the 1321 has jumpers to set gain, in 2 stages, the final gain is the product
of both;
the daq, instead has a gain that is auto calculated according to the input
voltage limits.
is the total gain the product of tha gain of scxi and daq?
cause i've found a VI that gets the total gain but it doesnt work, another
one instead gets the DAQ gain and works fine
i've found a VI to close the shunt resistor of 1321 but it doesnt work, so
how i can close this resistor?
any help will be appreciated
thank you
haveanicetrip
--
Umberto
Most likely, for either of these programs to work the NI-DAQ software
would have to be aware that the terminal block is present as an
accessory. Make sure the 1321 is configured as the accessory to your
SCXI module in MAX. Bring up the properties window of the SCXI module
as it appears in the chassis list in MAX and click on the Accessory
tab. SCXI-1321 should be selected.
I hope this helps!
-Russell