SYBR green: melt curve Tm shift

4 views
Skip to first unread message

Ed Siefker

unread,
Dec 21, 2009, 10:57:39 AM12/21/09
to met...@magpie.bio.indiana.edu
I am having a small problem with my SYBR green qrtpcr.
I have 4 primer sets running against 2 different cDNA
samples. The primers are all at the same concentration
(300 nm). I prepared both cDNA samples side by side, in
exactly the same way and loaded 10ng of each.

The problem is, when I look at the melt curve, I'm seeing
a decrease in Tm of the product of about 1 degree C from
one cDNA to another. This Tm shift occurs for all 4 primer
sets. In both cases, there is a single sharp peak, so I'm
doubting I have primer dimers.

Does anyone have any ideas? I've used these primers before
on different samples, and the melting curves always exactly
coincided, no matter what the cDNA was.

Duncan Clark

unread,
Dec 22, 2009, 7:19:41 AM12/22/09
to
Historians believe that in newspost
<mailman.11.126142...@net.bio.net> on Mon, 21 Dec 2009,
Ed Siefker <ebs1...@creighton.edu> penned the following literary
masterpiece:

>
>Does anyone have any ideas? I've used these primers before
>on different samples, and the melting curves always exactly
>coincided, no matter what the cDNA was.

Something in one cDNA sample that isn't in the other. Could be salt, Mg,
EDTA, glycerol etc. It doesn't take much to shift 1C.

I bet if you could desalt it somehow (G25 or G50 spin column), it would
be fine.

Duncan
--
I love deadlines. I especially like the whooshing noise they make as
they go flying by.