problems with Dicom transfer to GE EchoPAC software

1,867 views
Skip to first unread message

Butch Tyrone

unread,
Nov 16, 2017, 1:59:27 PM11/16/17
to dcm4che
Good afternoon,

I have inherited a support role for DCM4CHEE so I am learning this system as I type.

Our DCM4CHEE server specs, 
VM with Zenon 2.67 GHZ 
6.0 GB RAM
HDD 50GB
Storage HD 1.74TB with 315GB free
OS is Windows Server 2008 Enterprise 64 bit

DCM4CHEE version 3.0.1.2
DCM4CHEE MSSQL 2.12.1
Jboss 4.2.3.GA

GE EchoPAC PC is a HP Z240 workstation with Windows 7 X64 and a 500GB NVME drive. 
All network connections are Gigabit.

For reasons I cannot disclose, we will be unable to update any of these versions without extensive validation testing.

We have a GE Vivid E95 Ultrasound machine. to go with this machine, we have purchased GE EchoPAC software. The system is used as follows.

Technicians take ultrasound images and movies using the GE Vivid E95 Ultrasound machine. These images and movies are stored locally, then transferred to DCM4CHEE where they remain.

The GE EchoPAC PC has a listener configured to look for images and movies previously transferred from the GE Vivid E95 Ultrasound to DCM4CHEE. once every month or so, a Scientist travels here and downloads the images and videos on a study by study basis from DCM4CHEE to the GE EchoPAC PC using the Listener on the EchoPAC PC. Most of them transfer just fine, the scientist can then add his notes and measurements. 

The problem is, there are entire studies that, when selected and a download is attempted, the GE EchoPAC PC will transfer the patient information, then display an error that no images are available. These images can be viewed in the Weasis viewer. They can also be viewed from the GE EchoPAC software, so images and videos are there, but will not transfer down to the GE EchoPAC PC. This could be larger videos causing this problem. 

I have a case opened with GE, they looked at their logs and told me they felt that the DCM4CHEE server was blocking the traffic. They also admitted that the version of EchoPAC software we are using is known to have transfer problems, though they felt that it would affect all patients and studies, not just some of them. I would guess maybe 1/4 of the images uploaded from early 2016 to the present, scattered throughout that time are having this problem. We must get these images downloaded as the scientist can not do his analysis when viewing them from the DCM4CHEE server. 

I have our server admins looking into this issue as well. 

Any help would be appreciated. Let me know if there is any more information I can provide that would be helpful.



 


fleetwoodfc

unread,
Nov 17, 2017, 9:53:44 AM11/17/17
to dcm4che
First thing to look at is the dcm4chee server.log as this will usually show what the issue is - look for ERROR entries related to the transfer.
My suspicion would be that the transfer times out if you are sending large files - search this forum for 'DIMSETimeout' for further info. if this is the case.
p.s. there is no version 2.12.1 (probably is 2.17.1). 

Butch Tyrone

unread,
Nov 21, 2017, 4:37:32 PM11/21/17
to dcm...@googlegroups.com
Thanks, I'll look into this. We are also starting to suspect these are studies with larger or more images that will not transfer.


Butch Tyrone

unread,
Dec 29, 2017, 3:04:22 PM12/29/17
to dcm...@googlegroups.com
So It has been a while since I have been able to get i and take another look. 

I made a failed transfer, then copied the log entries for that activity. I will attach them below. I see no errors in any of the lince, only info and warn. the warns appear in successful transfers as well, so I dont think they are the problem. I see nothing in the log that mentions the DIMSETimeout. 

There is a lot of info to absorb, so I am looking.

Just in case, I looked up the DSIMETimeout and it is set at 60000 ms.

Any help would be appreciated.

2017-12-29 11:39:53,781 INFO  -> (TCPServer-1-2) [org.dcm4cheri.server.ServerImpl] handle - Socket[addr=/10.10.186.29,port=56352,localport=11112]
2017-12-29 11:39:53,781 INFO  -> (TCPServer-1-2) [org.dcm4cheri.net.FsmImpl] Socket[addr=/10.10.186.29,port=56352,localport=11112]
2017-12-29 11:39:53,781 INFO  MPC02090->DCM4CHEE (TCPServer-1-2) [org.dcm4cheri.net.FsmImpl] received AAssociateRQ
appCtxName: 1.2.840.10008.3.1.1.1/DICOM Application Context Name
implClass: 1.2.840.113619.6.118
implVersion: ECHOPAC_201
calledAET: DCM4CHEE
callingAET: MPC02090
maxPDULen: 32768
asyncOpsWindow:
pc-1: as=1.2.840.10008.5.1.4.1.2.2.1/Study Root Query/Retrieve Information Model - FIND
ts=1.2.840.10008.1.2.1/Explicit VR Little Endian
ts=1.2.840.10008.1.2/Implicit VR Little Endian
ts=1.2.840.10008.1.2.2/Explicit VR Big Endian
pc-3: as=1.2.840.10008.5.1.4.1.2.2.2/Study Root Query/Retrieve Information Model - MOVE
ts=1.2.840.10008.1.2.1/Explicit VR Little Endian
ts=1.2.840.10008.1.2/Implicit VR Little Endian
ts=1.2.840.10008.1.2.2/Explicit VR Big Endian
2017-12-29 11:39:53,781 INFO  MPC02090->DCM4CHEE (TCPServer-1-2) [org.dcm4cheri.net.FsmImpl] sending AAssociateAC
appCtxName: 1.2.840.10008.3.1.1.1/DICOM Application Context Name
implClass: 1.2.40.0.13.1.1.1
implVersion: dcm4che-1.4.34
calledAET: DCM4CHEE
callingAET: MPC02090
maxPDULen: 16352
asyncOpsWindow:
pc-1: 0 - acceptance
ts=1.2.840.10008.1.2/Implicit VR Little Endian
pc-3: 0 - acceptance
ts=1.2.840.10008.1.2/Implicit VR Little Endian
2017-12-29 11:39:53,781 INFO  MPC02090->DCM4CHEE (TCPServer-1-2) [org.dcm4cheri.net.FsmImpl] received [pc-1] 4:C_FIND_RQ with Dataset
class: 1.2.840.10008.5.1.4.1.2.2.1/Study Root Query/Retrieve Information Model - FIND
2017-12-29 11:39:53,781 WARN  MPC02090->DCM4CHEE (TCPServer-1-2) [org.dcm4chex.archive.ejb.jdbc.BaseCmd] (0008,1070) Operator's Name,PN,*0,#0,[] not supported for existence and/or matching
2017-12-29 11:39:53,781 WARN  MPC02090->DCM4CHEE (TCPServer-1-2) [org.dcm4chex.archive.ejb.jdbc.BaseCmd] (0018,1020) Software Version(s),LO,*0,#0,[] not supported for existence and/or matching
2017-12-29 11:39:53,781 WARN  MPC02090->DCM4CHEE (TCPServer-1-2) [org.dcm4chex.archive.ejb.jdbc.BaseCmd] (0018,1030) Protocol Name,LO,*0,#0,[] not supported for existence and/or matching
2017-12-29 11:39:53,796 INFO  MPC02090->DCM4CHEE (TCPServer-1-2) [org.dcm4cheri.net.FsmImpl] sending [pc-1] 4:C_FIND_RSP with Dataset
class: 1.2.840.10008.5.1.4.1.2.2.1/Study Root Query/Retrieve Information Model - FIND
status: ff01
2017-12-29 11:39:53,796 INFO  MPC02090->DCM4CHEE (TCPServer-1-2) [org.dcm4cheri.net.FsmImpl] sending [pc-1] 4:C_FIND_RSP
class: 1.2.840.10008.5.1.4.1.2.2.1/Study Root Query/Retrieve Information Model - FIND
status: 0
2017-12-29 11:39:53,796 INFO  MPC02090->DCM4CHEE (ActiveAssoc-1399-1) [org.dcm4cheri.net.FsmImpl] received A-RELEASE-RQ
2017-12-29 11:39:53,796 INFO  MPC02090->DCM4CHEE (ActiveAssoc-1399-1) [org.dcm4cheri.net.FsmImpl] sending A-RELEASE-RP
2017-12-29 11:39:53,796 INFO  -> (TCPServer-1-2) [org.dcm4cheri.server.ServerImpl] finished - Socket[addr=/10.10.186.29,port=56352,localport=11112]
2017-12-29 11:39:53,812 INFO  -> (TCPServer-1-4) [org.dcm4cheri.server.ServerImpl] handle - Socket[addr=/10.10.186.29,port=56353,localport=11112]
2017-12-29 11:39:53,812 INFO  -> (TCPServer-1-4) [org.dcm4cheri.net.FsmImpl] Socket[addr=/10.10.186.29,port=56353,localport=11112]
2017-12-29 11:39:53,812 INFO  MPC02090->DCM4CHEE (TCPServer-1-4) [org.dcm4cheri.net.FsmImpl] received AAssociateRQ
appCtxName: 1.2.840.10008.3.1.1.1/DICOM Application Context Name
implClass: 1.2.840.113619.6.118
implVersion: ECHOPAC_201
calledAET: DCM4CHEE
callingAET: MPC02090
maxPDULen: 32768
asyncOpsWindow:
pc-1: as=1.2.840.10008.5.1.4.1.2.2.1/Study Root Query/Retrieve Information Model - FIND
ts=1.2.840.10008.1.2.1/Explicit VR Little Endian
ts=1.2.840.10008.1.2/Implicit VR Little Endian
ts=1.2.840.10008.1.2.2/Explicit VR Big Endian
pc-3: as=1.2.840.10008.5.1.4.1.2.2.2/Study Root Query/Retrieve Information Model - MOVE
ts=1.2.840.10008.1.2.1/Explicit VR Little Endian
ts=1.2.840.10008.1.2/Implicit VR Little Endian
ts=1.2.840.10008.1.2.2/Explicit VR Big Endian
2017-12-29 11:39:53,812 INFO  MPC02090->DCM4CHEE (TCPServer-1-4) [org.dcm4cheri.net.FsmImpl] sending AAssociateAC
appCtxName: 1.2.840.10008.3.1.1.1/DICOM Application Context Name
implClass: 1.2.40.0.13.1.1.1
implVersion: dcm4che-1.4.34
calledAET: DCM4CHEE
callingAET: MPC02090
maxPDULen: 16352
asyncOpsWindow:
pc-1: 0 - acceptance
ts=1.2.840.10008.1.2/Implicit VR Little Endian
pc-3: 0 - acceptance
ts=1.2.840.10008.1.2/Implicit VR Little Endian
2017-12-29 11:39:53,812 INFO  MPC02090->DCM4CHEE (TCPServer-1-4) [org.dcm4cheri.net.FsmImpl] received [pc-1] 5:C_FIND_RQ with Dataset
class: 1.2.840.10008.5.1.4.1.2.2.1/Study Root Query/Retrieve Information Model - FIND
2017-12-29 11:39:53,812 WARN  MPC02090->DCM4CHEE (TCPServer-1-4) [org.dcm4chex.archive.ejb.jdbc.BaseCmd] (0018,0010) Contrast/Bolus Agent,LO,*0,#0,[] not supported for existence and/or matching
2017-12-29 11:39:53,859 INFO  MPC02090->DCM4CHEE (ActiveAssoc-1399-1) [org.dcm4cheri.net.FsmImpl] closing connection - Socket[addr=/10.10.186.29,port=56352,localport=11112]
2017-12-29 11:39:53,859 INFO  MPC02090->DCM4CHEE (TCPServer-1-4) [org.dcm4cheri.net.FsmImpl] sending [pc-1] 5:C_FIND_RSP with Dataset
class: 1.2.840.10008.5.1.4.1.2.2.1/Study Root Query/Retrieve Information Model - FIND
status: ff01
2017-12-29 11:39:53,859 INFO  MPC02090->DCM4CHEE (TCPServer-1-4) [org.dcm4cheri.net.FsmImpl] sending [pc-1] 5:C_FIND_RSP with Dataset
class: 1.2.840.10008.5.1.4.1.2.2.1/Study Root Query/Retrieve Information Model - FIND
status: ff01
2017-12-29 11:39:53,859 INFO  MPC02090->DCM4CHEE (TCPServer-1-4) [org.dcm4cheri.net.FsmImpl] sending [pc-1] 5:C_FIND_RSP with Dataset
class: 1.2.840.10008.5.1.4.1.2.2.1/Study Root Query/Retrieve Information Model - FIND
status: ff01
2017-12-29 11:39:53,859 INFO  MPC02090->DCM4CHEE (TCPServer-1-4) [org.dcm4cheri.net.FsmImpl] sending [pc-1] 5:C_FIND_RSP with Dataset
class: 1.2.840.10008.5.1.4.1.2.2.1/Study Root Query/Retrieve Information Model - FIND
status: ff01
2017-12-29 11:39:53,859 INFO  MPC02090->DCM4CHEE (TCPServer-1-4) [org.dcm4cheri.net.FsmImpl] sending [pc-1] 5:C_FIND_RSP with Dataset
class: 1.2.840.10008.5.1.4.1.2.2.1/Study Root Query/Retrieve Information Model - FIND
status: ff01
2017-12-29 11:39:53,859 INFO  MPC02090->DCM4CHEE (TCPServer-1-4) [org.dcm4cheri.net.FsmImpl] sending [pc-1] 5:C_FIND_RSP with Dataset
class: 1.2.840.10008.5.1.4.1.2.2.1/Study Root Query/Retrieve Information Model - FIND
status: ff01
2017-12-29 11:39:53,859 INFO  MPC02090->DCM4CHEE (TCPServer-1-4) [org.dcm4cheri.net.FsmImpl] sending [pc-1] 5:C_FIND_RSP with Dataset
class: 1.2.840.10008.5.1.4.1.2.2.1/Study Root Query/Retrieve Information Model - FIND
status: ff01
2017-12-29 11:39:53,859 INFO  MPC02090->DCM4CHEE (TCPServer-1-4) [org.dcm4cheri.net.FsmImpl] sending [pc-1] 5:C_FIND_RSP with Dataset
class: 1.2.840.10008.5.1.4.1.2.2.1/Study Root Query/Retrieve Information Model - FIND
status: ff01
2017-12-29 11:39:53,859 INFO  MPC02090->DCM4CHEE (TCPServer-1-4) [org.dcm4cheri.net.FsmImpl] sending [pc-1] 5:C_FIND_RSP with Dataset
class: 1.2.840.10008.5.1.4.1.2.2.1/Study Root Query/Retrieve Information Model - FIND
status: ff01
2017-12-29 11:39:53,859 INFO  MPC02090->DCM4CHEE (TCPServer-1-4) [org.dcm4cheri.net.FsmImpl] sending [pc-1] 5:C_FIND_RSP with Dataset
class: 1.2.840.10008.5.1.4.1.2.2.1/Study Root Query/Retrieve Information Model - FIND
status: ff01
2017-12-29 11:39:53,859 INFO  MPC02090->DCM4CHEE (TCPServer-1-4) [org.dcm4cheri.net.FsmImpl] sending [pc-1] 5:C_FIND_RSP with Dataset
class: 1.2.840.10008.5.1.4.1.2.2.1/Study Root Query/Retrieve Information Model - FIND
status: ff01
2017-12-29 11:39:53,859 INFO  MPC02090->DCM4CHEE (TCPServer-1-4) [org.dcm4cheri.net.FsmImpl] sending [pc-1] 5:C_FIND_RSP with Dataset
class: 1.2.840.10008.5.1.4.1.2.2.1/Study Root Query/Retrieve Information Model - FIND
status: ff01
2017-12-29 11:39:53,859 INFO  MPC02090->DCM4CHEE (TCPServer-1-4) [org.dcm4cheri.net.FsmImpl] sending [pc-1] 5:C_FIND_RSP with Dataset
class: 1.2.840.10008.5.1.4.1.2.2.1/Study Root Query/Retrieve Information Model - FIND
status: ff01
2017-12-29 11:39:53,859 INFO  MPC02090->DCM4CHEE (TCPServer-1-4) [org.dcm4cheri.net.FsmImpl] sending [pc-1] 5:C_FIND_RSP with Dataset
class: 1.2.840.10008.5.1.4.1.2.2.1/Study Root Query/Retrieve Information Model - FIND
status: ff01
2017-12-29 11:39:53,859 INFO  MPC02090->DCM4CHEE (TCPServer-1-4) [org.dcm4cheri.net.FsmImpl] sending [pc-1] 5:C_FIND_RSP with Dataset
class: 1.2.840.10008.5.1.4.1.2.2.1/Study Root Query/Retrieve Information Model - FIND
status: ff01
2017-12-29 11:39:53,859 INFO  MPC02090->DCM4CHEE (TCPServer-1-4) [org.dcm4cheri.net.FsmImpl] sending [pc-1] 5:C_FIND_RSP with Dataset
class: 1.2.840.10008.5.1.4.1.2.2.1/Study Root Query/Retrieve Information Model - FIND
status: ff01
2017-12-29 11:39:53,859 INFO  MPC02090->DCM4CHEE (TCPServer-1-4) [org.dcm4cheri.net.FsmImpl] sending [pc-1] 5:C_FIND_RSP with Dataset
class: 1.2.840.10008.5.1.4.1.2.2.1/Study Root Query/Retrieve Information Model - FIND
status: ff01
2017-12-29 11:39:53,859 INFO  MPC02090->DCM4CHEE (TCPServer-1-4) [org.dcm4cheri.net.FsmImpl] sending [pc-1] 5:C_FIND_RSP with Dataset
class: 1.2.840.10008.5.1.4.1.2.2.1/Study Root Query/Retrieve Information Model - FIND
status: ff01
2017-12-29 11:39:53,859 INFO  MPC02090->DCM4CHEE (TCPServer-1-4) [org.dcm4cheri.net.FsmImpl] sending [pc-1] 5:C_FIND_RSP with Dataset
class: 1.2.840.10008.5.1.4.1.2.2.1/Study Root Query/Retrieve Information Model - FIND
status: ff01
2017-12-29 11:39:53,859 INFO  MPC02090->DCM4CHEE (TCPServer-1-4) [org.dcm4cheri.net.FsmImpl] sending [pc-1] 5:C_FIND_RSP with Dataset
class: 1.2.840.10008.5.1.4.1.2.2.1/Study Root Query/Retrieve Information Model - FIND
status: ff01
2017-12-29 11:39:53,859 INFO  MPC02090->DCM4CHEE (TCPServer-1-4) [org.dcm4cheri.net.FsmImpl] sending [pc-1] 5:C_FIND_RSP with Dataset
class: 1.2.840.10008.5.1.4.1.2.2.1/Study Root Query/Retrieve Information Model - FIND
status: ff01
2017-12-29 11:39:53,859 INFO  MPC02090->DCM4CHEE (TCPServer-1-4) [org.dcm4cheri.net.FsmImpl] sending [pc-1] 5:C_FIND_RSP with Dataset
class: 1.2.840.10008.5.1.4.1.2.2.1/Study Root Query/Retrieve Information Model - FIND
status: ff01
2017-12-29 11:39:53,859 INFO  MPC02090->DCM4CHEE (TCPServer-1-4) [org.dcm4cheri.net.FsmImpl] sending [pc-1] 5:C_FIND_RSP with Dataset
class: 1.2.840.10008.5.1.4.1.2.2.1/Study Root Query/Retrieve Information Model - FIND
status: ff01
2017-12-29 11:39:53,859 INFO  MPC02090->DCM4CHEE (TCPServer-1-4) [org.dcm4cheri.net.FsmImpl] sending [pc-1] 5:C_FIND_RSP with Dataset
class: 1.2.840.10008.5.1.4.1.2.2.1/Study Root Query/Retrieve Information Model - FIND
status: ff01
2017-12-29 11:39:53,859 INFO  MPC02090->DCM4CHEE (TCPServer-1-4) [org.dcm4cheri.net.FsmImpl] sending [pc-1] 5:C_FIND_RSP with Dataset
class: 1.2.840.10008.5.1.4.1.2.2.1/Study Root Query/Retrieve Information Model - FIND
status: ff01
2017-12-29 11:39:53,859 INFO  MPC02090->DCM4CHEE (TCPServer-1-4) [org.dcm4cheri.net.FsmImpl] sending [pc-1] 5:C_FIND_RSP with Dataset
class: 1.2.840.10008.5.1.4.1.2.2.1/Study Root Query/Retrieve Information Model - FIND
status: ff01
2017-12-29 11:39:53,859 INFO  MPC02090->DCM4CHEE (TCPServer-1-4) [org.dcm4cheri.net.FsmImpl] sending [pc-1] 5:C_FIND_RSP with Dataset
class: 1.2.840.10008.5.1.4.1.2.2.1/Study Root Query/Retrieve Information Model - FIND
status: ff01
2017-12-29 11:39:53,859 INFO  MPC02090->DCM4CHEE (TCPServer-1-4) [org.dcm4cheri.net.FsmImpl] sending [pc-1] 5:C_FIND_RSP with Dataset
class: 1.2.840.10008.5.1.4.1.2.2.1/Study Root Query/Retrieve Information Model - FIND
status: ff01
2017-12-29 11:39:53,859 INFO  MPC02090->DCM4CHEE (TCPServer-1-4) [org.dcm4cheri.net.FsmImpl] sending [pc-1] 5:C_FIND_RSP with Dataset
class: 1.2.840.10008.5.1.4.1.2.2.1/Study Root Query/Retrieve Information Model - FIND
status: ff01
2017-12-29 11:39:53,859 INFO  MPC02090->DCM4CHEE (TCPServer-1-4) [org.dcm4cheri.net.FsmImpl] sending [pc-1] 5:C_FIND_RSP with Dataset
class: 1.2.840.10008.5.1.4.1.2.2.1/Study Root Query/Retrieve Information Model - FIND
status: ff01
2017-12-29 11:39:53,859 INFO  MPC02090->DCM4CHEE (TCPServer-1-4) [org.dcm4cheri.net.FsmImpl] sending [pc-1] 5:C_FIND_RSP with Dataset
class: 1.2.840.10008.5.1.4.1.2.2.1/Study Root Query/Retrieve Information Model - FIND
status: ff01
2017-12-29 11:39:53,859 INFO  MPC02090->DCM4CHEE (TCPServer-1-4) [org.dcm4cheri.net.FsmImpl] sending [pc-1] 5:C_FIND_RSP with Dataset
class: 1.2.840.10008.5.1.4.1.2.2.1/Study Root Query/Retrieve Information Model - FIND
status: ff01
2017-12-29 11:39:53,859 INFO  MPC02090->DCM4CHEE (TCPServer-1-4) [org.dcm4cheri.net.FsmImpl] sending [pc-1] 5:C_FIND_RSP with Dataset
class: 1.2.840.10008.5.1.4.1.2.2.1/Study Root Query/Retrieve Information Model - FIND
status: ff01
2017-12-29 11:39:53,859 INFO  MPC02090->DCM4CHEE (TCPServer-1-4) [org.dcm4cheri.net.FsmImpl] sending [pc-1] 5:C_FIND_RSP with Dataset
class: 1.2.840.10008.5.1.4.1.2.2.1/Study Root Query/Retrieve Information Model - FIND
status: ff01
2017-12-29 11:39:53,859 INFO  MPC02090->DCM4CHEE (TCPServer-1-4) [org.dcm4cheri.net.FsmImpl] sending [pc-1] 5:C_FIND_RSP with Dataset
class: 1.2.840.10008.5.1.4.1.2.2.1/Study Root Query/Retrieve Information Model - FIND
status: ff01
2017-12-29 11:39:53,859 INFO  MPC02090->DCM4CHEE (TCPServer-1-4) [org.dcm4cheri.net.FsmImpl] sending [pc-1] 5:C_FIND_RSP
class: 1.2.840.10008.5.1.4.1.2.2.1/Study Root Query/Retrieve Information Model - FIND
status: 0
2017-12-29 11:39:53,906 INFO  MPC02090->DCM4CHEE (ActiveAssoc-1400-1) [org.dcm4cheri.net.FsmImpl] received A-RELEASE-RQ
2017-12-29 11:39:53,906 INFO  MPC02090->DCM4CHEE (ActiveAssoc-1400-1) [org.dcm4cheri.net.FsmImpl] sending A-RELEASE-RP
2017-12-29 11:39:53,906 INFO  -> (TCPServer-1-4) [org.dcm4cheri.server.ServerImpl] finished - Socket[addr=/10.10.186.29,port=56353,localport=11112]
2017-12-29 11:39:53,968 INFO  MPC02090->DCM4CHEE (ActiveAssoc-1400-1) [org.dcm4cheri.net.FsmImpl] closing connection - Socket[addr=/10.10.186.29,port=56353,localport=11112]
2017-12-29 11:39:54,062 INFO  -> (TCPServer-1-3) [org.dcm4cheri.server.ServerImpl] handle - Socket[addr=/10.10.186.29,port=56354,localport=11112]
2017-12-29 11:39:54,062 INFO  -> (TCPServer-1-3) [org.dcm4cheri.net.FsmImpl] Socket[addr=/10.10.186.29,port=56354,localport=11112]
2017-12-29 11:39:54,062 INFO  MPC02090->DCM4CHEE (TCPServer-1-3) [org.dcm4cheri.net.FsmImpl] received AAssociateRQ
appCtxName: 1.2.840.10008.3.1.1.1/DICOM Application Context Name
implClass: 1.2.840.113619.6.118
implVersion: ECHOPAC_201
calledAET: DCM4CHEE
callingAET: MPC02090
maxPDULen: 32768
asyncOpsWindow:
pc-1: as=1.2.840.10008.5.1.4.1.2.2.1/Study Root Query/Retrieve Information Model - FIND
ts=1.2.840.10008.1.2.1/Explicit VR Little Endian
ts=1.2.840.10008.1.2/Implicit VR Little Endian
ts=1.2.840.10008.1.2.2/Explicit VR Big Endian
pc-3: as=1.2.840.10008.5.1.4.1.2.2.2/Study Root Query/Retrieve Information Model - MOVE
ts=1.2.840.10008.1.2.1/Explicit VR Little Endian
ts=1.2.840.10008.1.2/Implicit VR Little Endian
ts=1.2.840.10008.1.2.2/Explicit VR Big Endian
2017-12-29 11:39:54,062 INFO  MPC02090->DCM4CHEE (TCPServer-1-3) [org.dcm4cheri.net.FsmImpl] sending AAssociateAC
appCtxName: 1.2.840.10008.3.1.1.1/DICOM Application Context Name
implClass: 1.2.40.0.13.1.1.1
implVersion: dcm4che-1.4.34
calledAET: DCM4CHEE
callingAET: MPC02090
maxPDULen: 16352
asyncOpsWindow:
pc-1: 0 - acceptance
ts=1.2.840.10008.1.2/Implicit VR Little Endian
pc-3: 0 - acceptance
ts=1.2.840.10008.1.2/Implicit VR Little Endian
2017-12-29 11:39:54,062 INFO  MPC02090->DCM4CHEE (TCPServer-1-3) [org.dcm4cheri.net.FsmImpl] received [pc-1] 6:C_FIND_RQ with Dataset
class: 1.2.840.10008.5.1.4.1.2.2.1/Study Root Query/Retrieve Information Model - FIND
2017-12-29 11:39:54,062 WARN  MPC02090->DCM4CHEE (TCPServer-1-3) [org.dcm4chex.archive.ejb.jdbc.BaseCmd] (0008,1070) Operator's Name,PN,*0,#0,[] not supported for existence and/or matching
2017-12-29 11:39:54,062 WARN  MPC02090->DCM4CHEE (TCPServer-1-3) [org.dcm4chex.archive.ejb.jdbc.BaseCmd] (0018,1020) Software Version(s),LO,*0,#0,[] not supported for existence and/or matching
2017-12-29 11:39:54,062 WARN  MPC02090->DCM4CHEE (TCPServer-1-3) [org.dcm4chex.archive.ejb.jdbc.BaseCmd] (0018,1030) Protocol Name,LO,*0,#0,[] not supported for existence and/or matching
2017-12-29 11:39:54,062 INFO  MPC02090->DCM4CHEE (TCPServer-1-3) [org.dcm4cheri.net.FsmImpl] sending [pc-1] 6:C_FIND_RSP with Dataset
class: 1.2.840.10008.5.1.4.1.2.2.1/Study Root Query/Retrieve Information Model - FIND
status: ff01
2017-12-29 11:39:54,062 INFO  MPC02090->DCM4CHEE (TCPServer-1-3) [org.dcm4cheri.net.FsmImpl] sending [pc-1] 6:C_FIND_RSP
class: 1.2.840.10008.5.1.4.1.2.2.1/Study Root Query/Retrieve Information Model - FIND
status: 0
2017-12-29 11:39:54,062 INFO  MPC02090->DCM4CHEE (ActiveAssoc-1401-1) [org.dcm4cheri.net.FsmImpl] received A-RELEASE-RQ
2017-12-29 11:39:54,062 INFO  MPC02090->DCM4CHEE (ActiveAssoc-1401-1) [org.dcm4cheri.net.FsmImpl] sending A-RELEASE-RP
2017-12-29 11:39:54,062 INFO  -> (TCPServer-1-3) [org.dcm4cheri.server.ServerImpl] finished - Socket[addr=/10.10.186.29,port=56354,localport=11112]
2017-12-29 11:39:54,077 INFO  -> (TCPServer-1) [org.dcm4cheri.server.ServerImpl] handle - Socket[addr=/10.10.186.29,port=56355,localport=11112]
2017-12-29 11:39:54,077 INFO  -> (TCPServer-1) [org.dcm4cheri.net.FsmImpl] Socket[addr=/10.10.186.29,port=56355,localport=11112]
2017-12-29 11:39:54,077 INFO  MPC02090->DCM4CHEE (TCPServer-1) [org.dcm4cheri.net.FsmImpl] received AAssociateRQ
appCtxName: 1.2.840.10008.3.1.1.1/DICOM Application Context Name
implClass: 1.2.840.113619.6.118
implVersion: ECHOPAC_201
calledAET: DCM4CHEE
callingAET: MPC02090
maxPDULen: 32768
asyncOpsWindow:
pc-1: as=1.2.840.10008.5.1.4.1.2.2.1/Study Root Query/Retrieve Information Model - FIND
ts=1.2.840.10008.1.2.1/Explicit VR Little Endian
ts=1.2.840.10008.1.2/Implicit VR Little Endian
ts=1.2.840.10008.1.2.2/Explicit VR Big Endian
pc-3: as=1.2.840.10008.5.1.4.1.2.2.2/Study Root Query/Retrieve Information Model - MOVE
ts=1.2.840.10008.1.2.1/Explicit VR Little Endian
ts=1.2.840.10008.1.2/Implicit VR Little Endian
ts=1.2.840.10008.1.2.2/Explicit VR Big Endian
2017-12-29 11:39:54,077 INFO  MPC02090->DCM4CHEE (TCPServer-1) [org.dcm4cheri.net.FsmImpl] sending AAssociateAC
appCtxName: 1.2.840.10008.3.1.1.1/DICOM Application Context Name
implClass: 1.2.40.0.13.1.1.1
implVersion: dcm4che-1.4.34
calledAET: DCM4CHEE
callingAET: MPC02090
maxPDULen: 16352
asyncOpsWindow:
pc-1: 0 - acceptance
ts=1.2.840.10008.1.2/Implicit VR Little Endian
pc-3: 0 - acceptance
ts=1.2.840.10008.1.2/Implicit VR Little Endian
2017-12-29 11:39:54,077 INFO  MPC02090->DCM4CHEE (TCPServer-1) [org.dcm4cheri.net.FsmImpl] received [pc-1] 7:C_FIND_RQ with Dataset
class: 1.2.840.10008.5.1.4.1.2.2.1/Study Root Query/Retrieve Information Model - FIND
2017-12-29 11:39:54,077 WARN  MPC02090->DCM4CHEE (TCPServer-1) [org.dcm4chex.archive.ejb.jdbc.BaseCmd] (0018,0010) Contrast/Bolus Agent,LO,*0,#0,[] not supported for existence and/or matching
2017-12-29 11:39:54,077 INFO  MPC02090->DCM4CHEE (TCPServer-1) [org.dcm4cheri.net.FsmImpl] sending [pc-1] 7:C_FIND_RSP with Dataset
class: 1.2.840.10008.5.1.4.1.2.2.1/Study Root Query/Retrieve Information Model - FIND
status: ff01
2017-12-29 11:39:54,077 INFO  MPC02090->DCM4CHEE (TCPServer-1) [org.dcm4cheri.net.FsmImpl] sending [pc-1] 7:C_FIND_RSP with Dataset
class: 1.2.840.10008.5.1.4.1.2.2.1/Study Root Query/Retrieve Information Model - FIND
status: ff01
2017-12-29 11:39:54,077 INFO  MPC02090->DCM4CHEE (TCPServer-1) [org.dcm4cheri.net.FsmImpl] sending [pc-1] 7:C_FIND_RSP with Dataset
class: 1.2.840.10008.5.1.4.1.2.2.1/Study Root Query/Retrieve Information Model - FIND
status: ff01
2017-12-29 11:39:54,077 INFO  MPC02090->DCM4CHEE (TCPServer-1) [org.dcm4cheri.net.FsmImpl] sending [pc-1] 7:C_FIND_RSP with Dataset
class: 1.2.840.10008.5.1.4.1.2.2.1/Study Root Query/Retrieve Information Model - FIND
status: ff01
2017-12-29 11:39:54,077 INFO  MPC02090->DCM4CHEE (TCPServer-1) [org.dcm4cheri.net.FsmImpl] sending [pc-1] 7:C_FIND_RSP with Dataset
class: 1.2.840.10008.5.1.4.1.2.2.1/Study Root Query/Retrieve Information Model - FIND
status: ff01
2017-12-29 11:39:54,077 INFO  MPC02090->DCM4CHEE (TCPServer-1) [org.dcm4cheri.net.FsmImpl] sending [pc-1] 7:C_FIND_RSP with Dataset
class: 1.2.840.10008.5.1.4.1.2.2.1/Study Root Query/Retrieve Information Model - FIND
status: ff01
2017-12-29 11:39:54,077 INFO  MPC02090->DCM4CHEE (TCPServer-1) [org.dcm4cheri.net.FsmImpl] sending [pc-1] 7:C_FIND_RSP with Dataset
class: 1.2.840.10008.5.1.4.1.2.2.1/Study Root Query/Retrieve Information Model - FIND
status: ff01
2017-12-29 11:39:54,077 INFO  MPC02090->DCM4CHEE (TCPServer-1) [org.dcm4cheri.net.FsmImpl] sending [pc-1] 7:C_FIND_RSP with Dataset
class: 1.2.840.10008.5.1.4.1.2.2.1/Study Root Query/Retrieve Information Model - FIND
status: ff01
2017-12-29 11:39:54,077 INFO  MPC02090->DCM4CHEE (TCPServer-1) [org.dcm4cheri.net.FsmImpl] sending [pc-1] 7:C_FIND_RSP with Dataset
class: 1.2.840.10008.5.1.4.1.2.2.1/Study Root Query/Retrieve Information Model - FIND
status: ff01
2017-12-29 11:39:54,077 INFO  MPC02090->DCM4CHEE (TCPServer-1) [org.dcm4cheri.net.FsmImpl] sending [pc-1] 7:C_FIND_RSP with Dataset
class: 1.2.840.10008.5.1.4.1.2.2.1/Study Root Query/Retrieve Information Model - FIND
status: ff01
2017-12-29 11:39:54,077 INFO  MPC02090->DCM4CHEE (TCPServer-1) [org.dcm4cheri.net.FsmImpl] sending [pc-1] 7:C_FIND_RSP with Dataset
class: 1.2.840.10008.5.1.4.1.2.2.1/Study Root Query/Retrieve Information Model - FIND
status: ff01
2017-12-29 11:39:54,077 INFO  MPC02090->DCM4CHEE (TCPServer-1) [org.dcm4cheri.net.FsmImpl] sending [pc-1] 7:C_FIND_RSP with Dataset
class: 1.2.840.10008.5.1.4.1.2.2.1/Study Root Query/Retrieve Information Model - FIND
status: ff01
2017-12-29 11:39:54,077 INFO  MPC02090->DCM4CHEE (TCPServer-1) [org.dcm4cheri.net.FsmImpl] sending [pc-1] 7:C_FIND_RSP with Dataset
class: 1.2.840.10008.5.1.4.1.2.2.1/Study Root Query/Retrieve Information Model - FIND
status: ff01
2017-12-29 11:39:54,077 INFO  MPC02090->DCM4CHEE (TCPServer-1) [org.dcm4cheri.net.FsmImpl] sending [pc-1] 7:C_FIND_RSP with Dataset
class: 1.2.840.10008.5.1.4.1.2.2.1/Study Root Query/Retrieve Information Model - FIND
status: ff01
2017-12-29 11:39:54,077 INFO  MPC02090->DCM4CHEE (TCPServer-1) [org.dcm4cheri.net.FsmImpl] sending [pc-1] 7:C_FIND_RSP with Dataset
class: 1.2.840.10008.5.1.4.1.2.2.1/Study Root Query/Retrieve Information Model - FIND
status: ff01
2017-12-29 11:39:54,077 INFO  MPC02090->DCM4CHEE (TCPServer-1) [org.dcm4cheri.net.FsmImpl] sending [pc-1] 7:C_FIND_RSP with Dataset
class: 1.2.840.10008.5.1.4.1.2.2.1/Study Root Query/Retrieve Information Model - FIND
status: ff01
2017-12-29 11:39:54,077 INFO  MPC02090->DCM4CHEE (TCPServer-1) [org.dcm4cheri.net.FsmImpl] sending [pc-1] 7:C_FIND_RSP with Dataset
class: 1.2.840.10008.5.1.4.1.2.2.1/Study Root Query/Retrieve Information Model - FIND
status: ff01
2017-12-29 11:39:54,077 INFO  MPC02090->DCM4CHEE (TCPServer-1) [org.dcm4cheri.net.FsmImpl] sending [pc-1] 7:C_FIND_RSP with Dataset
class: 1.2.840.10008.5.1.4.1.2.2.1/Study Root Query/Retrieve Information Model - FIND
status: ff01
2017-12-29 11:39:54,077 INFO  MPC02090->DCM4CHEE (TCPServer-1) [org.dcm4cheri.net.FsmImpl] sending [pc-1] 7:C_FIND_RSP with Dataset
class: 1.2.840.10008.5.1.4.1.2.2.1/Study Root Query/Retrieve Information Model - FIND
status: ff01
2017-12-29 11:39:54,077 INFO  MPC02090->DCM4CHEE (TCPServer-1) [org.dcm4cheri.net.FsmImpl] sending [pc-1] 7:C_FIND_RSP with Dataset
class: 1.2.840.10008.5.1.4.1.2.2.1/Study Root Query/Retrieve Information Model - FIND
status: ff01
2017-12-29 11:39:54,077 INFO  MPC02090->DCM4CHEE (TCPServer-1) [org.dcm4cheri.net.FsmImpl] sending [pc-1] 7:C_FIND_RSP with Dataset
class: 1.2.840.10008.5.1.4.1.2.2.1/Study Root Query/Retrieve Information Model - FIND
status: ff01
2017-12-29 11:39:54,077 INFO  MPC02090->DCM4CHEE (TCPServer-1) [org.dcm4cheri.net.FsmImpl] sending [pc-1] 7:C_FIND_RSP with Dataset
class: 1.2.840.10008.5.1.4.1.2.2.1/Study Root Query/Retrieve Information Model - FIND
status: ff01
2017-12-29 11:39:54,077 INFO  MPC02090->DCM4CHEE (TCPServer-1) [org.dcm4cheri.net.FsmImpl] sending [pc-1] 7:C_FIND_RSP with Dataset
class: 1.2.840.10008.5.1.4.1.2.2.1/Study Root Query/Retrieve Information Model - FIND
status: ff01
2017-12-29 11:39:54,077 INFO  MPC02090->DCM4CHEE (TCPServer-1) [org.dcm4cheri.net.FsmImpl] sending [pc-1] 7:C_FIND_RSP with Dataset
class: 1.2.840.10008.5.1.4.1.2.2.1/Study Root Query/Retrieve Information Model - FIND
status: ff01
2017-12-29 11:39:54,077 INFO  MPC02090->DCM4CHEE (TCPServer-1) [org.dcm4cheri.net.FsmImpl] sending [pc-1] 7:C_FIND_RSP with Dataset
class: 1.2.840.10008.5.1.4.1.2.2.1/Study Root Query/Retrieve Information Model - FIND
status: ff01
2017-12-29 11:39:54,077 INFO  MPC02090->DCM4CHEE (TCPServer-1) [org.dcm4cheri.net.FsmImpl] sending [pc-1] 7:C_FIND_RSP with Dataset
class: 1.2.840.10008.5.1.4.1.2.2.1/Study Root Query/Retrieve Information Model - FIND
status: ff01
2017-12-29 11:39:54,077 INFO  MPC02090->DCM4CHEE (TCPServer-1) [org.dcm4cheri.net.FsmImpl] sending [pc-1] 7:C_FIND_RSP with Dataset
class: 1.2.840.10008.5.1.4.1.2.2.1/Study Root Query/Retrieve Information Model - FIND
status: ff01
2017-12-29 11:39:54,077 INFO  MPC02090->DCM4CHEE (TCPServer-1) [org.dcm4cheri.net.FsmImpl] sending [pc-1] 7:C_FIND_RSP with Dataset
class: 1.2.840.10008.5.1.4.1.2.2.1/Study Root Query/Retrieve Information Model - FIND
status: ff01
2017-12-29 11:39:54,077 INFO  MPC02090->DCM4CHEE (TCPServer-1) [org.dcm4cheri.net.FsmImpl] sending [pc-1] 7:C_FIND_RSP with Dataset
class: 1.2.840.10008.5.1.4.1.2.2.1/Study Root Query/Retrieve Information Model - FIND
status: ff01
2017-12-29 11:39:54,077 INFO  MPC02090->DCM4CHEE (TCPServer-1) [org.dcm4cheri.net.FsmImpl] sending [pc-1] 7:C_FIND_RSP with Dataset
class: 1.2.840.10008.5.1.4.1.2.2.1/Study Root Query/Retrieve Information Model - FIND
status: ff01
2017-12-29 11:39:54,077 INFO  MPC02090->DCM4CHEE (TCPServer-1) [org.dcm4cheri.net.FsmImpl] sending [pc-1] 7:C_FIND_RSP with Dataset
class: 1.2.840.10008.5.1.4.1.2.2.1/Study Root Query/Retrieve Information Model - FIND
status: ff01
2017-12-29 11:39:54,077 INFO  MPC02090->DCM4CHEE (TCPServer-1) [org.dcm4cheri.net.FsmImpl] sending [pc-1] 7:C_FIND_RSP with Dataset
class: 1.2.840.10008.5.1.4.1.2.2.1/Study Root Query/Retrieve Information Model - FIND
status: ff01
2017-12-29 11:39:54,077 INFO  MPC02090->DCM4CHEE (TCPServer-1) [org.dcm4cheri.net.FsmImpl] sending [pc-1] 7:C_FIND_RSP with Dataset
class: 1.2.840.10008.5.1.4.1.2.2.1/Study Root Query/Retrieve Information Model - FIND
status: ff01
2017-12-29 11:39:54,077 INFO  MPC02090->DCM4CHEE (TCPServer-1) [org.dcm4cheri.net.FsmImpl] sending [pc-1] 7:C_FIND_RSP with Dataset
class: 1.2.840.10008.5.1.4.1.2.2.1/Study Root Query/Retrieve Information Model - FIND
status: ff01
2017-12-29 11:39:54,077 INFO  MPC02090->DCM4CHEE (TCPServer-1) [org.dcm4cheri.net.FsmImpl] sending [pc-1] 7:C_FIND_RSP with Dataset
class: 1.2.840.10008.5.1.4.1.2.2.1/Study Root Query/Retrieve Information Model - FIND
status: ff01
2017-12-29 11:39:54,077 INFO  MPC02090->DCM4CHEE (TCPServer-1) [org.dcm4cheri.net.FsmImpl] sending [pc-1] 7:C_FIND_RSP
class: 1.2.840.10008.5.1.4.1.2.2.1/Study Root Query/Retrieve Information Model - FIND
status: 0
2017-12-29 11:39:54,093 INFO  MPC02090->DCM4CHEE (ActiveAssoc-1402-1) [org.dcm4cheri.net.FsmImpl] received A-RELEASE-RQ
2017-12-29 11:39:54,093 INFO  MPC02090->DCM4CHEE (ActiveAssoc-1402-1) [org.dcm4cheri.net.FsmImpl] sending A-RELEASE-RP
2017-12-29 11:39:54,093 INFO  -> (TCPServer-1) [org.dcm4cheri.server.ServerImpl] finished - Socket[addr=/10.10.186.29,port=56355,localport=11112]
2017-12-29 11:39:54,124 INFO  MPC02090->DCM4CHEE (ActiveAssoc-1401-1) [org.dcm4cheri.net.FsmImpl] closing connection - Socket[addr=/10.10.186.29,port=56354,localport=11112]
2017-12-29 11:39:54,155 INFO  MPC02090->DCM4CHEE (ActiveAssoc-1402-1) [org.dcm4cheri.net.FsmImpl] closing connection - Socket[addr=/10.10.186.29,port=56355,localport=11112]
2017-12-29 11:39:54,155 INFO  -> (TCPServer-1-1) [org.dcm4cheri.server.ServerImpl] handle - Socket[addr=/10.10.186.29,port=56356,localport=11112]
2017-12-29 11:39:54,155 INFO  -> (TCPServer-1-1) [org.dcm4cheri.net.FsmImpl] Socket[addr=/10.10.186.29,port=56356,localport=11112]
2017-12-29 11:39:54,155 INFO  MPC02090->DCM4CHEE (TCPServer-1-1) [org.dcm4cheri.net.FsmImpl] received AAssociateRQ
appCtxName: 1.2.840.10008.3.1.1.1/DICOM Application Context Name
implClass: 1.2.840.113619.6.118
implVersion: ECHOPAC_201
calledAET: DCM4CHEE
callingAET: MPC02090
maxPDULen: 32768
asyncOpsWindow:
pc-1: as=1.2.840.10008.5.1.4.1.2.2.1/Study Root Query/Retrieve Information Model - FIND
ts=1.2.840.10008.1.2.1/Explicit VR Little Endian
ts=1.2.840.10008.1.2/Implicit VR Little Endian
ts=1.2.840.10008.1.2.2/Explicit VR Big Endian
pc-3: as=1.2.840.10008.5.1.4.1.2.2.2/Study Root Query/Retrieve Information Model - MOVE
ts=1.2.840.10008.1.2.1/Explicit VR Little Endian
ts=1.2.840.10008.1.2/Implicit VR Little Endian
ts=1.2.840.10008.1.2.2/Explicit VR Big Endian
2017-12-29 11:39:54,155 INFO  MPC02090->DCM4CHEE (TCPServer-1-1) [org.dcm4cheri.net.FsmImpl] sending AAssociateAC
appCtxName: 1.2.840.10008.3.1.1.1/DICOM Application Context Name
implClass: 1.2.40.0.13.1.1.1
implVersion: dcm4che-1.4.34
calledAET: DCM4CHEE
callingAET: MPC02090
maxPDULen: 16352
asyncOpsWindow:
pc-1: 0 - acceptance
ts=1.2.840.10008.1.2/Implicit VR Little Endian
pc-3: 0 - acceptance
ts=1.2.840.10008.1.2/Implicit VR Little Endian
2017-12-29 11:39:54,155 INFO  MPC02090->DCM4CHEE (TCPServer-1-1) [org.dcm4cheri.net.FsmImpl] received [pc-3] 8:C_MOVE_RQ with Dataset
class: 1.2.840.10008.5.1.4.1.2.2.2/Study Root Query/Retrieve Information Model - MOVE
move dest: MPC02090
2017-12-29 11:39:54,171 INFO  MPC02090->DCM4CHEE (Timer-2) [org.dcm4cheri.net.FsmImpl] sending [pc-3] 8:C_MOVE_RSP
class: 1.2.840.10008.5.1.4.1.2.2.2/Study Root Query/Retrieve Information Model - MOVE
status: ff00
2017-12-29 11:39:54,171 INFO  MPC02090->DCM4CHEE (Thread-55385) [org.dcm4cheri.net.FsmImpl] Socket[addr=mpc02090.somedomain.com/10.10.186.29,port=104,localport=64148]
2017-12-29 11:39:54,171 INFO  DCM4CHEE->MPC02090 (Thread-55385) [org.dcm4cheri.net.FsmImpl] sending AAssociateRQ
appCtxName: 1.2.840.10008.3.1.1.1/DICOM Application Context Name
implClass: 1.2.40.0.13.1.1.1
implVersion: dcm4che-1.4.34
calledAET: MPC02090
callingAET: DCM4CHEE
maxPDULen: 16352
asyncOpsWindow: maxOpsInvoked=0, maxOpsPerformed=1
pc-1: as=1.2.840.10008.5.1.4.1.1.6.1/Ultrasound Image Storage
ts=1.2.840.10008.1.2/Implicit VR Little Endian
pc-3: as=1.2.840.10008.5.1.4.1.1.6.1/Ultrasound Image Storage
ts=1.2.840.10008.1.2.1/Explicit VR Little Endian
pc-5: as=1.2.840.10008.5.1.4.1.1.6.1/Ultrasound Image Storage
ts=1.2.840.10008.1.2.4.50/JPEG Baseline (Process 1)
pc-7: as=1.2.840.10008.5.1.4.1.1.3.1/Ultrasound Multi-frame Image Storage
ts=1.2.840.10008.1.2/Implicit VR Little Endian
pc-9: as=1.2.840.10008.5.1.4.1.1.3.1/Ultrasound Multi-frame Image Storage
ts=1.2.840.10008.1.2.1/Explicit VR Little Endian
pc-11: as=1.2.840.10008.5.1.4.1.1.3.1/Ultrasound Multi-frame Image Storage
ts=1.2.840.10008.1.2.4.50/JPEG Baseline (Process 1)
2017-12-29 11:39:55,013 INFO  DCM4CHEE->MPC02090 (Thread-55385) [org.dcm4cheri.net.FsmImpl] received AAssociateAC
appCtxName: 1.2.840.10008.3.1.1.1/DICOM Application Context Name
implClass: 1.2.840.113619.6.118
implVersion: ECHOPAC_201
calledAET: MPC02090
callingAET: DCM4CHEE
maxPDULen: 32768
asyncOpsWindow: maxOpsInvoked=1, maxOpsPerformed=1
pc-1: 0 - acceptance
ts=1.2.840.10008.1.2/Implicit VR Little Endian
pc-3: 0 - acceptance
ts=1.2.840.10008.1.2.1/Explicit VR Little Endian
pc-5: 0 - acceptance
ts=1.2.840.10008.1.2.4.50/JPEG Baseline (Process 1)
pc-7: 0 - acceptance
ts=1.2.840.10008.1.2/Implicit VR Little Endian
pc-9: 0 - acceptance
ts=1.2.840.10008.1.2.1/Explicit VR Little Endian
pc-11: 0 - acceptance
ts=1.2.840.10008.1.2.4.50/JPEG Baseline (Process 1)
2017-12-29 11:39:55,544 INFO  DCM4CHEE->MPC02090 (Thread-55385) [org.dcm4cheri.net.FsmImpl] sending [pc-11] 1:C_STORE_RQ with Dataset
class: 1.2.840.10008.5.1.4.1.1.3.1/Ultrasound Multi-frame Image Storage
inst: 1.2.840.113619.2.391.1050.1495006362.40.1.512/?
2017-12-29 11:39:55,544 INFO  DCM4CHEE->MPC02090 (Thread-55385) [org.dcm4chex.archive.util.FileDataSource] M-READ file:D:\DICOM\2017\5\17\12\05990E76\05991237\65822898
2017-12-29 11:39:59,179 INFO  MPC02090->DCM4CHEE (Timer-2) [org.dcm4cheri.net.FsmImpl] sending [pc-3] 8:C_MOVE_RSP
class: 1.2.840.10008.5.1.4.1.2.2.2/Study Root Query/Retrieve Information Model - MOVE
status: ff00
2017-12-29 11:39:59,335 INFO  DCM4CHEE->MPC02090 (ActiveAssoc-1404) [org.dcm4cheri.net.FsmImpl] received [pc-11] 1:C_STORE_RSP
class: 1.2.840.10008.5.1.4.1.1.3.1/Ultrasound Multi-frame Image Storage
inst: 1.2.840.113619.2.391.1050.1495006362.40.1.512/?
status: 0
2017-12-29 11:39:59,335 INFO  DCM4CHEE->MPC02090 (Thread-55385) [org.dcm4cheri.net.FsmImpl] sending [pc-11] 2:C_STORE_RQ with Dataset
class: 1.2.840.10008.5.1.4.1.1.3.1/Ultrasound Multi-frame Image Storage
inst: 1.2.840.113619.2.391.1050.1495006367.41.1.512/?
2017-12-29 11:39:59,335 INFO  DCM4CHEE->MPC02090 (Thread-55385) [org.dcm4chex.archive.util.FileDataSource] M-READ file:D:\DICOM\2017\5\17\12\05990E76\05991237\5FD79974
2017-12-29 11:40:03,718 INFO  DCM4CHEE->MPC02090 (ActiveAssoc-1404) [org.dcm4cheri.net.FsmImpl] received [pc-11] 2:C_STORE_RSP
class: 1.2.840.10008.5.1.4.1.1.3.1/Ultrasound Multi-frame Image Storage
inst: 1.2.840.113619.2.391.1050.1495006367.41.1.512/?
status: 0
2017-12-29 11:40:03,718 INFO  DCM4CHEE->MPC02090 (Thread-55385) [org.dcm4cheri.net.FsmImpl] sending [pc-11] 3:C_STORE_RQ with Dataset
class: 1.2.840.10008.5.1.4.1.1.3.1/Ultrasound Multi-frame Image Storage
inst: 1.2.840.113619.2.391.1050.1495006375.42.1.512/?
2017-12-29 11:40:03,718 INFO  DCM4CHEE->MPC02090 (Thread-55385) [org.dcm4chex.archive.util.FileDataSource] M-READ file:D:\DICOM\2017\5\17\12\05990E76\05991237\410FA538
2017-12-29 11:40:04,186 INFO  MPC02090->DCM4CHEE (Timer-2) [org.dcm4cheri.net.FsmImpl] sending [pc-3] 8:C_MOVE_RSP
class: 1.2.840.10008.5.1.4.1.2.2.2/Study Root Query/Retrieve Information Model - MOVE
status: ff00
2017-12-29 11:40:04,639 INFO  -> (Thread-55387) [org.dcm4chex.archive.mbean.FileSystemMgt2Service] Check file system group ONLINE_STORAGE for deletion of orphaned private files
2017-12-29 11:40:04,639 INFO  -> (Thread-55388) [org.dcm4chex.archive.mbean.FileSystemMgt2Service] Check file system group LOSSY_STORAGE for deletion of orphaned private files
2017-12-29 11:40:07,275 INFO  DCM4CHEE->MPC02090 (ActiveAssoc-1404) [org.dcm4cheri.net.FsmImpl] received [pc-11] 3:C_STORE_RSP
class: 1.2.840.10008.5.1.4.1.1.3.1/Ultrasound Multi-frame Image Storage
inst: 1.2.840.113619.2.391.1050.1495006375.42.1.512/?
status: 0
2017-12-29 11:40:07,275 INFO  DCM4CHEE->MPC02090 (Thread-55385) [org.dcm4cheri.net.FsmImpl] sending [pc-11] 4:C_STORE_RQ with Dataset
class: 1.2.840.10008.5.1.4.1.1.3.1/Ultrasound Multi-frame Image Storage
inst: 1.2.840.113619.2.391.1050.1495006386.43.1.512/?
2017-12-29 11:40:07,275 INFO  DCM4CHEE->MPC02090 (Thread-55385) [org.dcm4chex.archive.util.FileDataSource] M-READ file:D:\DICOM\2017\5\17\12\05990E76\05991237\FF240459
2017-12-29 11:40:09,194 INFO  MPC02090->DCM4CHEE (Timer-2) [org.dcm4cheri.net.FsmImpl] sending [pc-3] 8:C_MOVE_RSP
class: 1.2.840.10008.5.1.4.1.2.2.2/Study Root Query/Retrieve Information Model - MOVE
status: ff00
2017-12-29 11:40:11,596 INFO  DCM4CHEE->MPC02090 (ActiveAssoc-1404) [org.dcm4cheri.net.FsmImpl] received [pc-11] 4:C_STORE_RSP
class: 1.2.840.10008.5.1.4.1.1.3.1/Ultrasound Multi-frame Image Storage
inst: 1.2.840.113619.2.391.1050.1495006386.43.1.512/?
status: 0
2017-12-29 11:40:11,596 INFO  DCM4CHEE->MPC02090 (Thread-55385) [org.dcm4cheri.net.FsmImpl] sending [pc-5] 5:C_STORE_RQ with Dataset
class: 1.2.840.10008.5.1.4.1.1.6.1/Ultrasound Image Storage
inst: 1.2.840.113619.2.391.1050.1495006397.44.1/?
2017-12-29 11:40:11,596 INFO  DCM4CHEE->MPC02090 (Thread-55385) [org.dcm4chex.archive.util.FileDataSource] M-READ file:D:\DICOM\2017\5\17\12\05990E76\05991237\82222652
2017-12-29 11:40:11,596 INFO  DCM4CHEE->MPC02090 (ActiveAssoc-1404) [org.dcm4cheri.net.FsmImpl] received [pc-5] 5:C_STORE_RSP
class: 1.2.840.10008.5.1.4.1.1.6.1/Ultrasound Image Storage
inst: 1.2.840.113619.2.391.1050.1495006397.44.1/?
status: 0
2017-12-29 11:40:11,596 INFO  DCM4CHEE->MPC02090 (Thread-55385) [org.dcm4cheri.net.FsmImpl] sending [pc-5] 6:C_STORE_RQ with Dataset
class: 1.2.840.10008.5.1.4.1.1.6.1/Ultrasound Image Storage
inst: 1.2.840.113619.2.391.1050.1495006446.45.1/?
2017-12-29 11:40:11,596 INFO  DCM4CHEE->MPC02090 (Thread-55385) [org.dcm4chex.archive.util.FileDataSource] M-READ file:D:\DICOM\2017\5\17\12\05990E76\05991237\DFCF560E
2017-12-29 11:40:11,659 INFO  DCM4CHEE->MPC02090 (ActiveAssoc-1404) [org.dcm4cheri.net.FsmImpl] received [pc-5] 6:C_STORE_RSP
class: 1.2.840.10008.5.1.4.1.1.6.1/Ultrasound Image Storage
inst: 1.2.840.113619.2.391.1050.1495006446.45.1/?
status: 0
2017-12-29 11:40:11,659 INFO  DCM4CHEE->MPC02090 (Thread-55385) [org.dcm4cheri.net.FsmImpl] sending [pc-5] 7:C_STORE_RQ with Dataset
class: 1.2.840.10008.5.1.4.1.1.6.1/Ultrasound Image Storage
inst: 1.2.840.113619.2.391.1050.1495006478.46.1/?
2017-12-29 11:40:11,659 INFO  DCM4CHEE->MPC02090 (Thread-55385) [org.dcm4chex.archive.util.FileDataSource] M-READ file:D:\DICOM\2017\5\17\12\05990E76\05991237\81EBBCD0
2017-12-29 11:40:11,721 INFO  DCM4CHEE->MPC02090 (ActiveAssoc-1404) [org.dcm4cheri.net.FsmImpl] received [pc-5] 7:C_STORE_RSP
class: 1.2.840.10008.5.1.4.1.1.6.1/Ultrasound Image Storage
inst: 1.2.840.113619.2.391.1050.1495006478.46.1/?
status: 0
2017-12-29 11:40:11,721 INFO  DCM4CHEE->MPC02090 (Thread-55385) [org.dcm4cheri.net.FsmImpl] sending [pc-5] 8:C_STORE_RQ with Dataset
class: 1.2.840.10008.5.1.4.1.1.6.1/Ultrasound Image Storage
inst: 1.2.840.113619.2.391.1050.1495006525.47.1/?
2017-12-29 11:40:11,721 INFO  DCM4CHEE->MPC02090 (Thread-55385) [org.dcm4chex.archive.util.FileDataSource] M-READ file:D:\DICOM\2017\5\17\12\05990E76\05991237\DC2F3B4E
2017-12-29 11:40:11,784 INFO  DCM4CHEE->MPC02090 (ActiveAssoc-1404) [org.dcm4cheri.net.FsmImpl] received [pc-5] 8:C_STORE_RSP
class: 1.2.840.10008.5.1.4.1.1.6.1/Ultrasound Image Storage
inst: 1.2.840.113619.2.391.1050.1495006525.47.1/?
status: 0
2017-12-29 11:40:11,784 INFO  DCM4CHEE->MPC02090 (Thread-55385) [org.dcm4cheri.net.FsmImpl] sending [pc-5] 9:C_STORE_RQ with Dataset
class: 1.2.840.10008.5.1.4.1.1.6.1/Ultrasound Image Storage
inst: 1.2.840.113619.2.391.1050.1495006554.48.1/?
2017-12-29 11:40:11,784 INFO  DCM4CHEE->MPC02090 (Thread-55385) [org.dcm4chex.archive.util.FileDataSource] M-READ file:D:\DICOM\2017\5\17\12\05990E76\05991237\792D1833
2017-12-29 11:40:11,784 INFO  DCM4CHEE->MPC02090 (ActiveAssoc-1404) [org.dcm4cheri.net.FsmImpl] received [pc-5] 9:C_STORE_RSP
class: 1.2.840.10008.5.1.4.1.1.6.1/Ultrasound Image Storage
inst: 1.2.840.113619.2.391.1050.1495006554.48.1/?
status: 0
2017-12-29 11:40:11,784 INFO  DCM4CHEE->MPC02090 (Thread-55385) [org.dcm4cheri.net.FsmImpl] sending [pc-5] 10:C_STORE_RQ with Dataset
class: 1.2.840.10008.5.1.4.1.1.6.1/Ultrasound Image Storage
inst: 1.2.840.113619.2.391.1050.1495006589.49.1/?
2017-12-29 11:40:11,784 INFO  DCM4CHEE->MPC02090 (Thread-55385) [org.dcm4chex.archive.util.FileDataSource] M-READ file:D:\DICOM\2017\5\17\12\05990E76\05991237\206808D2
2017-12-29 11:40:11,846 INFO  DCM4CHEE->MPC02090 (ActiveAssoc-1404) [org.dcm4cheri.net.FsmImpl] received [pc-5] 10:C_STORE_RSP
class: 1.2.840.10008.5.1.4.1.1.6.1/Ultrasound Image Storage
inst: 1.2.840.113619.2.391.1050.1495006589.49.1/?
status: 0
2017-12-29 11:40:11,846 INFO  DCM4CHEE->MPC02090 (Thread-55385) [org.dcm4cheri.net.FsmImpl] sending [pc-5] 11:C_STORE_RQ with Dataset
class: 1.2.840.10008.5.1.4.1.1.6.1/Ultrasound Image Storage
inst: 1.2.840.113619.2.391.1050.1495006624.50.1/?
2017-12-29 11:40:11,846 INFO  DCM4CHEE->MPC02090 (Thread-55385) [org.dcm4chex.archive.util.FileDataSource] M-READ file:D:\DICOM\2017\5\17\12\05990E76\05991237\425BE9A6
2017-12-29 11:40:11,862 INFO  DCM4CHEE->MPC02090 (ActiveAssoc-1404) [org.dcm4cheri.net.FsmImpl] received [pc-5] 11:C_STORE_RSP
class: 1.2.840.10008.5.1.4.1.1.6.1/Ultrasound Image Storage
inst: 1.2.840.113619.2.391.1050.1495006624.50.1/?
status: 0
2017-12-29 11:40:11,877 INFO  DCM4CHEE->MPC02090 (Thread-55385) [org.dcm4cheri.net.FsmImpl] sending [pc-11] 12:C_STORE_RQ with Dataset
class: 1.2.840.10008.5.1.4.1.1.3.1/Ultrasound Multi-frame Image Storage
inst: 1.2.840.113619.2.391.1050.1495006651.51.1.512/?
2017-12-29 11:40:11,877 INFO  DCM4CHEE->MPC02090 (Thread-55385) [org.dcm4chex.archive.util.FileDataSource] M-READ file:D:\DICOM\2017\5\17\12\05990E76\05991237\90305CF5
2017-12-29 11:40:14,202 INFO  MPC02090->DCM4CHEE (Timer-2) [org.dcm4cheri.net.FsmImpl] sending [pc-3] 8:C_MOVE_RSP
class: 1.2.840.10008.5.1.4.1.2.2.2/Study Root Query/Retrieve Information Model - MOVE
status: ff00
2017-12-29 11:40:15,200 INFO  MPC02090->DCM4CHEE (ActiveAssoc-1403-1) [org.dcm4cheri.net.FsmImpl] received [pc-3] 8:C_CANCEL_RQ
2017-12-29 11:40:15,465 INFO  DCM4CHEE->MPC02090 (ActiveAssoc-1404) [org.dcm4cheri.net.FsmImpl] received [pc-11] 12:C_STORE_RSP
class: 1.2.840.10008.5.1.4.1.1.3.1/Ultrasound Multi-frame Image Storage
inst: 1.2.840.113619.2.391.1050.1495006651.51.1.512/?
status: 0
2017-12-29 11:40:15,465 INFO  DCM4CHEE->MPC02090 (Thread-55385) [org.dcm4cheri.net.FsmImpl] sending A-RELEASE-RQ
2017-12-29 11:40:15,465 INFO  DCM4CHEE->MPC02090 (ActiveAssoc-1404) [org.dcm4cheri.net.FsmImpl] received A-RELEASE-RP
2017-12-29 11:40:15,465 INFO  DCM4CHEE->MPC02090 (ActiveAssoc-1404) [org.dcm4cheri.net.FsmImpl] closing connection - Socket[addr=mpc02090.somedomain.com/10.10.186.29,port=104,localport=64148]
2017-12-29 11:40:15,481 INFO  MPC02090->DCM4CHEE (Thread-55385) [org.dcm4cheri.net.FsmImpl] sending [pc-3] 8:C_MOVE_RSP
class: 1.2.840.10008.5.1.4.1.2.2.2/Study Root Query/Retrieve Information Model - MOVE
status: fe00
2017-12-29 11:40:15,481 INFO  MPC02090->DCM4CHEE (ActiveAssoc-1403-1) [org.dcm4cheri.net.FsmImpl] received A-RELEASE-RQ
2017-12-29 11:40:15,481 INFO  MPC02090->DCM4CHEE (ActiveAssoc-1403-1) [org.dcm4cheri.net.FsmImpl] sending A-RELEASE-RP
2017-12-29 11:40:15,481 INFO  -> (TCPServer-1-1) [org.dcm4cheri.server.ServerImpl] finished - Socket[addr=/10.10.186.29,port=56356,localport=11112]
2017-12-29 11:40:15,543 INFO  MPC02090->DCM4CHEE (ActiveAssoc-1403-1) [org.dcm4cheri.net.FsmImpl] closing connection - Socket[addr=/10.10.186.29,port=56356,localport=11112]
Reply all
Reply to author
Forward
0 new messages