UMIK-1 calibration file 90deg won't load

200 views
Skip to first unread message

Curbo 34

unread,
Mar 28, 2022, 4:09:21 PM3/28/22
to AudioTool Discussion Group
Hey guys, love the app, so much utility in such a small package.

Anyways, I've had a UMIK-1 and used it with the standard calibration file to great success, loads the calibration file provided by minidsp instantly by changing the .txt extensions to .cal, however I cannot get it to load the 90deg calibration file no matter what. Simply says "Cannot read from storage" this is true for all devices tested (S20 ultra, S22 ultra, and an amazon fire HD tablet). 
I attempted the fix mentioned to have it read as a Dayton file (Adding * to the start of the comment line) with no success. 
Since the normal cal will load, I loaded it then exported it, and opened it in a text editor to try and make sense of what it was doing to the file. I was under the impression it was just averaging the data into a 1/3 octave scale, but through my attempts to do this via a spreadsheet editor, the data I get back is VASTLY different than the data in the exported sheet.
 Attached below are all files mentioned (Cals straight from minidsp with extension changed, output of straight cal from audiotool)

Any help is greatly appreciated, I'm somewhat out of time to play with trying to figure it out so hopefully someone knows.
 I'd also appreciate if you could add instructions directly for the umik-1 to the audiotool manual or generalize the minidsp section to include it.
Thanks everyone!
Tttd.cal
7090114_90deg.txt
7090114.txt

Julian Bunn

unread,
Mar 29, 2022, 5:31:22 PM3/29/22
to audiotool-dis...@googlegroups.com
The 90 degree file should work fine if you remove the second comment line in the file ... AudioTool is not expecting that!

Julian

--
You received this message because you are subscribed to the Google Groups "AudioTool Discussion Group" group.
To unsubscribe from this group and stop receiving emails from it, send an email to audiotool-discussio...@googlegroups.com.
To view this discussion on the web visit https://groups.google.com/d/msgid/audiotool-discussion-group/b7b4795f-602b-4a7e-944b-bb698f75dc75n%40googlegroups.com.
Reply all
Reply to author
Forward
0 new messages