issue with ETT tool for 170.315_b1_toc_inp_svap_uscdiv3_sample1_v1

69 views
Skip to first unread message

Vishal Patel

unread,
Nov 26, 2025, 8:41:14 AMNov 26
to Edge Test Tool (ETT)
Hi team,
we're testing CDA for 170.315_b1_toc_inp_svap_uscdiv3_sample1_v1 sample and found below 2 issues, based on the data we sending the correct values but it failed.

PF beloww error, PFA CDA for the reference.
170.315_b1_toc_inp_svap_uscdiv3_sample1_v1_error.jpg

sample data:170.315_b1_toc_inp_svap_uscdiv3_sample1_datasample.jpg
Regards
Vishal.
170.315_b1_ToC_Inpatient_Sample_1_(v1)_new12.xml

Mat Davis

unread,
Nov 26, 2025, 7:41:04 PMNov 26
to Edge Test Tool (ETT)
Hi Vish.

Issue #1
  • Review your Entry for Food insecurity > Entry and EntryRelationship on line # 2905
  • The validator is expecting a separate Entry and EntryRelationship to reflect "238131007" aka Overweight
  • This is likely implied in the " USCDI Data Class/Element: HealthConcerns: " section

Issue #2
Thanks - Mat

Vishal Patel

unread,
Nov 27, 2025, 8:26:36 AMNov 27
to Edge Test Tool (ETT)
Hi Mat,
for issue#1,
this looks like as per the standard we are passing the data, hence it should not give any error from the validator.
ref: https://www.hl7.org/ccdasearch/pdfs/2.16.840.1.113883.10.20.22.4.132.pdf

HealthConcernAct_ProblemObservation.jpg

Regards,
Vishal.

Mat Davis

unread,
Nov 28, 2025, 4:55:26 AMNov 28
to Edge Test Tool (ETT)
Hi Vishal - just a clarification...

Your Food insecurity is fine based off what I saw in your original post - I was just using that as a "good example" to review.

I believe you're missing additional machine readable info for "overweight" is my current assumption.

Thanks - Mat