Good evening Lex,

Love tinkering with this device and I really want to take it up a notch for our lab. I was playing with the code in Free Feeding v1.8.5, but seemed to be stumped. I'm pairing the FED3 with an external hardware that will be connected via BNC. I want the hardware to receive input from FED3 when a mouse removes a pellet by sending a pulse. A step further would be to have FED3 not dispense another pellet for a set amount of time because we are using fiber photometry to record the before, during, and after of each epoch of pellet retrieval. How might one code for this? I tried the timeout function and thought it would solve it but I end up telling it to timeout before it sends a pulse. 

Thank you Dr. Kravitz for your help and support and letting me join the group!

Fond regards,

Miguel D.  

Lex Kravitz

unread,
Oct 25, 2021, 11:34:18 PM10/25/21
to FEDforum
Hi Miguel!
You want the FED3 to send a pulse immediately after the pellet is removed, and then also do a Timeout?  There are three functions we need to juggle into the right order.  (BTW you can see all the functions for the FED3 library here). The three relevant functions are:
1) Feed, which runs the feeding routine to drop a pellet.
2) BNC, which controls pulses from the BNC output.  For instance, you can call the BNC function as:
fed3.BNC(200,1) to deliver a single 200ms pulse or fed3.BNC(100,10) to deliver ten 100ms pulses
3) Timeout runs a timeout routine, basically pausing the FED3 for a specified time.

The complication you're experiencing is because the Feed routine automatically calls Timeout.  But you want Feed, then BNC, and then Timeout.  Long-term I should probably make it so Feed doesn't automatically call Timeout but for now that's how it works. I did it that way because it makes the high level code clean.  I made the following edits to the Free-feeding script to do what you want (noted in yellow arrows below):
1) I turned off the Timeout call in the Feed function, by setting fed3.timeout = 0.  Now when Feed calls TImeout the timeout duration will be 0s, or essentially off.
2) I added fed3.BNC(200,1) to the void loop() after the Feed function, which will deliver a single 200ms pulse immediately after the pellet is removed.  You can edit this pulse call how you want.
3) I added delay (115000) to the code to replace the timeout function we removed - delay is a built in Arduino function that pauses the microprocessor.  So this will pause the FED3 for 115 seconds here, after which it will resume at the top of the void loop().