Hi,
I also saw this section on the webtool troubleshooting about low R2:
---
Low R2 value
A low R2 can be caused when the settings are not optimal or when the sequence quality is not good.
Settings
By default, the decomposition window is set to its maximum size and the
Indel size range is set to 10. The settings can be adjusted in advanced
settings.
- Large indels are present in the sample. By default the
decomposition is calculated with a maximal indel size of 10. When larger
indels are present, they will cannot be modeled, which will result in a
low R2. Try to increase the Indel size range to test if this improves the fit (see figure Indel size range)
- Poor local quality of the sequence trace. Often the end of the
sequence is of low quality. This can be observed in the quality plot
that shows a high aberrant sequence signal at the end of the sequence
trace (see figure Poor quality sequence end). This compromises the
decomposition of the sequence trace. Adjust the boundaries of the
decomposition window in such a way that it will not overlap with the
region that is of low quality.
- Repetitive regions in the sequence trace. These regions can be
observed in the quality plot as a sudden stretch without aberrant
nucleotides (see figure Repetitive region). This region might interfere
with the decomposition of the sequence trace. Adjust the boundaries of
the decomposition window in such a way that it will not overlap with the
repetitive sequence part.
Wrongly annotated nucleotides
Sometimes the quality of the peaks in chromatogram looks fine, but the
file has some wrongly unannotated or wrongly additional annotated
nucleotides. These will interfere with the indel spectra (see figure
wrongly unannotated nucleotide). TIDE gives a warning when the spacing
between the nucleotides in the chromatogram of the sequence trace are
not consistent, which is often an indication for wrongly unannotated or
wrongly additional annotated nucleotides. Then the sequence file cannot
be used for a reliable TIDE analysis. Try if possible to set the right
boundary of the decomposition window lower. In case this warning stays,
carefully investigate your chromatogram.
Poor sequence quality
Poor sequence quality can be observed in chromatogram (see figure Poor
sequence quality). There is more noise present in the data with results
in a lower R
2.
----
So in general I think your idea that "TIDE-calculated efficiency/R-squared x 100%" is probably not the case and you need to adjust the settings and/or possibly obtain new sequencing trace files if R2 is low.
If you would like, send me an email directly with your sequence traces that give low R2 (and the guide sequence too) and I'll see if I can troubleshoot?
Cheers,
Mike