Hi, Benjamin. Did anybody reply yet?
In the past, when I've used Comet with Orbi hi-res MS/MS scans, I've used an MS/MS m/z bin size of 0.02. This corresponds to something Jimmy published: "
In fact, the optimum mass bin width is 1.0005 for low resolution data
such as that acquired on an ion trap detector, and we recommend a mass
bin width of 0.02 for high resolution spectra." (
https://doi.org/10.1007/s13361-015-1179-x)
If you're asking about a precursor mass tolerance denominated in ppm, most well-calibrated Orbi data can be searched with 10ppm or 20ppm tolerance. I have on a couple of occasions dealt with data from an Orbi WAY out of calibration, where the ppm tolerance exceeded 20ppm. In a case like that, the proper answer in an ideal world is "run maintenance on your instrument and recollect the data!"
Good luck!
Dave