Dear AudioTool Development Team,
I would like to suggest an improvement to the SPL calibration process in AudioTool.
Currently, AudioTool allows the user to manually adjust the Global Offset during calibration. I believe this process could be made much easier and more reliable by adding an Automatic Calibration function.
Proposed featureThe idea is to make the calibration process similar to the procedure used by professional sound level meters and dosimeters.
The user would follow these steps:
AudioTool asks the user:
"What is the reference SPL level?"
The user enters the exact value provided by the calibrator's calibration certificate, for example:
94.08 dB SPL
Suppose the calibrated acoustic calibrator is certified at:
94.08 dB SPL
When connected to the microphone, AudioTool currently reads:
92.6 dB SPL
The application would automatically calculate the difference:
94.08 − 92.60 = +1.48 dB
AudioTool would then automatically apply a +1.48 dB Global Offset.
After calibration, the application should display approximately:
94.08 dB SPL
The user would not need to manually adjust the offset and repeatedly check the result.
Why I believe this would be usefulThis would make the calibration process:
It would also be useful to allow decimal values such as 94.08 dB, rather than requiring the user to calibrate only to an integer value such as 94 dB.
Possible user interfaceFor example, the current calibration menu could have an additional option:
Automatic SPL Calibration
After selecting it, AudioTool could display:
Place the microphone in the acoustic calibrator and start the calibrator.
Reference level:
[ 94.08 ] dB SPLCurrent measurement:
92.60 dB SPL[ CALIBRATE ]
After pressing CALIBRATE, AudioTool would calculate and apply the necessary Global Offset automatically.
It could then display:
Important considerationCalibration completed
Reference: 94.08 dB SPL
Measured: 94.08 dB SPL
Applied Offset: +1.48 dB
I am specifically suggesting this as an automation of the existing Global Offset calibration process, rather than replacing the existing 1/3-octave microphone calibration functionality.
The purpose would simply be to establish the absolute SPL reference level automatically using a known acoustic calibrator.
For example, if the user has an acoustic calibrator with a calibration certificate stating that its actual output is 94.08 dB SPL, AudioTool should be able to use that exact value as the reference and automatically determine the required offset.
I believe this would be a relatively simple but very useful improvement to the calibration workflow, especially for users using calibrated measurement microphones and external acoustic calibrators.
Thank you for considering this feature.
Best regards,
Fernando