Hi Kris
Station Name (0008,1010) is an Equipment level attribute, so any
receiving system (like a PACS) should not assume that it is the
same for an entire Study (if that's what your PACS is doing),
since images in different Series in the same Study may be created
on different equipment.
If you scan and send the documents to the PACS first, before acquiring
the images on the modality and sending them to the PACS, does the station
name of the modality get displayed instead?
What PACS do you have?
Where, and at what level in your browser or display is this information
being shown?
I suspect it would be easier to configure your document scanning
system to not send Station Name at all, rather than get every modality
to change!
Note that Station Name (0008,1010) is optional (Type 3), so not
sending it at all is legal.
If you can't get the document scanning software vendor to configure
it off, then you could route all output from the document scanning
software through some kind of routing or filtering software that
removed the offending attribute and then sent it on to PACS.
Messing with attributes like this can get you into trouble ("tag
morphing considered harmful"), but sometimes if operational needs
are incompatible with dysfunctional systems ...
What have your field engineers for the PACS, document scanners and
modalities said when asked about this?
David