Problem: Chamber A cannot backfill to 70 Torr

0 views
Skip to first unread message

EE

unread,
Oct 23, 2007, 1:58:43 AM10/23/07
to EPI2
Ch. A failed initializing pressure subsystem....Other alarms on the
system were:
- Ch. A pressure control valve failed to close
- Ch. A press. cont. valve never reached home in minus direction
- Ch. A reduced pressure stepper failed

EE

unread,
Nov 6, 2007, 5:19:02 PM11/6/07
to EPI2
This alarm was caused by a dirty throttle. We removed the throttle
valve and cleaned it by soaking it DI water. Placed a dummy throttle
valve in placed of the dirty one and pumped down the foreline as soon
as possible in order to minimize exposing the line to air. Performed
a throttle valve calibration after re-installing the clean valve.