Slit lamp beam

13 views
Skip to first unread message

Matthew Kwong

unread,
Mar 24, 2021, 8:01:32 AM3/24/21
to Eye Care Equipment Maintainers

HI group, hope you are all well
I have a slit reichert slit lamp that does not focus a slit optic precisely.  It looks like 2 beams when narrowed to 1 mm optic sections.  I have not found anything on this site, but if there is such discussion on how a slit lamp works, or how to adjust, I would appreciate such link.

Thanks

Rob Martin

unread,
Mar 24, 2021, 6:56:55 PM3/24/21
to eye-care-equipm...@googlegroups.com
Hi Matthew,

Is it possible the lamp or reflector is not aligned and you could be seeing two light sources, the filament and its reflection from the reflector? There should be just one filament with the reflected filament directly behind the lamp filament. You could use a paper to determine if the two beams are two filaments at a low light setting. As you move the paper away from the slit lamp the filiments will become visible.  The 1 mm slit is not narrow enough to form a single slit diffraction pattern on the paper.



RobM

--
You received this message because you are subscribed to the Google Groups "Eye Care Equipment Maintainers" group.
To unsubscribe from this group and stop receiving emails from it, send an email to eye-care-equipment-ma...@googlegroups.com.
To view this discussion on the web visit https://groups.google.com/d/msgid/eye-care-equipment-maintainers/7cf0c75c-07d8-48e6-aef3-fe1bea6a51b6n%40googlegroups.com.
Screenshot_20210324-195530_Samsung Notes.jpg

Ganga Saravanaa

unread,
Mar 30, 2021, 6:48:13 AM3/30/21
to eye-care-equipm...@googlegroups.com
Hi Matthew,
I hope, you solved your slitlamp complaint.
If not yet, please  check the binocular holding screw (very important never open or disturb the binocular optics).
Please do the following 
Step 1: fix the slit lamp test rod.
Step 2: keep the thin (narrow) slit beam on test rod.
Step 3: slightly open the binocular holding screw then move the binocular front and back slowly upto you get the focused slit beam on the test rod.(make sure always hold binocular after open the screw
and never take away).  

With regards 
Ganga 
 

--

Srinivasan

unread,
Apr 4, 2021, 10:26:10 AM4/4/21
to Eye Care Equipment Maintainers
Dear all: I should have responded to this long ago. Because of my age I am not able to check my mail regularly and some mail gets unattended. If you have an assess to the Aravind CD please look at it and you would find the solution. I do not have facilities to send you the CD. Please request some one in Madurai to send you the full CD or the section relating to slit lamp. Best wishes Srinivasan