There are many factors:
- microscopes with LAB6 filaments rather than FEGs have a much stronger dependency between defocus and envelope function
- use of a K2 in counting mode dramatically improves SNR across the entire spectrum, extending the achievable resolution on any microscope not hard-limited by some specific parameter
- as you get further from focus, the CTF oscillations get much closer together at high resolution. There is a point at which this becomes uncorrectable, but this point depends on the box size the phase flipping is done on. This is one reason why some strategies for phase-flipping operate on the entire micrograph. Of course that won't save you if your particle is too close to the edge...
- The threshold for information recovery is highly specimen dependent. An icosahedral virus, for example, provides a strong enough per-particle alignment that fairly high resolution information may still be recoverable even when the envelope has fallen off substantially.
Yes, to an extent, but this is an imperfect science. If you have a highly monodisperse specimen frozen over holes with minimal degraded protein and minimal scattering from the buffer, then yes. You can look at the 1-D power spectrum computed by e2evalimage.py and assess where the CTF oscillations are no longer visible (or get too close together).
However, if you have a carbon substrate, or a lot of scattering from the buffer, then you need to carefully subtract these effects before you can assess the resolution potential. Typically this is done by boxing out particles and looking at the SSNR curves in e2ctf.py (Visual CTF in projectmanager). The SSNR explicitly takes the background into account, and is a pretty accurate estimate if the defocus is fit properly (which it normally should be). Our rule of thumb in the past has been that once the SSNR falls below ~0.02, the information is not readily recoverable, but there is a bit of variability in this cutoff.
In crystallography, the fact that you have crystal scattering peaks which you know must occur in specific locations gives you excellent separation from the background, and allows easy resolution assessment. Since the data is continuous and much noisier, there simply isn't a direct equivalent for CryoEM.
----------------------------------------------------------------------------
Steven Ludtke, Ph.D.
Co-Director National Center For Macromolecular Imaging
(
ncmi.bcm.edu)
Baylor College of Medicine