Revision Manual y plan del Mantenimiento preventivo 2/2

10 views
Skip to first unread message

Adriana Mateos Jasso

unread,
Jun 20, 2013, 2:56:59 PM6/20/13
to Boya Holbox-Mexico, Ismael Mariño, Ismael Marino, Sergio Cerdeira Estrada, Eduardo Santamaria del Angel, Mauricio Israel Martínez Clorio, info...@semar.gob.mx

Hola a Todos

Les mando el manual preventivo para que lo revisen y me hagan llegar sus comentarios, observaciones y/o sugerencia.
asi como tambien el plan de Mantenimiento preventivo que se realizaria cada 2 meses.


Saludos

Adriana


M. C. Adriana Mateos Jasso
________________________________________________________________
Monitoreo Marino
Subcoordinación de Monitoreo Marino
CONABIO
ama...@conabio.gob.mx
+52 (55) 5004-4987
Liga Periferico - Insurgentes Sur #4903
México, D.F.



Manual_Mantenimiento_Preventivo_Boya_Holbox_20jun13.pdf

Victoria, Ivan - Xylem

unread,
Jun 20, 2013, 3:58:19 PM6/20/13
to Adriana Mateos Jasso, Boya Holbox-Mexico, Ismael Mariño, Ismael Marino, Sergio Cerdeira Estrada, Eduardo Santamaria del Angel, Mauricio Israel Martínez Clorio, info...@semar.gob.mx, Sneddon, Shawn - Xylem

Hi

The only comment I have is about the concentration of the hydrochloric acid to clean the sensors.

You can use 5% concentration of pure HCl (not 30%!!)

Or you can use pure vinegar from the grocery store.

There shouldn’t be any damage for leaving the submersible sensors or instruments in it for hours.

Right Shawn?

 

 

 

Ivan Victoria

Latin America Sales

 

Xylem Analytics

275 Martine Street

Suite 200, Unit #12

Fall River, MA, 02723, USA

(937) 767-7241 x832, Toll free: (800) 765-4974

Cell: 508-369-1601

Ivan.V...@xyleminc.com

www.aanderaa.com

 

Xylem_tag_4c 

 


 

Aanderaa | SonTek | Waterlog | YSI

 

 

--
Has recibido este mensaje porque estás suscrito al grupo "Boya Holbox-Mexico" de Grupos de Google.
Para anular la suscripción a este grupo y dejar de recibir sus correos electrónicos, envía un correo electrónico a boya-holbox-mex...@googlegroups.com.
Para obtener más opciones, visita https://groups.google.com/groups/opt_out.
 
 

CONFIDENTIALITY NOTICE: This e-mail, including any attachments and/or linked documents, is intended for the sole use of the intended addressee and may contain information that is privileged, confidential, proprietary, or otherwise protected by law. Any unauthorized review, dissemination, distribution, or copying is prohibited. If you have received this communication in error, please contact the original sender immediately by reply email and destroy all copies of the original message and any attachments. Please note that any views or opinions presented in this e-mail are solely those of the author and do not necessarily represent those of Xylem Inc.
image003.emz

Victoria, Ivan - Xylem

unread,
Jun 20, 2013, 4:10:08 PM6/20/13
to Sneddon, Shawn - Xylem, Adriana Mateos Jasso, Boya Holbox-Mexico, Ismael Mariño, Ismael Marino, Sergio Cerdeira Estrada, Eduardo Santamaria del Angel, Mauricio Israel Martínez Clorio, info...@semar.gob.mx
How much water do they need to put in the 30% HCl?

Ivan Victoria
Sent from my mobile phone

On Jun 20, 2013, at 4:05 PM, "Sneddon, Shawn - Xylem" <shawn....@Xyleminc.com> wrote:

They want 30% Hydrochloric acid but it must be diluted with water thereafter.  I believe it is sold in different concentrations.  Industrial grade is roughly 30% to 34%.  We use the 30% and dilute with water.  Pure vinegar will also do the trick.

 

From: Victoria, Ivan - Xylem
Sent: Thursday, June 20, 2013 3:58 PM
To: Adriana Mateos Jasso; Boya Holbox-Mexico; Ismael Mariño; Ismael Marino; Sergio Cerdeira Estrada; Eduardo Santamaria del Angel; Mauricio Israel Martínez Clorio; info...@semar.gob.mx
Cc: Sneddon, Shawn - Xylem
Subject: RE: Revision Manual y plan del Mantenimiento preventivo 2/2

 

Hi

The only comment I have is about the concentration of the hydrochloric acid to clean the sensors.

You can use 5% concentration of pure HCl (not 30%!!)

Or you can use pure vinegar from the grocery store.

There shouldn’t be any damage for leaving the submersible sensors or instruments in it for hours.

Right Shawn?

 

 

 

Ivan Victoria

Latin America Sales

 

Xylem Analytics

275 Martine Street

Suite 200, Unit #12

Fall River, MA, 02723, USA

(937) 767-7241 x832, Toll free: (800) 765-4974

Cell: 508-369-1601

Ivan.V...@xyleminc.com

www.aanderaa.com

 

<image004.jpg> 

 


 

Aanderaa | SonTek | Waterlog | YSI

 

 

 

From: boya-holb...@googlegroups.com [mailto:boya-holb...@googlegroups.com] On Behalf Of Adriana Mateos Jasso
Sent: Thursday, June 20, 2013 2:57 PM
To: Boya Holbox-Mexico; Ismael Mariño; Ismael Marino; Sergio Cerdeira Estrada; Eduardo Santamaria del Angel; Mauricio Israel Martínez Clorio; info...@semar.gob.mx
Subject: Revision Manual y plan del Mantenimiento preventivo 2/2

 

 

Hola a Todos

Les mando el manual preventivo para que lo revisen y me hagan llegar sus comentarios, observaciones y/o sugerencia.
asi como tambien el plan de Mantenimiento preventivo que se realizaria cada 2 meses.


Saludos

Adriana

M. C. Adriana Mateos Jasso
________________________________________________________________
Monitoreo Marino
Subcoordinación de Monitoreo Marino
CONABIO
ama...@conabio.gob.mx
+52 (55) 5004-4987
Liga Periferico - Insurgentes Sur #4903
México, D.F.

 

 

--
Has recibido este mensaje porque estás suscrito al grupo "Boya Holbox-Mexico" de Grupos de Google.
Para anular la suscripción a este grupo y dejar de recibir sus correos electrónicos, envía un correo electrónico a boya-holbox-mex...@googlegroups.com.
Para obtener más opciones, visita https://groups.google.com/groups/opt_out.
 
 

Reply all
Reply to author
Forward
0 new messages