Hi Sahil,
I checked the DICOM in detail, and I found some problems here.
First, it seems that the header does not store the bvector
condition. (Are you using an experimental sequence that does not
output bvector information?). A typical SIEMENS diffusion DICOM has
b-vector information stored in (0x0019,0x100E), (0x0019,0x1027), or
the CSA header under "DiffusionGradientDirection". There is no
information stored in either of these three fields.
Second, the "dsi q4 half" usually has 128 sampling directions, but
the dataset only have 48 directions. I cannot match it with my
existing table.
Thirds, SIEMENS diffusion DICOM often save image as mosaic, but
here the data are stored for each slice. DSI Studio can handle this,
but I just curious which sequence you used to acquire the data.
Last but the most important, the signal attenuation for different
b-value does not behave normally. Usually in high b-value acquisition,
there is no signal from gray matter. But in your data, there is still
signals from gray matter at b-value=7,000. It seems that the images
were acquired in about 3,000. This is why I mentioned that the images
look like a simulated/interpolated one.
I am sorry that I cannot come up with a solution here. It seems
that there is something to sort out at the image acquisition side.
Maybe you may provide more information about how the sequence is
setup.
Best regards,
Frank
On Tue, Oct 28, 2014 at 8:01 PM, Fang-Cheng Yeh <
fran...@gmail.com> wrote:
> Thanks a lot for your patience in helping me figure out the problem.
>
> I will let you know once I come up with a solution.
>
> Best regards,
> Frank
>
> On Tue, Oct 28, 2014 at 7:59 PM, Sahil Bajaj <
sahil...@gmail.com> wrote:
>>
>> Hi Frank,
>>
>> Please find the attached folder "DSI_LC.zip', containing the .dcm files
>> which I used to create SRC file:
>>
>> Thank you so much Frank for all your time and great help in figuring out
>> this issue.
>>
>> Thanks.
>> S