After finishing the teaching session I tried to pick one tip from the SBS box using the “Get DiTi” direct command from the Evoware standard menu. Although the cone went in the tip, mounted it and lifted it up without any problems, I got the following error message from Evoware standard.
I have tried to talk to TECAN helpdesk but after a lot of emails bak and forth it seems they do not have an idea and they suggested we order one of their technicians to look into it on site.
I was wondering whether anyone has had a similar experience and whether there is a way to overcome this problem.
Thank you for your time.
All the best,
Panos
--
-- You received this message because you are subscribed to the Google Groups DIYbio group. To post to this group, send email to diy...@googlegroups.com. To unsubscribe from this group, send email to diybio+un...@googlegroups.com. For more options, visit this group at https://groups.google.com/d/forum/diybio?hl=en
Learn more at www.diybio.org
---
You received this message because you are subscribed to the Google Groups "DIYbio" group.
To unsubscribe from this group and stop receiving emails from it, send an email to diybio+un...@googlegroups.com.
To view this discussion on the web visit https://groups.google.com/d/msgid/diybio/70d5fd07-1522-4b16-846b-b8f8685caa15n%40googlegroups.com.
To view this discussion on the web visit https://groups.google.com/d/msgid/diybio/86e4053a-af97-4a99-8fb3-1286156201ecn%40googlegroups.com.
The Z-Start value must be by set without Diti approx. 50 Steps over the Diti Edge and the golden Diti Cone (5mm).
Then Z-Max is then set to a value of 210-250 (maybe even 300) steps (or tenths of mm) from Z-Start. Remember to unlock the Z values from each other before adjusting the values.
BUT THE MOST IMPORTANT DETAIL IS THE FOLLOWING:
When the values of Z-Start and Z-max are set DO NOT MOVE THE Diti CONE TO Z-Max when teaching (by using the per "move to" button). This will definitely lead to a crash message! Instead save the settings and test the set values with the Direct Command "Get DiTis".
We observed that only when we tested the Diti Cone Z-max settings using the "move to" button, all the problems that I described in the beginning of this thread appeared.
--
-- You received this message because you are subscribed to the Google Groups DIYbio group. To post to this group, send email to diy...@googlegroups.com. To unsubscribe from this group, send email to diybio+un...@googlegroups.com. For more options, visit this group at https://groups.google.com/d/forum/diybio?hl=en
Learn more at www.diybio.org
---
You received this message because you are subscribed to the Google Groups "DIYbio" group.
To unsubscribe from this group and stop receiving emails from it, send an email to diybio+un...@googlegroups.com.
To view this discussion on the web visit https://groups.google.com/d/msgid/diybio/d57520c8-6830-448c-a541-5c0bac462422n%40googlegroups.com.