It is definitely true that the PDF will give more accurate sizes for
smaller particles, but where the cutoff occurs depends on the
resolution of your measurement. It also depends on your structural
model being correct.
1) instrument resolution issue:
Peaks in the PDF are attenuated by particle size effects and by
instrument resolution effects. The recommended approach is to have a
bulk material that has the same structure as your nanoparticle (as
near as possible). Measure it at roughly the same time, and with the
diffractometer in the same setup, as when you measure your NPs. User
PDFgui to refine the bulk data. Turn off spdiameter so there is no
particle size effect in there and refine sigma_Q which is a parameter
that is accounting for the instrument resolution. Then fix sigma and
refine your nanoparticle data, letting spdiameter refine. You should
get a good estimate of the NP size. However, if your instrument
resolution is very broad and your NPs are very big, then even this
approach runs into problems of loss of accuracy.
2) accuracy of the model issue:
If the structural model you are using is incorrect, then all bets are
off. The program will use spdiameter to damp out ripples that it
doesn't like at high-r and you will get a hugely incorrect NP size.
You can check for this by looking at a plot of the data-PDF with the
model on top. If at high-r there are clear ripples in the data but
not in the model you need to beware. These could be inter-particle
correlation peaks (and therefore OK) or they could be that your
structural model is not perfect and spdiameter has damped the model.
How do you know if the ripples are particle-particle correlations?
You don't! But if the particles are orientationally or spatially
disordered any inter-particle correlations are broad, much broader
than intra-particle correlations.
Hope it helps,
Simon
--
Prof. Simon Billinge
Applied Physics & Applied Mathematics
Columbia University
500 West 120th Street
Room 200 Mudd, MC 4701
New York, NY 10027
Tel:
(212)-854-2918 (o) 851-7428 (lab)
Condensed Matter and Materials Science
Brookhaven National Laboratory
P.O. Box 5000
Upton, NY 11973-5000
(631)-344-5387
email: sb2896 at columbia dot edu
home:
http://nirt.pa.msu.edu/