lightbulb / air-flow convection thermocycler for PCR - $70

227 views
Skip to first unread message

Mackenzie Cowell

unread,
Jul 9, 2011, 3:52:35 PM7/9/11
to diybio
even better than the light-bulb PCR reported earlier (http://futurelabcamp.com/2011/2011/nyc-projects#Lightbulb-PCR).  $67 thermocycler (3000 INR).


Design of a thermocycler based on light and air having optimal heat capacity

Pramod Upadhyay

This report describes the design of a ‘fast’ thermocycler for polymerase chain reactions (PCR). In this design of the thermocycler, the sample holders were placed over a copper foil made by electroforming, having an optimal heat capacity, and showing exact fitting of the sample holders. While an electric bulb was used as the source of heat, a muffin fan was used for cooling. A matched copper-wire resistance thermometer was used to monitor the temperature. Conventional poly-propylene tubes or ordinary glass capillaries were used as sample holders for DNA amplification. For exercising control over various timed events necessary for conducting PCR, the apparatus made use of the printer adapter of an IBM personal computer. The ‘fast’ thermocycler was evaluated for its response time, temperature profiles generated, and efficiency in amplifying DNA in the laboratory. Cost of the materials, excluding a PC, of this thermocycler works out around Rs 3000 only.

The polymerase chain reaction (PCR) technique has emerged as one of the most widely used techniques in molecular biology in the recent years1. The increase in the popularity of this technique has been accompanied by corresponding improvements in the apparatus used for performing it. A variety of engineering approaches to control temperature have been investigated for their applicability to the design of a PCR apparatus, including resistive heating and cooling by means of a refrigerant2; temperature cycling, using the Peltier effect3, through the use of heated and chilled air-streams4,5, and most recently, in a continuous flow manner6.

In most of these designs the sample holder is placed inside a temperature-controlled metallic block, often weighing more than 500 g. Assuming that the specific heat of the material of metallic block is similar to that of aluminium, its heat capacity would be around 450 JK–1 which would be rather too high to heat or cool, for example, 10 samples, each of 50 µl, having heat capacity of 2.1 JK–1 only. The high heat capacity of the metallic block often limits the rate at which it could be heated or cooled. In thermocyclers based on heated and chilled air-streams, while the heat capacity of medium is quite small due to the low thermal conductivity of air, an air cyclone would be required to achieve uniformity of temperature.

In this report a thermocycler having a unique design of copper foil that holds sample holders having approximately 10 JK–1 heat capacity is described. In this design, copper foil is heated
by placing it close to a halogen lamp, and subsequently cooling it by blowing a gentle stream of air over it. Owing to the optimal heat capacity of the copper foil, it can be heated and cooled quite rapidly. Furthermore, the high thermal conductivity of copper (unlike air) assures uniformity of temperature. A thermally matched temperature sensor is used to take the full advantage of fast temperature cycling of copper foil.

Features of the thermocycler

In this thermocycler samples can either be placed inside conventional poly-propylene tubes or else placed inside capillary tubes, and then placed on a blackened foil made by electroforming copper over poly-propylene tubes or capillary tubes. In the electroformed copper foil, tight fitting of the sample holder, and uniform and controlled thickness is ensured. A muffin fan blows air of ambient temperature over this copper foil. The temperature of the copper foil is maintained by controlling the light energy emitted by a 500 W halogen lamp. A ‘matched’ temperature transducer is used for monitoring temperature changes and communicating with the controller. The control is built around the parallel printer adapter of an IBM PC. Attaining of fast temperature cycling becomes possible with this thermocycler due to low heat capacity of the copper foil. For the incorporation of the above-outlined features in the design proposed here, the following rationale and overview of advantages were considered.

Optimal heat capacity of platform for sample holder

For the optimal heat capacity of the platform for sample holder, the heat capacity of such a platform should be nearly ten times the heat capacity of the sample holders to ensure heat transfer from the platform to the sample holders without any significant temperature change. A foil made by electroforming copper over poly-propylene tubes or capillary tubes served as a platform for sample holders. Thus by using an electroformed foil, the fitting of poly-propylene tubes or capillary tubes was observed to be exact, resulting thereby in maximum heat transfer. Furthermore, we observed that while a foil made by conventional machining process required highly specialized tools, the foil made by electroforming copper gave uniform deposition of copper with thickness (mass) that could be controlled.

Using light energy for heating

An electric bulb has a very fast response compared to nichrome wire heaters used in most designs. Faster ramp rate with practically zero ‘dead time’ was attained by heating the copper foil by the light of a 500 W halogen lamp. Phase control of AC mains was used to control the light intensity of the bulb, and thus the temperature of the sample holder.

Using an ordinary fan for cooling

An industry standard ‘muffin fan’ of 3 1/8" size and having an air blowing rate of 35 cubic feet per minute (CFM) (the cooling fan used in most electronic equipment) was employed for blowing air over the sample holder for cooling. This obviated the need for a specially designed fan.

Using a thermally matched copper-wire resistance thermometer as a
temperature sensor

A thermally matched temperature sensor was developed. An insulated copper wire was packed inside the capillary and its resistance was used as a measure of temperature. The weight of the copper wire inserted inside the capillary was such that the heat capacity of the inserted wire became equal to the heat capacity of 20 µl (volume of the reaction mix) of the sample. This is a major simplification, since a temperature sensor of any desired heat capacity can be made very easily by this method.

Choosing a PC printer adapter for control

The parallel printer adapter of an IBM PC was used as it is fairly ubiquitous and well standardized. It is well suited for moderate data transfer rates. Furthermore, it required minimal hardware support, and very simple programming.

Using commonly available materials and electronic components

All electronic components used were ‘Commercial’ grade, and could be obtained from a hobby store.

Intelligible circuit and program

A very simple circuit with minimum components was designed. The accompanying program, written in Power
Basic, is a popular version of BASIC. It is a fairly straightforward program with no special algorithms.

Design and fabrication of
thermocycler

Copper foil

Poly-propylene tubes or glass capillaries (outer diameter 1.5 mm and inner diameter 1.0 mm, Top Syringes, India), employed in our laboratory for obtaining blood samples from rabbits by retro-orbital puncture, were glued on an acrylic sheet. Electrically conducting silver paint was applied on the surface on which copper was to be electroformed and was connected to a power supply as the cathode. The anode was made from a sheet of copper. Electroforming was carried out in 20% copper sulphate solution in 5% sulphuric acid at 1 A/dm2. The temperature was maintained at 55° C. As the current efficiency of electrolysis was not 100%, it was therefore desirable to periodically weigh and find out the amount of copper deposited. Approximately 30 g copper was deposited on the surface to make a platform for 16 sample holders. After electroforming, the copper foil was removed from the acrylic sheet by subjecting it to hot and cold water a couple of times.

The assembly

A schematic diagram of the fast temperature cycler is shown in Figure 1. The copper foil made by electroforming was mounted on a frame. One of the depressions was used to place the temperature sensor. The foil was blackened by permanent marker ink. A 500 W halogen lamp (Philips, India) was mounted below at a height of 5 cm, and

516.gif (9369 bytes)

 a muffin fan (Rexnord, India, 12 Volts DC; 0.2 A) was mounted over the foil at a height of 5 cm.

The sensor

In order to make a matched temperature sensor, for example, for a 20 µl sample, first the heat capacity of 20 µl sample was calculated. Assuming that the specific heat and density of the sample were the same as for water (4.18 Jg–1 K–1 and 1 gml–1, respectively) the heat capacity would be:

 

20 µl/1000 µl g–1 ´  4.18 Jg–1 K–1

= 0.0836 JK–1.

 

Given the specific heat of copper as 0.384 Jg–1 K–1, for a length of copper wire to have 0.836 JK–1 heat capacity, it should weigh 0.0836 JK–1/0.384 Jg–1 K–1 = 0.217 g.

Thus the heat capacity of 217 mg of copper wire and 20 µl sample was found to be similar.

This weight (217 mg) of thin insulated copper wire was therefore packed inside a capillary and its ends were sealed with minimum amount of RTV silicone adhesive, such that both the ends of copper wire protruded out of the capillary. These ends were meant to be connected to the circuit board.

In order to calibrate the sensor, it was held at a series of temperature values as

read by a mercury thermometer, and the number presented by the analog to digital converter in response to each value was recorded (see data multiplexer below). A calibration equation was constructed using these numerical values of temperature and response.

In order to compare the temperature–time profiles of the sensor and the sample, another sensor was constructed to record the temperature of the PCR mix. In such a sensor, the electrical conductance of this sealed PCR mix was recorded at different temperatures and a calibration curve was plotted. Both the sensors, the ‘PCR mix’ temperature sensor and the ‘copper wire’ temperature sensor, were placed on the copper foil kept under the lamp.

The controller

Construction of the circuit: All the components were purchased over the counter from entertainment electronics stores. The circuit was wired on a ‘solder breadboard’ and instructions given in ref. 7 were followed. The circuit comprised of 3 parts: the data
multiplexer, the timer and the power supply.

The data multiplexer: The block diagram of the data multiplexer, and its working are shown in (Figure 2). Four bits of the control port were decoded into 16-control lines, and an eight-bit data bus was fed to three transparent octal latches. A 12-bits error signal, calculated from the set point, and temperature of the sensor, was written on a latch. The error signal was transferred from layer 1 by the data transfer pulse and loaded to counters by the load pulse. At the end of the countdown, the carry out ‘fired’ the triac. Two layers of latches were used as a 12-bit word that was to be transferred in one go8.

To read the temperature of the sensor, the resistance of the copper wire was converted into voltage, scaled, and was amplified. This voltage was fed to a sample and hold, and then to a 12-bit analog digital converter (ADC). By using a 2 ´  4 bits tristate, the lower 8 bits of the converted number were read. The higher 4 bits were selected by one of the bits of the latch and were read in a similar manner.

517.gif (20045 bytes) 

 

The timer: The block diagramme of the timer is shown in Figure 2. Since a 12-bits error signal was calculated, every half cycle of the AC mains was divided into 4096 points. AC-mains-synchronous 100 Hz was multiplied by 4096 by the phase-locked loop. The AC-mains-synchronous 100 Hz signal was used to transfer data from latch layer 2 to latch layer 1. The computer was interrupted at 10 Hz, also derived from AC-mains-synchronous 100 Hz, by the fifth bit of the status port.

Power supply: A conventional three-pin-regulator-based power supply was made to power the circuit.

The software

The interface described above was made to run by a program written in ‘Power Basic’. An ‘assembly routine’ acquired data from the ADC. The program calculates the error signal, using a ‘proportional and integral’ algorithm9, and sends it to the respective latches.

518.gif (35714 bytes) 

DNA amplification

For testing the efficiency of the instrument in the amplification of a 307-bp amplicon for the diagnosis of Hepatitis B virus10, fifty microliter of the reaction mixture (3 m l, 25 mM MgCl2; 4 m l, 2.5 mM dNTP; 5 m l, 10´ buffer; and 0.2 m l taq polymerase and primers;

DNA and water) was loaded in polypropylene tubes and was subjected to either of the two protocols: (i) conventional cycling, i.e. 1 min for denaturing DNA, 1 min for annealing, and 2 min for extension, for 30 cycles and, (ii) fast cycling with 10 s denaturing, 30 s annealing, and 30 s extension for 30 cycles.

Results and discussion

Assembly

Space constraints must necessarily limit the details of assembling and operating the thermocycler reported here. It is nevertheless believed that sufficient information is contained in this report to enable the construction of such a thermocycler11.

Evaluation of the temperature sensor

Figure 3 shows a comparison of the temperature–time profiles generated by the PCR-mix sensor and copper-wire sensor. The temperature of the PCR-mix sensor did not show a change as smoothly as that for the copper wire sensor, and the maximum time lag was of the order of 5 s. Thus we conclude that the duration of the time for which the PCR samples were to be maintained at a particular temperature should not be less than 5 s. The maximum rate of temperature change in the temperature–time profiles shown in Figure 3 was of the order of 3° C/s.

Uniformity of the temperature

Two additional temperature sensors were kept at the two extreme corners on the copper foil, and their temperatures recorded. Figure 4 shows the difference between the temperatures recorded by the two sensors as less than 1° C. It was therefore concluded that the apparatus achieved a uniformity of temperature distribution of the order of 1° C across the samples.

DNA amplification

Figure 5 shows the results of the amplification experiment. Lane 1 shows a 100-bp ladder. In lanes 2 and 6, negative controls with conventional cycling and fast cycling, respectively, are shown. Positive controls are shown in lanes 3, 4 and 5. Lane 3 shows amplification with conventional cycling. And in the lanes 4 and 5, amplification with fast cycling, carried out at two different times at different locations, are shown. Comparing these with the yield of the PCR product when conventional timings were used, it is obvious that it is more than sufficient for visualizing the amplified product on the ethidium bromide-stained gel.

Conclusions

A unique design thermocycler to carry out polymerase chain reaction has been presented in this paper. The thermocycler is designed with locally available materials and is evaluated for its uniformity of temperature, temperature ramp rate, and efficiency of amplifications. Due to faster temperature ramp rate, the time required for amplification could be reduced to 45 min, which is one third of conventional time. Cost of the materials, excluding a PC, of this thermocycler works out around Rs 3000 only.

 


  1. Wittwer, C. T., Reed, G. B. and Ririe, K. M., in The Polymerase Chain Reaction (eds Mullis, K. B., Ferre, F. and Gibbs, R. A.), Birkhaüser, Boston, 1994, pp. 174–181.
  2. Upadhyay, P. K. and Gadre, D. V., Meas. Sci. Technol., 1995, 6, 588–
    592.
  3. Collasius, M., Falk, H., Ciester, C. and Valet, G., Anal. Biochem., 1989, 181, 163–166.
  4. Wittwer, C. T., Fillmore, G. C. and Hillyard, D. R., Nucleic Acids Res., 1989, 17, 4353–4357.
  5. Wittwer, C. T., Hillyard, D. R. and Ririe, K. M., US Patent 5455175, 1995.
  6. Martin, U. K., Andrew, J. de Mello and Andreas, M., Science, 1998, 280, 1046–1048.
  7. Horowitz, P. and Hill, W., in The Art of Electronics, Cambridge University Press, Cambridge, 1989, pp. 827–862.
  8. Pramod, U., Ele. Des., 1997, 45, 100–102.
  9. Corripio, A. B., in Tuning of Industrial Control System, Instrument Society of America, North Carolina, 1990, pp. 111–133.
  10. Details of the primers to be published later.
  11. We are planning to hold a training workshop to disseminate the knowhow of this thermocycler, in October 1999. Readers, who are keen on participating in this workshop, should write to the author.

ACKNOWLEDGEMENTS.  During the designing phase, Amit Misra made a number of valuable suggestions, in the process of evaluation of the thermocycler, the contributions from Amol Amin, Sunder Bhist, S. N. Kazim, Salma Wakil, and Saman Habib are gratefully acknowledged. This work was supported by the core grant from Department of Biotechnology, New Delhi, to the
National Institute of Immunology, New Delhi.


Pramod Upadhyay is in the National Institute of Immunology, Aruna Asaf Ali Marg, New Delhi 110 067, India


--
+1.231.313.9062 / m...@diybio.org / @100ideas

Cory Tobin

unread,
Jul 9, 2011, 7:04:43 PM7/9/11
to diy...@googlegroups.com
Based on Fig 3 it looks like they are getting a ramp rate of about
2C/s. That's pretty good!

-cory

Russell Durrett

unread,
Jul 10, 2011, 2:41:41 AM7/10/11
to DIYbio
This design is lightyears ahead of any previous light bulb thermal
cycler. I mean they even electroplated their own copper foil!

On Jul 9, 12:52 pm, Mackenzie Cowell <m...@diybio.org> wrote:
> even better than the light-bulb PCR reported earlier (http://futurelabcamp.com/2011/2011/nyc-projects#Lightbulb-PCR).  $67
> thermocycler (3000 INR).
>
> http://www.ias.ac.in/currsci/aug25/articles18.htm
>
> *Design of a thermocycler based on light and air having optimal heat
> capacity*
>
> **
>
> *Pramod Upadhyay*
> **
>
> *This report describes the design of a ‘fast’ thermocycler for polymerase
> chain reactions (PCR). In this design of the thermocycler, the sample
> holders were placed over a copper foil made by electroforming, having an
> optimal heat capacity, and showing exact fitting of the sample holders.
> While an electric bulb was used as the source of heat, a muffin fan was used
> for cooling. A matched copper-wire resistance thermometer was used to
> monitor the temperature. Conventional poly-propylene tubes or ordinary glass
> capillaries were used as sample holders for DNA amplification. For
> exercising control over various timed events necessary for conducting PCR,
> the apparatus made use of the printer adapter of an IBM personal computer.
> The ‘fast’ thermocycler was evaluated for its response time, temperature
> profiles generated, and efficiency in amplifying DNA in the laboratory. Cost
> of the materials, excluding a PC, of this thermocycler works out around Rs
> 3000 only.*
> advantage of fast temperature cycling of copper foil.**
> **
>
> *Features of the thermocycler*
>
> In this thermocycler samples can either be placed inside conventional
> poly-propylene tubes or else placed inside capillary tubes, and then placed
> on a blackened foil made by electroforming copper over poly-propylene tubes
> or capillary tubes. In the electroformed copper foil, tight fitting of the
> sample holder, and uniform and controlled thickness is ensured. A muffin fan
> blows air of ambient temperature over this copper foil. The temperature of
> the copper foil is maintained by controlling the light energy emitted by a
> 500 W halogen lamp. A ‘matched’ temperature transducer is used for
> monitoring temperature changes and communicating with the controller. The
> control is built around the parallel printer adapter of an IBM PC. Attaining
> of fast temperature cycling becomes possible with this thermocycler due to
> low heat capacity of the copper foil. For the incorporation of the
> above-outlined features in the design proposed here, the following rationale
> and overview of advantages were considered.**
> **
>
> *Optimal heat capacity of platform for sample holder*
>
> For the optimal heat capacity of the platform for sample holder, the heat
> capacity of such a platform should be nearly ten times the heat capacity of
> the sample holders to ensure heat transfer from the platform to the sample
> holders without any significant temperature change. A foil made by
> electroforming copper over poly-propylene tubes or capillary tubes served as
> a platform for sample holders. Thus by using an electroformed foil, the
> fitting of poly-propylene tubes or capillary tubes was observed to be exact,
> resulting thereby in maximum heat transfer. Furthermore, we observed that
> while a foil made by conventional machining process required highly
> specialized tools, the foil made by electroforming copper gave uniform
> deposition of copper with thickness (mass) that could be controlled.**
> **
>
> *Using light energy for heating*
>
> An electric bulb has a very fast response compared to nichrome wire heaters
> used in most designs. Faster ramp rate with practically zero ‘dead time’ was
> attained by heating the copper foil by the light of a 500 W halogen lamp.
> Phase control of AC mains was used to control the light intensity of the
> bulb, and thus the temperature of the sample holder.**
> **
>
> *Using an ordinary fan for cooling*
>
> An industry standard ‘muffin fan’ of 3 1/8" size and having an air blowing
> rate of 35 cubic feet per minute (CFM) (the cooling fan used in most
> electronic equipment) was employed for blowing air over the sample holder
> for cooling. This obviated the need for a specially designed fan.**
> **
>
> *Using a thermally matched copper-wire resistance thermometer as a
> temperature sensor*
>
> A thermally matched temperature sensor was developed. An insulated copper
> wire was packed inside the capillary and its resistance was used as a
> measure of temperature. The weight of the copper wire inserted inside the
> capillary was such that the heat capacity of the inserted wire became equal
> to the heat capacity of 20 µl (volume of the reaction mix) of the sample.
> This is a major simplification, since a temperature sensor of any desired
> heat capacity can be made very easily by this method.**
> **
>
> *Choosing a PC printer adapter for control*
>
> The parallel printer adapter of an IBM PC was used as it is fairly
> ubiquitous and well standardized. It is well suited for moderate data
> transfer rates. Furthermore, it required minimal hardware support, and very
> simple programming.**
> **
>
> *Using commonly available materials and electronic components*
>
> All electronic components used were ‘Commercial’ grade, and could be
> obtained from a hobby store.**
> **
>
> *Intelligible circuit and program*
>
> A very simple circuit with minimum components was designed. The accompanying
> program, written in Power
> Basic, is a popular version of BASIC. It is a fairly straightforward program
> with no special algorithms.**
> **
>
> *Design and fabrication of
> thermocycler***
> **
>
> *Copper foil*
>
> Poly-propylene tubes or glass capillaries (outer diameter 1.5 mm and inner
> diameter 1.0 mm, Top Syringes, India), employed in our laboratory for
> obtaining blood samples from rabbits by retro-orbital puncture, were glued
> on an acrylic sheet. Electrically conducting silver paint was applied on the
> surface on which copper was to be electroformed and was connected to a power
> supply as the cathode. The anode was made from a sheet of copper.
> Electroforming was carried out in 20% copper sulphate solution in 5%
> sulphuric acid at 1 A/dm2. The temperature was maintained at 55° C. As the
> current efficiency of electrolysis was not 100%, it was therefore desirable
> to periodically weigh and find out the amount of copper deposited.
> Approximately 30 g copper was deposited on the surface to make a platform
> for 16 sample holders. After electroforming, the copper foil was removed
> from the acrylic sheet by subjecting it to hot and cold water a couple of
> times.**
> **
>
> *The assembly*
>
> A schematic diagram of the fast temperature cycler is shown in Figure 1. The
> copper foil made by electroforming was mounted on a frame. One of the
> depressions was used to place the temperature sensor. The foil was blackened
> by permanent marker ink. A 500 W halogen lamp (Philips, India) was mounted
> below at a height of 5 cm, and
>
> [image: 516.gif (9369 bytes)]
>
>  a muffin fan (Rexnord, India, 12 Volts DC; 0.2 A) was mounted over the foil
> at a height of 5 cm.**
> ****
>
> *The sensor*
>
> In order to make a matched temperature sensor, for example, for a 20 µl
> sample, first the heat capacity of 20 µl sample was calculated. Assuming
> that the specific heat and density of the sample were the same as for water
> (4.18 Jg–1 K–1 and 1 gml–1, respectively) the heat capacity would be:
>
> 20 µl/1000 µl g–1 ´  4.18 Jg–1 K–1
>
> = 0.0836 JK–1.
>
> Given the specific heat of copper as 0.384 Jg–1 K–1, for a length of copper
> wire to have 0.836 JK–1 heat capacity, it should weigh 0.0836 JK–1/0.384 Jg
> –1 K–1 = 0.217 g.
>
> Thus the heat capacity of 217 mg of copper wire and 20 µl sample was found
> to be similar.
>
> This weight (217 mg) of thin insulated copper wire was therefore packed
> inside a capillary and its ends were sealed with minimum amount of RTV
> silicone adhesive, such that both the ends of copper wire protruded out of
> the capillary. These ends were meant to be connected to the circuit board.
>
> In order to calibrate the sensor, it was held at a series of temperature
> values as
>
> read by a mercury thermometer, and the number presented by the analog to
> digital converter in response to each value was recorded (see *data
> multiplexer* below). A calibration equation was constructed using these
> numerical values of ...
>
> read more »

Mackenzie Cowell

unread,
Jul 10, 2011, 3:02:34 AM7/10/11
to diy...@googlegroups.com
The authors are pcr biohacker badasses.  It was published circa 1999!  Man, I want to track them down.


--
You received this message because you are subscribed to the Google Groups "DIYbio" group.
To post to this group, send email to diy...@googlegroups.com.
To unsubscribe from this group, send email to diybio+un...@googlegroups.com.
For more options, visit this group at http://groups.google.com/group/diybio?hl=en.

Dakota Hamill

unread,
Jul 10, 2011, 12:47:15 PM7/10/11
to diy...@googlegroups.com
When this was first posted I skimmed over the first couple of top lines, not realizing how far it went down.  I just read through most of it, pretty amazing build.  Now I'm all amped to copper coat some tubes.  Just yesterday I finished copper plating some coins and nails, and I have left-over copper acetate ready to use for another batch of copper plating.  I don't have silver paint, but I do have a silver based Arctic Silver CPU heat sink paste, just no PCR tubes to try it on.

That might be the goal for the day, copper plating something not metal

Nathan McCorkle

unread,
Jul 11, 2011, 1:16:59 PM7/11/11
to diy...@googlegroups.com
Well seems that after the article was published, work stopped, see my
query and the response:
--------------------------------------------------------------
Dr. Upadhyay,

I am tracking a paper, "Design of a thermocycler based on light and
air having optimal heat capacity", published here:
http://www.ias.ac.in/currsci/aug25/articles18.htm

I am wondering if you are the same author, and if there has been any
progress on the research. I would like to make a similar device,
trying to use heated and stirred air as a simplification.

If you can provide any more papers/references to help me, I would
greatly appreciate this.

Best Regards,
-Nathan
--------------------------------------------------------------
Dear Mr. Nathan,

Thank you for the mail and also for interest in my work. We stopped
this project many years ago due to many unfortunate reasons.

At the industrial design level, there have been many up gradation in
the thermocycler designs - they are more compact and reliable and
cheap now.
Further, in the Rea Time PCR the quantification of amplified product
can be done while the reaction is proceeding.

It is not possible for me to suggest you an article or two for update,
please do a google search - you will get many papers. Should you have
any question related to work i have published, i will be delighted to
answer.

Thanks and best wishes.

--
Pramod Upadhyay
Scientist
National Institute of Immunology
--------------------------------------------------------------

> --
> You received this message because you are subscribed to the Google Groups
> "DIYbio" group.
> To post to this group, send email to diy...@googlegroups.com.
> To unsubscribe from this group, send email to
> diybio+un...@googlegroups.com.
> For more options, visit this group at
> http://groups.google.com/group/diybio?hl=en.
>

--
Nathan McCorkle
Rochester Institute of Technology
College of Science, Biotechnology/Bioinformatics

Mackenzie Cowell

unread,
Jul 11, 2011, 2:10:30 PM7/11/11
to diy...@googlegroups.com
there have been many up gradation in the thermocycler designs - they are more compact and reliable and cheap now.

lol.  I'd like to know the thermocyclers he's referring to.  Sub $500 thermocyclers only available in India?

Mac
Reply all
Reply to author
Forward