different values assigned to same Dicom TAGS for same study but two different versions of dcm4chee 2.13.6 and 2.018.1

19 views
Skip to first unread message

gian78567

unread,
Feb 19, 2016, 9:28:04 AM2/19/16
to dcm...@googlegroups.com
Hi all,

I have  noticed an issue that doesn't allow me to use the  new version of dcm4chee (2.018.1) I have just installed (B)
So far we have been using version 2.13.6(A) without having the issue described  below.
The modality in question is a CT.(CT) and the STUDY we will talk about it is the same for both cases 1 and 2

1-When CT sends the study to A I can then forward the study to other destination systems (called tps or Treatment Planning systems) where among other tags,(0020,0032),(0020,0037),(0020,1041) seem to be very important to maintain what for the TPS is also known as  the ZERO slice.
From CT to A everything works  and attached you find the dicom dump for the study  archived on this version of dcm4chee.(2.13.6)

2- When CT sends to B and then I forward this same study to the same destination mentioned above , this latter doesn't recognize the right zero slice.

Looking at the two attached dicom dumps, one for 2.13.6 and the other for .2.0.18.1, I can see the different  values  in some tags, those mentioned before in this email. I also attached two (the same) dicom objects I could download from both dcm4chee.

I'd like to use the newer version of dcm4chee, how do I correct this problem though

thanks for your help.

Gian


2.0.18.1.txt
2.13.6.txt
0270B166file 2.13.6
file 2.018.1