use common nomenclature.During the meeting yesterday, it was said that there is not always a 1:1 mapping. Can some one provide an example of when there is not a 1:1 mapping between an IEEE UOM and a UCUM UOM?Thanks,Dan
From: Steven Dain <sd...@rogers.com>
Sent: Wednesday, May 4, 2022 11:55 AM
To: ihe-pcd-...@googlegroups.com <ihe-pcd-...@googlegroups.com>; Malcolm Clarke <mclar...@gmail.com>; Paul Schluter <pssch...@gmail.com>; ihe-p...@googlegroups.com <ihe-p...@googlegroups.com>; ihe-p...@googlegroups.com<ihe-p...@googlegroups.com>; IHE DEV Tooling and Testing <ihe-dev...@googlegroups.com>; PCD Technical Committee GG <ihepc...@googlegroups.com>; device...@lists.hl7.org <device...@lists.hl7.org>
Cc: Monroe Pattillo <monroe_...@bellsouth.net>; Mathieu Roullet <mrou...@enovacom.fr>; Stefan Schlichting <stefan.sc...@ornet.org>; Dalibor Pokrajac <Dalibor....@guardrfid.com>; Walsh John L.M.D. <john....@mgh.harvard.edu>; Paul Sherman <paulrsh...@gmail.com>; Garguilo, John J. <john.g...@nist.gov>; Crouzier, Nicolas <nicolas....@nist.gov>; Todd Cooper <toddco...@gmail.com>; John Rhoads <johnr...@johnrhoads.net>; Chris Courville <ccou...@epic.com>; Faughn, Michael R. (Fed) <michael...@nist.gov>; Tom.Ko...@bbraun.com <Tom.Ko...@bbraun.com>; Norman Jones <nsjo...@btinternet.com>; Hassing, Kai <kai.h...@philips.com>; Klotz, Tobias <Tobias...@draeger.com>; Ken Fuchs <kenj....@gmail.com>; Gisch, Simon <Simon...@draeger.com>; Abell, Josh <Josh....@draeger.com>; Staudenmaier, Greg (Stod) <Greg.Sta...@va.gov>; Tracy Rausch <tr...@docboxinc.com>; Daidi Zhong <daidi...@ieee.org>; Michael J. Kirwan <mki...@dsheet.com>; Lisa Perry <l.p...@ieee.org>; Sabo, Perry (Carefusion, non-GE) <perry...@carefusion.com>; di...@moberg.com <di...@moberg.com>; Marks, Kenneth E <KEMa...@madisoncollege.edu>; Mike Krajnak <michael...@med.ge.com>; Neal Seidl <Neal....@med.ge.com>; Comunale, Carter <ccom...@bernoullihealth.com>; Szente, Andras <asz...@bernoullihealth.com>; Iglesias, Brian <bigl...@bernoullihealth.com>; Hadi Rahemi <rah...@circulationconcepts.com>; shimp...@marshfieldresearch.org <shimp...@marshfieldresearch.org>; Andrew Norton <andrew...@tiscali.co.uk>; Tietz, Michael <Michae...@draeger.com>; Beier, Manfred <Manfre...@draeger.com>; Per Johansson <per.jo...@getinge.com>; Anders Häggström <anders.h...@getinge.com>; Jeff Rinda <jeffre...@gmail.com>; Nichols, Steven (GE Healthcare) <Steven....@ge.com>; Ramachandran, Karthik R (GE Healthcare) <r.karthikr...@med.ge.com>; Brauer, Kimberly (GE Healthcare) <Kimberl...@ge.com>; Peeters, Wouter <w.pe...@philips.com>; Philipsen, Jan-Alrik <Jan-Alrik...@draeger.com>; Daniel Rubery <Daniel...@freseniusmedicalcare.com>; Paul Schluter <psch...@center4mi.org>
Subject: [EXTERNAL EMAIL] Re: [ihe-pcd-rtm-vent:1459] Please join our RTM tcon tomorrow, Wednesday, May 4th, at 11:00 AM CDT - additional information to be included in IEEE P11073-10101b (follow-up)
NOT an FMCNA email - External emailHi Paul I think Malcolm suggestion in his last paragraph is reasonable and will save lots of space reducing the cost of standardBest regards,
Steven
Please excuse the typos
On Wed, May 4, 2022 at 11:39, Malcolm Clarke<mclar...@gmail.com> wrote:
Dear PaulMy take.In PHD we have always included the list of units that may (should) be used with a metric (normally SCADA), assuming this is informative, ie it is not exhaustive nor mandatory, and could be extended in the future. Likewise we have listed the enumerations that may/should be used in a metric.However the case for including UCUM in an IEEE 11073 document is not so clear, as we do not encourage its use in place of MDC units in any IEEE based protocol, and indeed for some applications we require and use only 11073 numeric codes.I am assuming there is a 1:1 mapping, and therefore information on mapping could be included, but it is not clear why a list of UCUM measures should be given for every SCADA term, a mapping table would "save much space" as many terms use common nomenclature.RegardsMalcolm
--
You received this message because you are subscribed to the Google Groups "IHE PCD Rosetta Term Mapping - Ventilator Task Group" group.
To unsubscribe from this group and stop receiving emails from it, send an email to ihe-pcd-rtm-ve...@googlegroups.com.
To view this discussion on the web visit https://groups.google.com/d/msgid/ihe-pcd-rtm-vent/BE8B0D97-88CE-4A7A-B648-99B96C629D31%40gmail.com.
UOM_MDC: MDC_DIM_BEAT_PER_MIN ‘beats per minute’ UOM_UCUM: {beat}/min {beats}/min 1/min /min
Enum_Values lists the enumerated tokens that would appear in a text-encoded message, such as HL7 V2.Enum_Descriptions list a brief description of the enumerated value
Table A.x.y.z – Dialysis Mode of Operation [MDC_HDIALY_MACH_MODE_OF_OPERATION]
Enum_Values |
Enum_Descriptions |
PRETX |
Pre-Treatment – Preparing for dialysis but no blood in extracorporeal circuit |
TX |
Treatment – Blood in the extracorporeal circuit |
POSTTX |
Post-Treatment – After dialysis but no longer any blood in extracorporeal circuit |
DIS |
Disinfection or rinse of dialysis fluid path. |
IDL |
Idle |
SVC |
Service mode |
Dear Paul
There are two aspects to this discussion:
1. The physics, convention or definition of a term will also define the single unit that is accepted (or equivalents). It is to be expected that every coding scheme will give that unit with that term. It is hard to conceive otherwise. Why does one scheme define the unit for a term differently? They must be the same.
2. In implementing a generic gateway it is conceivable only to provide rules or a simple mapping table of MDC_unit to UCUM, it is not realistic to include the entire RTMMS with the UCUM code for every term, nor hunt RTMMS for exceptions and code for these. We must attain 1:1 mapping for sanity and for respect (why are these IEEE people always different?).
Regards
Malcolm
On May 9, 2022, at 4:24 PM, Paul Schluter <pssch...@gmail.com> wrote:
Greetings!
The RTM tcon normally scheduled for this Wednesday, May 11th, is cancelled due to upcoming IEEE 11073 PoCD and HL7 DEV SIG Working Group meetings to be held (virtually) this week.I plan to provide a brief update regarding the status of IEEE P11073-10101b on Tuesday, May 10th, from 11:30 AM to 12:00 Noon US EDT (based on the current draft agenda) and my primary topic will be the addition of the units-of-measure and enumerated values to the draft IEEE P11073-10101b standard, plus other topics. The meetings continue to Friday this week.The draft agenda for the IEEE 11073 PoCD and HL7 DEV SIG Working Group meetings is available at https://confluence.hl7.org/display/HCD/Agenda%3A+2022-May+9+thru+13+-+IEEE+11073+and+HL7+DEVICES+Joint+Working+Group+Meetings%2C+Virtual .
Although the majority of NMT measurements are reported as relative values (e.g. TOF%) rather than as absolute measurements, there are devices and systems that report absolute measurements as well. The NMT response can be measured using a variety of methods listed below, and my question to any of you (that design or market or interface to) an NMT system are what are the additional units-of-measure that could be reported by commercially available systems that should be listed along with the units-of-measure highlighted with yellow below?The set of units will be collectively represented by the _UOM_NMT_AMPL unit-of-measure group referenced the twelve NMT “absolute” numeric measurements, including evoked response. Representing the set of units as a _UOM_GROUP will facilitate adding units later on and/or as new sensing technologies for the NMT evoked response are developed in the future.
Technique
Acronym
UOM_MDC
UOM_UCUM
Notes
Generic or undisclosed method
MDC_DIM_NOS
{unknown}
Electromyography
EMG
MDC_DIM_MILLI_VOLT
mV
Mechanomyography (force, tension)
MMG
MDC_DIM_G_FORCE
gf
Note: gram-force is now an MDC unit!
Aceceleromyography (linear, angular or 3D acceleration of thumb)
AMG, 3D AMG
MDC_DIM_CENTI_M_PER_SEC_SQ
cm/s2
Assume conversion of angular to linear acceleration is performed. Also, assume conversion of [g] G’s to cgs cm/s2.
Kinemyography (piezo) (angular velocity and/or displacement of thumb)
KMG
Phonomyography (lower frequency sounds, condenser or piezo)
PMG
Compressomyography (Modified NIBP cuff)
CMG
MDC_QRY_HDIALY_RX_QUERY^Dialysis Prescription Query^MDC (hemodialysis)
MDC_QRY_PDIALY_RX_QUERY^Peritoneal Dialysis Prescription Query^MDC (peritoneal dialysis)
The EMR system copies this value into the QAK-3 filed in the response.
The QPD-1 MDC_QRY codes play a similar role to the ORB-4 MDC_OBS codes we have defined in the past - they provide a high-level indication of the clinical intent of the QPD-1 query or OBR-4 observation set that can facilitate more precise message modeling and validation as well as facilitate rapid decoding by receivers.
I plan to define the two MDC_QRY_ codes starting at 1::3648, which is 64 codes above 1::3584 used for MDC_OBS Observation sets that can be conveyed by OBR-4. If you have any other guidance, please let me know.
Table Metadata
Table | Steward | V3 Harmonization | V3 Equivalent | Where used | Status |
0471 | InM | TBD | TBD | QPD-1, QAK-3, QID-2 | Active |
User-defined Table 0471 - Query Name
Value | Description | Comment |
| No suggested values defined |
|
All new 10101b MDC_EVT terms with be listed together with MDC_EVT terms from ISO/IEEE 11073-10101:2020, with full Definitions, Systematic Names, etc. A summary REFID, Disc, Part::Code and CF_CODE10 listing will also be provided, similar to what can be found in 10101:2020 Annex C. An example HTML listing with term grouping is attached.
The principal rationale for this approach is that the existing 545 MDC_EVT terms from 10101:2020 would not be included for brevity, focusing only only the 819 new terms for 10101b (which, after all, is an amendment). The summary REFID and code listing would include both with grouping of like-kind terms.
<MakeEventList.x1d.2022-07-12T17.html>
Stefan Schlichting, IEEE 11073 PoCD Working Group chair, will initiate the online PoCD WG ballot tomorrow. During the next two RTM tcons (this Wednesday, September 7, and next) I will provide a brief overview and answer any questions that you have. We will also devote part of the IEEE 11073 PoCD and GC meeting on September 19 or 20 in Baltimore and successfully conclude WG and SC “proceed with balloting” votes at that time.
I would like to thank all of you in the IHE DEV, IHE DEV PCD and RTM community for contributions over these many years!
Thanks and regards,
This nomenclature amendment includes significant extensions to support:
Again, I would like to thank all of you in the IHE DEV, IHE DEV PCD and RTM and IEEE 11073 community for your contributions over the past many years to this effort!
Thanks and regards,
During the Baltimore meeting, Elliot Silver (ResMed) mentioned that we were missing several patient positions, such as right- and left-lateral (recumbent), and that they should be added to our list of enumerated values for [MDC_ATTR_PT_BODY_POSN]. I have also added more precise Fowlers terms that can be used in lieu of “sitting" and “supine”. For the four Fowlers positions, I have also added specific angles for both beds and wheelchairs, since they are reported differently.I believe the list below is sufficient for the majority of acute and home care settings, but we should consider additional patient positioning terms (with references that define them along with their clinical utility) if we are missing any. [The original list had seven terms, based on the earlier IHE PCD SpO2 supplement, and were appropriate for reporting patient position for SpO2, NIBP and many other measurements in most acute care settings.] For example, should we consider adding common surgical positioning terms like “Lithotomy Position”, “Jackknife Position”, “Kidney Position”, “Sim’s Position” and possibly others? [This barely scratches the surface of the positioning capabilities of modern-day surgical tables, and it may be better to leave MDC_ATTR_PT_BODY_POSN to common acute and home-health care positions for now.]Table A.3.2.1.2.1.1—Enumerated values for [MDC_ATTR_PT_BODY_POSN]
Enum_Values
Enum_Descriptions
Comments
sitting
Sitting (also known as Fowler’s position)
standing
Standing
supine
Supine, lying face upward.
prone
Prone, lying face downward.
reverseTrendelenburg
The reverse Trendelenburg position is surgical position in which the lower extremities are leveled lower than the head and neck. It is the opposite of the Trendelenburg position, in which the head and the neck are below the lower extremities.
TrendelenburgPosition
In the Trendelenburg position the body is laid flat on the back (supine position) with the feet higher than the head by 15 to 30 degrees. This is a standard position used in abdominal and gynecological surgery. It allows better access to the pelvic organs as gravity pulls the intestines away from the pelvis. It was named after the German surgeon Friedrich Trendelenburg.
lowFowlersPosition
The Low-Fowler's position is when a patient is lying in bed in a supine position with the head of the bed is raised at approximately 15 to 30 degrees. [This is equivalent to 150 to 175 degrees of open seat to back angle in a wheelchair.]
semiFowlersPosition
The Semi-Fowler's position is when a patient is lying in bed in a supine position with the head of the bed at approximately 30 to 45 degrees. Patients who are on tube feedings are typically placed in the Semi-Fowler's position. Were the patient to lay flat the tube feeding fluid can possibly run into the lungs. Elevating the head of the bed to 30 degrees decreases the risk of the patient aspirating the tube feeding fluid. [This is equivalent to 135 to 150 degrees of open seat to back angle in a wheelchair.]
standardFowlers
The Standard-Fowler's position is when a patient is sitting in bed with the head of the bed at approximately 45 to 60 degrees. [This corresponds to 120 to 145 degrees of open seat to back angle in a wheelchair.]
highFowlers
The High-Fowler's position is when a patient is sitting in bed with the head of the bed at approximately 90 degrees. [This corresponds to 90 degrees of open seat to back angle in a wheelchair.]
rightLateralRecumbent
The Right Lateral Recumbent (RLR) is when the patient is lying on their right side.
lefttLateralRecumbent
The Left Lateral Recumbent, or LLR, is when the patient is lying on their left side.
Several people recommended that we assign globally unique numeric codes for enumerated values. This would be more aligned with other vocabulary standards such as LOINC, where LOINC “answer” strings and groups are now assigned unique alpha-numeric identifier codes.Here’s an example of how this will be specified in IEEE P11073-10101b:For the HL7 V2 message examples shown below, the following enumerated values for MDC_NCC_WIRELESS_AUTH are used. The Part::Code and CF_CODE10 values are example values.
Table A.3.2.13.2.3—Enumerated values for [MDC_NCC_WIRELESS_AUTH]
Enum_Values
Enum_Descriptions
Part::Code
CF_CODE10
OPEN
No security
7::18471
477223
WEP
Wired Equivalent Privacy
7::18472
477224
WPA-PSK
Wi-Fi Protected Access (WPA) - Pre-shared key exchange
7::18473
477225
WPA-ENT
Wi-Fi Protected Access (WPA) – Enterprise authentication
7::18474
477226
WPA2-PSK
Wi-Fi Protected Access II (WPA2) - Pre-shared key exchange
7::18475
477227
WPA2-ENT
Wi-Fi Protected Access II (WPA2) – Enterprise authentication
7::18476
477228
WPA3-PSK
Wi-Fi Protected Access® (WPA3) – Personal mode, Simultaneous Authentication of Equals (WPA3-SAE)
7::18477
477229
WPA3-ENT
Wi-Fi Protected Access® 3 (WPA3) – Enterprise authentication
7::18478
477230
This will support both the traditional encoding of sending the Enum_Value token as a constrained string HL7 v2 ‘ST’ datatype:
OBX|3|ST|69418^MDC_NCC_WIRELESS_AUTH^MDC|1.0.1.5|WPA2-ENT||||||F
Or as a fully populated HL7 V2 ‘CWE’ datatype with a globally unique context-free numeric code defined in Partition 7:OBX|3|CWE|69418^MDC_NCC_WIRELESS_AUTH^MDC|1.0.1.5|477228^WPA2-ENT^MDC||||||F
Although it would be possible to automatically generate a unique REFID for each { _ENUM_GROUP :: Enum_Value } pair, it is not clear that sending it the message would provide much validation benefit nor clarity compared to sending the shorter Enum_Value string as we currently do today.
<P11073_10101b.TopicSummary.1e.2022-09-22T13.pdf>
Enum_Descriptions |
|
unknown |
The patient position is unknown or cannot be determined based on the information available to the device (sensor, manual entry or by other means). |
upright |
The patient's torso is in an upright or nearly vertical orientation, regardless of the position of the lower extremeties (e.g. sitting or standing). |
sitting |
Sitting (also known as Fowler’s position) |
standing |
Standing |
supine |
Supine, lying face upward. |
prone |
Prone, lying face downward. |
reverseTrendelenburg |
In the reverse Trendelenburg position, the patient is lying flat on the back (supine position) with the head and neck elevated with respect to the lower extremeties. It is the opposite of the Trendelenburg position, in which the head and the neck are below the lower extremities. |
Trendelenburg |
In the Trendelenburg position, the patient is lying flat on the back (supine position) with the feet higher than the head by 15 to 30 degrees. This is a standard position used in abdominal and gynecological surgery. It allows better access to the pelvic organs as gravity pulls the intestines away from the pelvis. It was named after the German surgeon Friedrich Trendelenburg. |
lowFowlers |
In the low Fowler's position, the patient is lying in bed in a supine position with the head of the bed is raised at approximately 15 to 30 degrees. [This is equivalent to 150 to 175 degrees of open seat to back angle in a wheelchair.] |
semiFowlers |
In the semi-Fowler's position, the patient is lying in bed in a supine position with the head of the bed at approximately 30 to 45 degrees. Patients who are on tube feedings are typically placed in the Semi-Fowler's position. Were the patient to lay flat the tube feeding fluid can possibly run into the lungs. Elevating the head of the bed to 30 degrees decreases the risk of the patient aspirating the tube feeding fluid. [This is equivalent to 135 to 150 degrees of open seat to back angle in a wheelchair.] |
standardFowlers |
In the standard Fowler's position, the patient is sitting in bed with the head of the bed at approximately 45 to 60 degrees. [This corresponds to 120 to 145 degrees of open seat to back angle in a wheelchair.] |
highFowlers |
In the high Fowler's position, the patient is sitting in bed with the head of the bed at approximately 90 degrees. [This corresponds to 90 degrees of open seat to back angle in a wheelchair.] |
orthopneicPosition |
In the orthopneic or tripod position, the patient is in a sitting position or on the side of the bed with an overbed table in front to lean on and several pillows on the table to rest on. |
lateralRecumbent |
In the lateral recumbent (decubitus) position, the patient lies on one side of their body with the top leg in front of the bottom leg and the hip and knee flexed. Flexing the top hip and knee and placing this leg in front of the body creates a wider, triangular base of support and achieves greater stability. |
rightLateralRecumbent |
In the right lateral recumbent (decubitus) position, the patient is lying on their right side. |
lefttLateralRecumbent |
In the left lateral recumbent (decubitus) position, the patient is lying on their left side. |
SimsPosition |
In the Sim's (semi-prone) position, the patient assumes a posture halfway between the lateral and the prone positions. The lower arm is positioned behind the client, and the upper arm is flexed at the shoulder and the elbow. The upper leg is more acutely flexed at both the hip and the knee, than is the lower one. |
lithotomyPosition |
in the lithotomy position, the patient is on their back with hips and knees flexed and thighs apart. |
knee-chestPosition |
The knee-chest position can be in lateral or prone position. In lateral the patient lies on their side, torso lies diagonally across the table, hips and knees are flexed. In prone, the patient kneels on the table and lower shoulders on to the table so chest and face rests on the table. |
jackknifePosition |
In the jackknife position (also known as Kraske position), the patient's abdomen lies flat on the bed. The bed is scissored so the hip is lifted and the legs and head are low. |
kidneyPosition |
In the kidney position, the patient assumes a modified lateral position wherein the abdomen is placed over a lift in the operating table that bends the body. Patient is turned on their contralateral side with their back placed on the edge of the table. |
On Dec 2, 2022, at 10:51 AM, Paul Schluter <pssch...@gmail.com> wrote:
Greetings!
As those of you that are on the ballot group are likely aware, balloting was opened this morning for IEEE P11073-10101b-D1.The ballot closes at 23:59 UTC-12 on December 31, 2022.If you have any questions, please do not hesitate to contact me, Stefan Schlichting <stefan.sc...@ornet.org> or Tom Thompson" < thomas....@ieee.org>. For those of you who have not joined the IEEE-SA balloting group for IEEE P11073-10101b but would like to obtain a copy of the document as a potential “Public Reviewer”, please contact Tom Thompson or the IEEE-SA regarding that option.Thank you all for your thoughtful contributions and discussions over these many years, and have an enjoyable Holiday Season!
On Dec 7, 2022, at 8:43 AM, Paul Schluter <pssch...@gmail.com> wrote:
Greetings!
The RTM tcon for this Wednesday, December 7th, is cancelled. I do not have any new topics to discuss.
Balloting was opened for IEEE P11073-10101 on December 2nd and will close on December 31st. Please feel free to contact me or Stefan Schlichting <stefan.sc...@ornet.org> if you have any questions.
Thanks again for all your help and best regards,
Paul Schluter
email: pssch...@gmail.comcell: (414) 702-2026On Dec 2, 2022, at 10:51 AM, Paul Schluter <pssch...@gmail.com> wrote:Greetings!As those of you that are on the ballot group are likely aware, balloting was opened this morning for IEEE P11073-10101b-D1.The ballot closes at 23:59 UTC-12 on December 31, 2022.If you have any questions, please do not hesitate to contact me, Stefan Schlichting <stefan.sc...@ornet.org> or Tom Thompson" < thomas....@ieee.org>. For those of you who have not joined the IEEE-SA balloting group for IEEE P11073-10101b but would like to obtain a copy of the document as a potential “Public Reviewer”, please contact Tom Thompson or the IEEE-SA regarding that option.Thank you all for your thoughtful contributions and discussions over these many years, and have an enjoyable Holiday Season!
On Mar 1, 2023, at 8:35 AM, Paul Schluter <pssch...@gmail.com> wrote:
Greetings!
First, the RTM tcons for today, Wednesday, March 1st and next Wednesday, March 8th, are cancelled, since IEEE P11073-10101b-D2 is out for recirculation ballot until 23:59 UTC-12 on 08 Mar 2023.It is likely that the next RTM tcon will be held on Wednesday, March 15th, since by then we will have had time to review new comments and propose preliminary resolutions.
I would like to thank everyone who have taken the time to review IEEE P11073-10101b-D2. If you have any questions about the ballot, please feel free to contact me, Tom Thompson at thomas....@ieee.org or Stefan Schlichting at stefan.sc...@ornet.org .
And finally, Kathy and I had an absolutely wonderful time in Egypt and Jordan, and we’re now fully refreshed and looking forward to reengaging with all our activities, friends and colleagues back home.
Thanks and regards,
Paul Schluter
email: pssch...@gmail.comcell: (414) 702-2026
On Feb 1, 2023, at 8:17 AM, Paul Schluter <pssch...@gmail.com> wrote:
Greetings!First, the RTM tcons for today, Wednesday, February 1st and the next three Wednesdays, February 8, 15 and 22nd, are cancelled.My wife Kathy and I are looking forward to our trip to Egypt and Jordan over the next several weeks. We plan to return on February 22nd and it is likely that the next RTM tcon will be held on Wednesday, March 1st.Second, I am very happy to say that IEEE P11073-10101b-D2 has been submitted to the IEEE-SA and we expect the recirculation ballot to begin very shortly.I would like to thank everyone who submitted ballot comments and participated in our ballot reconciliation discussions last week. We were able to come to a resolution on practically all of the comments and believe that the updated draft is ready for recirculation. If you have any questions about the ballot, please contact Tom Thompson at thomas....@ieee.org or Stefan Schlichting at stefan.sc...@ornet.org .
Join on your computer, mobile app or room device (RTM meeting time: Wednesdays at 11:00 AM Central US)
11073 PoCD Working Group / STANDARDS DEV / NOMENCLATURE SG / 10101b
<<P11073-10101b-D3.2023-05-01T11.clean.pdf>><<P11073-10101b-D3.2023-05-01T11.redline.pdf>>
Monday, May 8th, 2023 11:00 AM to 5 PM US CSTPlease refer to the attached <2023-05 POCD 11073--IEEE SA_draft Agenda-members v1.doc> agenda, drafted by Konstantinos “Kosta” Makrodimitris, PhD, of the FDA.I will give a short presentation of IEEE P11073-10101b as part of the RTM update, starting at 13:45 PM US CST, followed by more in-depth discussion of selected comments and proposed dispositions.Tuesday, May 9th - Thursday, May 11th 9:00 AM to 5 PM US CST, using HL7 “Quarters” (there could be a meeting on Friday, too)The draft agenda for the DEV SIG meetings is available at: https://confluence.hl7.org/pages/viewpage.action?pageId=161061969
Due to a planned NIST Systems Outage for maintenance, RTMMS will be unavailable for use this weekend (starting June 16th and ending June 20th).
Michael Faughn
11073 PoCD Working Group / STANDARDS DEV / NOMENCLATURE SG / 10101b
<<P11073-10101b-D4.2023-06-12T10.clean.pdf>><<P11073-10101b-D4.2023-06-12T10.tracked.pdf>>