McMillan, Darren

unread,
Oct 9, 2015, 8:11:52 AM10/9/15
to eye-care-equipm...@googlegroups.com

Hi all,

 do any group members have instructions on how to recalibrate  a Perkins style tonometer?

 

I have lots of experience recalibration goldmann stlye but no experience with Perkins.

 

to be clear its is recalibration of error I'm interested in learning, not verifying the calibration :)

 

Thanks in advance

 

greetings from Scotland

 

Darren

****************************************************************************
NHSGG&C Disclaimer

The information contained within this e-mail and in any attachment is
confidential and may be privileged. If you are not the intended
recipient, please destroy this message, delete any copies held on your
systems and notify the sender immediately; you should not retain, copy
or use this e-mail for any purpose, nor disclose all or any part of its
content to any other person.

All messages passing through this gateway are checked for viruses, but
we strongly recommend that you check for viruses using your own virus
scanner as NHS Greater Glasgow & Clyde will not take responsibility for
any damage caused as a result of virus infection.

**************************************************************************

Srinivasan

unread,
Oct 9, 2015, 11:59:10 PM10/9/15
to Eye Care Equipment Maintainers
Dear Darren: Goldmann tonometer is all mechanical instrument and due to wear and tare or displacement of moving parts the instruments need recalibration and it can be done. However Perkins tonometer is an electronic instrument. It should not require recalibration as long as the electronics functions. There is no wear and tare. If electronics fails the instrument will have to be discarded unless one can get the spares for the electronic board with the IC chips.Srinivasan

McMillan, Darren

unread,
Oct 12, 2015, 4:43:18 AM10/12/15
to eye-care-equipm...@googlegroups.com

Hi thanks for your reply, I was under the impression that this style of perkins were mechanical and could be recalibrated.

 

Image result for perkins tonometer instruction manual

 

I know there are electronic perkins which you can do nothing with.

 

                Thanks again

 

Darren

From: eye-care-equipm...@googlegroups.com [mailto:eye-care-equipm...@googlegroups.com] On Behalf Of Srinivasan
Sent: 10 October 2015 04:59
To: Eye Care Equipment Maintainers
Subject: [eye-care-equipment-maintainers:326] Re: RECALIBRATION OF PERKINS TONOMETER

 

Dear Darren: Goldmann tonometer is all mechanical instrument and due to wear and tare or displacement of moving parts the instruments need recalibration and it can be done. However Perkins tonometer is an electronic instrument. It should not require recalibration as long as the electronics functions. There is no wear and tare. If electronics fails the instrument will have to be discarded unless one can get the spares for the electronic board with the IC chips.Srinivasan

On Friday, 9 October 2015 05:11:52 UTC-7, big daddy wrote:

--
You received this message because you are subscribed to the Google Groups "Eye Care Equipment Maintainers" group.
To unsubscribe from this group and stop receiving emails from it, send an email to eye-care-equipment-ma...@googlegroups.com.
For more options, visit https://groups.google.com/d/optout.

Philip Hoare

unread,
Oct 12, 2015, 5:06:18 AM10/12/15
to eye-care-equipm...@googlegroups.com

Dear Darren,

 

I have been in touch with my friends at HS UK (formerly Clement Clarke) and they have just emailed me back asking if your instrument is a MK2 or MK 3. I am not sure whether I would get a better answer than the one received from Prof Srinivasan who I acknowledge is the fountain of all knowledge on such equipment, but you never know.

 

I look forward to hearing from you.l

 

Kind regards,

 

Phil

 

Philip Hoare

Procurement Coalition Manager

International Agency for the Prevention of Blindness (IAPB)

 

Tel:      +44 (0)7738-785-326

Email: pho...@iapb.org

Skype: philhoare_iapb

IAPB Standard List: http://iapb.standardlist.org

Member of the Inter-Agency Procurement Group (IAPG): www.iapg.org.uk

10GA LOGO 27-30 OCT (00000002)

www.iapb.org   Follow us on Twitter and Facebook

McMillan, Darren

unread,
Oct 12, 2015, 6:19:57 AM10/12/15
to eye-care-equipm...@googlegroups.com

hi thanks, we have both models in use but MK 2 is more common.

 

 this is not to repair specific item but more for my own info and for future use.  I'm sure HS will recalibrate (for a price) so not to sure if they will give instructions…

 

thanks for your efforts

 

Darren

 


Sent: 12 October 2015 10:06
To: eye-care-equipm...@googlegroups.com

Subject: RE: [eye-care-equipment-maintainers:327] Re: RECALIBRATION OF PERKINS TONOMETER

 

Warning: This message contains unverified links which may not be safe.  You should only click links if you are sure they are from a trusted source.

Dear Darren,

 

I have been in touch with my friends at HS UK (formerly Clement Clarke) and they have just emailed me back asking if your instrument is a MK2 or MK 3. I am not sure whether I would get a better answer than the one received from Prof Srinivasan who I acknowledge is the fountain of all knowledge on such equipment, but you never know.

 

I look forward to hearing from you.l

 

Kind regards,

 

Phil

 

Philip Hoare

Procurement Coalition Manager

International Agency for the Prevention of Blindness (IAPB)

 

Tel:      +44 (0)7738-785-326

Email: pho...@iapb.org

Skype: philhoare_iapb

IAPB Standard List: http://iapb.standardlist.org

Member of the Inter-Agency Procurement Group (IAPG): www.iapg.org.uk

10GA LOGO 27-30 OCT (00000002)

www.iapb.org   Follow us on Twitter and Facebook

 

From: eye-care-equipm...@googlegroups.com [mailto:eye-care-equipm...@googlegroups.com] On Behalf Of McMillan, Darren


Sent: 12 October 2015 09:43
To: 'eye-care-equipm...@googlegroups.com'

Philip Hoare

unread,
Oct 12, 2015, 6:22:43 AM10/12/15
to eye-care-equipm...@googlegroups.com

Let’s see – I will come back to you as soon as I get a response from HS UK.

Philip Hoare

unread,
Oct 12, 2015, 6:50:04 AM10/12/15
to eye-care-equipm...@googlegroups.com

Dear Darren,

 

Please find attached the PERKINS Tonometer MKll Instructions Manual – Calibration is on Page 11 onwards.

 

Kind regards,

 

Phil

 

Philip Hoare

Procurement Coalition Manager

International Agency for the Prevention of Blindness (IAPB)

 

Tel:      +44 (0)7738-785-326

Email: pho...@iapb.org

Skype: philhoare_iapb

IAPB Standard List: http://iapb.standardlist.org

Member of the Inter-Agency Procurement Group (IAPG): www.iapg.org.uk

10GA LOGO 27-30 OCT (00000002)

www.iapb.org   Follow us on Twitter and Facebook

 

From: Philip Hoare

Sent: 12 October 2015 11:22
To: 'eye-care-equipm...@googlegroups.com'

Subject: RE: [eye-care-equipment-maintainers:328] Re: RECALIBRATION OF PERKINS TONOMETER

 

Let’s see – I will come back to you as soon as I get a response from HS UK.

 

Perkins MKII Instructions.pdf

McMillan, Darren

unread,
Oct 12, 2015, 7:05:46 AM10/12/15
to eye-care-equipm...@googlegroups.com

Hi Phil, this is the procedure to verify the calibration. I am familiar with this.

 

 what I am looking for is instructions on how to recalibrate if device fails this check.

 

really appreciate your efforts

 

Thanks

 

Darren

 

From: eye-care-equipm...@googlegroups.com [mailto:eye-care-equipm...@googlegroups.com] On Behalf Of Philip Hoare
Sent: 12 October 2015 11:50
To: eye-care-equipm...@googlegroups.com
Subject: RE: [eye-care-equipment-maintainers:330] Re: RECALIBRATION OF PERKINS TONOMETER

 

Warning: This message contains unverified links which may not be safe.  You should only click links if you are sure they are from a trusted source.

Dear Darren,

 

Please find attached the PERKINS Tonometer MKll Instructions Manual – Calibration is on Page 11 onwards.

 

Kind regards,

 

Phil

 

Philip Hoare

Procurement Coalition Manager

International Agency for the Prevention of Blindness (IAPB)

 

Tel:      +44 (0)7738-785-326

Email: pho...@iapb.org

Skype: philhoare_iapb

IAPB Standard List: http://iapb.standardlist.org

Member of the Inter-Agency Procurement Group (IAPG): www.iapg.org.uk

10GA LOGO 27-30 OCT (00000002)

www.iapb.org   Follow us on Twitter and Facebook

 

From: Philip Hoare

Sent: 12 October 2015 11:22
To: 'eye-care-equipm...@googlegroups.com'

Subject: RE: [eye-care-equipment-maintainers:328] Re: RECALIBRATION OF PERKINS TONOMETER

 

Let’s see – I will come back to you as soon as I get a response from HS UK.

 

Philip Hoare

unread,
Oct 12, 2015, 7:07:32 AM10/12/15
to eye-care-equipm...@googlegroups.com

Oh ok. Must admit, I just turned it around quickly…in a middle of a 3 day conference. I will get back to HS and come back to you.

McMillan, Darren

unread,
Oct 12, 2015, 7:09:06 AM10/12/15
to eye-care-equipm...@googlegroups.com

no pressure… really appriciate your efforts

 

thanks again

 

Darren

 

From: eye-care-equipm...@googlegroups.com [mailto:eye-care-equipm...@googlegroups.com] On Behalf Of Philip Hoare
Sent: 12 October 2015 12:07
To: eye-care-equipm...@googlegroups.com
Subject: RE: [eye-care-equipment-maintainers:332] Re: RECALIBRATION OF PERKINS TONOMETER

 

Warning: This message contains unverified links which may not be safe.  You should only click links if you are sure they are from a trusted source.

Oh ok. Must admit, I just turned it around quickly…in a middle of a 3 day conference. I will get back to HS and come back to you.

Reply all
Reply to author
Forward
0 new messages