thanks a lot Keith/Gavin. let me reply to both of you:
Keith: for the most recent sample, we have insertion of a single A at the cut site see below. The top chromatogram is the edited, bottom is WT
I agree, a moderate indel might cause skewed amplification but I think that is less likley than with a 1 bp insertion.
the guide targets mouse chromosome 16, gene Ercc4. I dont routinely run the guides through an off target program because the one I use weighs that in to the output file that it gives me (
https://portals.broadinstitute.org/gpp/public/analysis-tools/sgrna-design). this crRNA happened to be the 5th best when accounting for predicted on target PLUS off target activity. of the ~400 guides that the site gave me, this one was ranked ~22nd for off target score. I honestly dont know if that is good or bad, but it would be in the top ~5% of guides for that gene, so I suppose it should have less off target effects than others I could have chosen. Again, I dont know exactly what 22nd place or top 5% means. Possible that all the other guides were not very good either and the whole bunch is bad.
Gavin: There is a homopolymer at the 5' end of the amplicon and we didnt appreciate that til after we sequenced it. you will notice in the trace that there is a little bit of uncleanliness and I think it is bc of that upstream homopolymer. we sequence the amplicon with the R primer and the baseline is cleaner but we still get the same result.
After looking at the traces from our other genes a little more closely, I think you are right Gavin, that these arent a replacement for NGS or TA cloning of a pool of amplicons. i think that is where this will go next. Maybe this is closer to 50/50 and ICE/TIDE overestimate the indel %.
thanks for your help guys!