UDI Medical equipment error

43 views
Skip to first unread message

Dale Owens

unread,
May 22, 2025, 11:45:25 AMMay 22
to Edge Test Tool (ETT)
UDI organizer template ID is currently 2.16.840.1.113883.10.20.22.4.312 where it should be 2.16.840.1.113883.10.20.22.4.311.  This is based on the Appendix B: UDI Organizer template from C-CDA Supplemental Templates for Unique Device Identifier (UDI) R4.
If we update the templateId root to 2.16.840.1.113883.10.20.22.4.311 it fails conformance, where 2.16.840.1.113883.10.20.22.4.312 does not fail.  Can someone confirm which templateID root should be used?
P-4013781 - E-3907381_transformed_data - 6616920439798243982.xml

Kim Poletti

unread,
May 22, 2025, 11:48:03 AMMay 22
to Edge Test Tool (ETT)
Hi - Thanks for reaching out. This has been logged for review and a member of the team will reach out in the near future. For internal purposes only, this ticket # is SITE-4635.

kylem...@gmail.com

unread,
May 22, 2025, 12:57:18 PMMay 22
to Edge Test Tool (ETT)
Looking at your file, the issue that you are missing the required <id> element and also your statusCode is "active" when the IG says it should be "completed". If you add those and then use the .311 template ID, it does not have any errors on this section of the CCDA.

James Spillman

unread,
Jun 4, 2025, 4:19:52 PMJun 4
to Edge Test Tool (ETT)
Locking this conversation. If you are still having an issue please start a new conversation.