Hi All,
can anyone provide a test and check procedure for an operating microscopes?
currently ours on a service contract with companies but looking to do the routine maintenance myself.
thanks in advance
Darren
medical physics
Tennant institute of Ophthalmology
Gartnavel Hospital
Glasgow
Scotland
****************************************************************************
NHSGG&C Disclaimer
The information contained within this e-mail and in any attachment is
confidential and may be privileged. If you are not the intended
recipient, please destroy this message, delete any copies held on your
systems and notify the sender immediately; you should not retain, copy
or use this e-mail for any purpose, nor disclose all or any part of its
content to any other person.
All messages passing through this gateway are checked for viruses, but
we strongly recommend that you check for viruses using your own virus
scanner as NHS Greater Glasgow & Clyde will not take responsibility for
any damage caused as a result of virus infection.
**************************************************************************
--
You received this message because you are subscribed to the Google Groups "Eye Care Equipment Maintainers" group.
To unsubscribe from this group and stop receiving emails from it, send an email to eye-care-equipment-ma...@googlegroups.com.
For more options, visit https://groups.google.com/d/optout.
Srinivasan & Robin
Thank you so much for your input. as I expected not out with my capabilities. not too dissimilar to slit lamp maintenance.
Robin very good point about software updates as some of the modern ziess scopes do get regular updates.
I do have the luxury of being able to contact the company for repairs if I run into serious difficulty.
thanks again
Darren
From:
eye-care-equipm...@googlegroups.com
[mailto:eye-care-equipm...@googlegroups.com] On Behalf Of Srinivasan
Sent: 12 March 2015 05:01
To: eye-care-equipm...@googlegroups.com
Subject: [eye-care-equipment-maintainers:309] Re: operating microscope
Dear Darren: I would have the following checklist with me for the routine maintenance of an operating microscope.
1.Check the free movement of the microscope as a whole on its wheels. Some times the movement may not be smooth. Remove the dust and lint that may be clogging the wheels, gently lubricate the axle of the wheels, remove excess lubrication
2.Check the key that arrests locks the above movement. Check the locking and unlocking mechanism once or twice and remove any obstruction
3. Check the up down movement of the body of the microscope on the vertical stand.
4. Check the locking mechanism
5. Check the free movement of the arm that carries the microscope (optics) and check its locking mechanism (side ways and up down on the rotating arm)
6. Check the light source assembly. Make sure that all the bulbs ( 2 or 3 as the case may be) are in good condition. Any fused bulb to be replaced,
7. Check the cooling fan and the air vent. Remove dust and accumulated lint
8.Check the on off switch
9. Check that the optic fiber is firmly located (pushed in) at the two ends and make sure the fiber doesn't come in the way of free movement of the arm or gets twisted as the arm is moved
10.Check the X-Y and Up-down movements of the optics of the microscope in response to food pedal operations
11. Check the objective lens. If it requires cleaning clean it using mild soap solution. If is clean don't disturb it. There is no need to remove the objective lens from its position
12. Check the eye piece assembly. Clean the eye piece lenses (outside only) and check the free movement to change the IPD as desired
13. Check the various additions like the beam splitter, the camera attachment, assistants scope etc. Remove them all if they are not to be used and close all ports
14. Use a bag to cover the optics of the microscope when not in use
15. Keep a small bag of silica gel in the vicinity of the microscope
16. Remove the microscope from the operation theatre to a dry room if the theatre is to be washed and cleaned. Shift the microscope in to the theatre only after the theatre is dry.
17. Very rarely the axis of the beam of light that comes out from the optic fiber through the objective lens and axis of the microscope may not be aligned properly. This requires manipulation of certain screws positioning the end of the optic fiber by trial and error. This requires some skill that one can pick up by practice.
Best of luck. V.Srinivasan
On Wednesday, 11 March 2015 03:05:18 UTC-7, big daddy wrote:
--
You received this message because you are subscribed to the Google Groups
"Eye Care Equipment Maintainers" group.
To unsubscribe from this group and stop receiving emails from it, send an email
to eye-care-equipment-ma...@googlegroups.com.
For more options, visit https://groups.google.com/d/optout.
****************************************************************************
NHSGG&C Disclaimer