Re: [diffpy-users] Value of Qmax ang Qmaxinst at different temperature

53 views
Skip to first unread message

Simon Billinge

unread,
Nov 19, 2022, 9:49:29 AM11/19/22
to diffpy...@googlegroups.com
Dear Vishal,

It is a bit late but I see that no-one answered you.

For Qmaxinst it is a good idea to select this once for a series of related scans (as in your case) and to not change it.   I usually look at the entire data-range, try and find the highest Q-value (looking at F(Q)) where I don't see any large glitches (sharp rapid changes in the data that are small edge effects on the detector but greatly amplified by the PDF processing) below it.  This is usually somewhere around 28-30 on our experiments with 60-70 keV x-rays and a Perkin Elmer detector but will vary with the setup.  Once this is set, fix it and don't vary it.

For Qmax, you select this based on your scientific question.  If you are looking for small signals, you want less noisy data and set it lower typically, if you want higher resolution and can tolerate more noise, set it higher (If you want high resolution and low noise, you may have to measure longer!).  When you are comparing scans with each other it is always advised to compare them when they have been processed with the same Qmax.

So in your case, you could set Qmax to capture all the signal at high-T.  You will lose resolution on your lower-T data but have lower noise data.  Or you could set it to capture all the signal at low-T.  You will have more noisy data at high-T (the noise will be the same but your signal is less so your signal-noise is worse for the high-T data).  But in this case you won't be throwing signal away at low-T.     You could also set Qmax differently for each temperature to get the "optimal" signal at each temperature, but then be careful about comparing features across different temperature scans.   Of course, you can do both or all three and use the differently processed data to answer different questions.

Summary:  Always decide what scientific question you are trying to answer and design the analysis guided by that.

S

On Thu, Sep 29, 2022 at 1:55 PM 'VISHAL KUMAR' via diffpy-users <diffpy...@googlegroups.com> wrote:
Hello Everyone,

 I am doing temperature-dependent PDF analysis [PDFGETX3]. So I am facing the problem to decide the exact value of Qmax & Qmaxinst because at high temperatures the meaningful intensity is up to a lower 2theta value like 23 deg but at lower temperatures, this value is increased up to 28 deg. So, how to decide the value of Qmax & Qmaxinst, Can I use different Qmax & Qmaxinst for different temperatures? 


THANKS
VISHAL

--
You received this message because you are subscribed to the Google Groups "diffpy-users" group.
To unsubscribe from this group and stop receiving emails from it, send an email to diffpy-users...@googlegroups.com.
To view this discussion on the web visit https://groups.google.com/d/msgid/diffpy-users/c3c15f84-2f9c-4b40-91d2-adcb9d481a5dn%40googlegroups.com.


--
Simon Billinge
Professor, Columbia University
Physicist, Brookhaven National Laboratory

VISHAL KUMAR

unread,
Nov 22, 2022, 12:59:29 AM11/22/22
to diffpy-users
Hi Prof Billinge,

Thank you for the response. It will really helpful for PDF data analysis. 