Why Does Weasis Display Three Sets of Measurement Values Compared to Single Values in Other DICOM Viewers?

33 views
Skip to first unread message

Yunus eşme

unread,
Jan 22, 2025, 6:15:55 AMJan 22
to dcm4che

Hello,

I performed similar measurements for the same region of a PET/CT image using Osirix, Weasis, and Horos. However, in Weasis, the min, max, mean, median, and StDev values appear three times, as shown in the attached image.

Why does Weasis display these values three times, while other DICOM viewers show a single value? Should we take the average of these three values in Weasis? If yes, why are the values displayed separately instead of as a single number? Is there a way to view only one set of data in Weasis, like in the other viewers?

Additionally, if StDev corresponds to SD, why does it show a greater deviation compared to other viewers?

Lastly, while other applications don’t show units for these values, Weasis displays them as HU. Why is this the case? Does this have any significance for PET/CT evaluations?

I would appreciate your guidance on these questions. Thank you!

HOROS.png
Osirix.png
Weasis.png

Nicolas Roduit

unread,
Jan 23, 2025, 12:56:27 PMJan 23
to dcm4che
The 3 values represent the 3 channels of a color image. I return the question, why display only one value, is it an average or a grayscale conversion (which is not exactly an average)?
However, It would be better to measure directly on the PET image, as this image is a fusion with reduced intensity values (16-bit to 8-bit). See also the other statistics values.

There are several ways to calculate the standard deviation. Weasis uses the OpenCV function.

For units, it's a bug because Weasis displays HU for all CT images. In this case, the modality after fusion is retained in CT. I've opened a bug for this specific case.