Hi Renee,
You are correct that PID-10.1 must be “2106-3” because the Data Categorization is Test Case Fixed Data.
The Data Categorization for PID-10.2 is Changeable, which means the exact content is not relevant for the Test Case, it can be changed for the purposes of testing, and that example data are provided. The test tool validates for presence of data, and the Tester will manually validate that the content is appropriate for the code 2106-3. When PID-10.1 = 2106-3, then PID-10.2 could be “White” or other descriptions that are equivalent such as “Caucasoid” or “Caucasian”.
On slide 18 in the 2015 Edition Understanding ELR ONC Certification Testing ppt slide deck (accessible on the Documentation Page) is a full explanation of the Data Categorization used for the NIST ELR Test Tool.
Thank you for your question,
Sheryl
NIST
--
You received this message because you are subscribed to the Google Groups "HL7v2 Reportable Lab Testing" group.
To unsubscribe from this group and stop receiving emails from it, send an email to
hl7v2-reportable-lab...@googlegroups.com.
To view this discussion on the web visit
https://groups.google.com/d/msgid/hl7v2-reportable-lab-testing/eac60cf1-8225-48fe-b5a7-62187a42caf4n%40googlegroups.com.
To view this discussion on the web visit https://groups.google.com/d/msgid/hl7v2-reportable-lab-testing/718ee09a-1cd4-4ab3-a0dd-a4ae4be29d7en%40googlegroups.com.
Yes.
A better term for this categorization is “Presence-Test Case Proper” instead of “Changeable data”. Meaning that the data is expected for this element and the data needs to be appropriate for the element within the context of the test case. So, in this case the text for this element needs to be appropriate for the coded element in PID-10.1. A test proctor may visually inspect this element to ensure this but that tool validation will not.
Later versions of our tooling uses the term “Presence-Test Case Proper” instead of “Changeable data” to better reflect this interpretation.
Rob
NIST
From: 'Renee Clark' via HL7v2 Reportable Lab Testing <hl7v2-reportab...@googlegroups.com>
Sent: Wednesday, May 25, 2022 9:35 AM
To: HL7v2 Reportable Lab Testing <hl7v2-reportab...@googlegroups.com>
Subject: Re: ELR field PID 10.2 for Race
ok. So If I run a text file with PID 10.1 = 2028-9 and PID 10.2 =West Asian, the text file will not give an error in the tool. Right? that is how I am interpreting "Changeable Data" for that field.