Hi Ivan,
The force Dmax to zero does essentially what you would think. If you set it for yes ('Y') then the P(r) function is mathematically constrained to zero at the selected maximum dimension. If you set it to no ('N') then it isn't. Obviously, from a physical point of view the protein should have some distance where there are no more pairs of electrons, and at that maximum distance (Dmax) the P(r) function (the r^2 weighted electron pair distance distribution) will naturally go to zero.
Problems finding Dmax usually mean there is some amount of polydispersity in your sample. This could be due to aggregation or oligomerization, so that you have some amount of larger object in solution, contributing to a long tail of the P(r) function out to high distances. The other common case is polydispersity in conformation, for example from flexibility in your system. If a protein is adopting some small amount of extended conformations, for example, those contribute to a long tail of the P(r) function out to high distances as well. You'll want to look at the profile in other ways (e.g. Guinier and Kratky plots) to determine which of those you're seeing.
It's okay to accept the tail of the P(r) function to be a tiny bit above zero without forcing. Most people, myself included, would be okay with the P(r) function you show with Dmax=100 and forcing turned off. Your 85 isn't great, it's quite visibly above zero. Probably somewhere between those two there's a point where it first comes down to mostly zero, and that's what I'd chose as the maximum dimension for this particular system. But with something like this there's room for reasonable people to disagree, which is why we say that uncertainty in dmax can be quite large, 10% is a common number people throw around.
For your second issue, I suspect it's because you don't have a good initial Dmax. 79 is obviously too short for that data (you can see this in the first image you sent comparing different values). When you truncate, you're fitting a different region of the profile, and that's changing it. Ideally if you have a good dmax chosen, truncating should have little or no effect on the shape of the P(r) function.
So, my recommendation is try to find a Dmax in the 85-100 range, and that will probably take care of the second issue.
All the best.
- Jesse
----
Jesse Hopkins, PhDDeputy Director
BioCAT, Sector 18
Advanced Photon Source