dpb wrote:
> >> Another thread on here mentions using a Kill-A-Watt meter
> >
> > For inductive loads like motors, I wouldn't trust those consumer-
> > grade "kill-o-watt" meters any further than I could throw them.
>
> They only claim 2% accuracy but they do measure true RMS and PF.
> I don't know why you'd think any less; it's surely not difficult
> nor expensive in today's microprocessor world to sample and compute
> _very_ inexpensively.
I found some interesting material from a couple of forum threads.
The first one:
http://forums.anandtech.com/showthread.php?t=137169
To quote from that thread:
===============
I had a friend over today that wanted to show me his new purchase, a
fluke 1735 , so I thought I would do a quick comparison with the
killawatt. The killawatt isn't a bad meter, especially for the price.
Its does have its flaws though. Comparing it with a fluke 1735 I found a
few differences.
The killawatt is terrible at inductive loads, so don't use it for
measuring those. Inductive loads would be things typically with motors,
compressors. So not good for measuring a refirgertator or washing
machine. It measures them , but its results are not .2% accurate.
Refrigerator with compressor running 921 watts, fluke 841.8 watts
Its not accurate at measurements of small wattages. Things like the
power usage of a dvd player in standby mode are not accurate. Killawatt
said dvd player was using about 5 watts, fluke 2.64 watts.
The sampling rate is very low compared to meters like the fluke, so it
can miss quick spikes or surges in usage . When I used it to measure the
power usage of a 51" hdtv at turn on, it constantly gave different
readings. Range from 410 watts to 504 watts. Fluke 448 to 453 watts.
=================
The second thread:
http://fuelzilla.com/homepower/Power-measurement-question-Kill-A-Watt-and-Watts-Up-Pro-33213-.htm
Quoting several posts from that thread:
=================
I have both a Kill-A-Watt and a Watts Up? Pro power meters. To check how
close they're against each other, I connected them in series with the
Kill-A-Watt going into the plug, then the Watts Up? Pro and the load.
They should read about the same or the Kill-A-Watt should read the power
consumption of the load plus the Watts Up? (no more than a watt or
two). Comparing the actual meaasured values, the Watts Up? Pro
consistently give a value about 10% higher than the Kill-A-Watt with
inductive and non-sinusoidal loads and not quite as much difference with
resistive loads.
With the computer I'm using to write this message connected as the load,
Kill-A-Watt is reading 174W and Watts Up? Pro is reading 190W.
Both devices agrees within a reasonable degree against a known good DMM
for voltage.
DMM: 120.3V
W: 121V (does not resolve to 100mV)
K: 120.4V
Current do not agree with each other:
DMM: Unable to measure, my DMM is not true RMS capable
W: 2.37A
K: 2.19A (w/ no load, device reads 0.02A, 0.0W)
PF: both devices reads 0.66
Here are the differences in construction:
Voltage measurement:
Watts Up? Pro: An isolating transformer drops the voltage used for both
measuring the voltage and powering the device.
Kill-A-Watt: It is directly powered from the AC line through a series
R-C circuit and a separate resistive divider is used for voltage
measurement.
Current measurement:
Watts Up?: Current transformer.
Kill-A-Watt: Shunt
-----------------
Realistically speaking, high frequency load (say something that draws
power in 25 15A spikes each half cycle, such as some copy machine/laser
printer heater controller) rich in harmonics and high in crest factor
would not give the meter same accuracy as measuring a plain resistive
load.
Examples of highly harmonic loads:
Almost ALL IT equipments
Most home electronics
Residential electronic ballasts and CFLs
These loads frequently have a THD greater than 60%.
-------------------
I purchase three kill-a-watt units last year and put them all into each
other for comparisons, They all matched each other as close as the
LSDigit would allow.
I also took one of the units and checked it against a lab standard
traceable to the NBS standards and compared it for Voltage, Current,
Power and Reactive Power and I can tell you this, you could use this
unit interchangeably with our lab standard. No digit showing on the
kill-a-watt unit to it finest resolution was out by even one count. Now
our lab standard has a few more digits.
The all Vars (reactive power) and no real power (watts) comparison may
be off a little on our lab standard and I did not compensate using known
documented accuracy tables. The accuracy formulae is always divided by
the PF which makes ???? accuracy but this lab standard is about as
accurate as it gets in Canada without controlled environments etc..
etc..
I am really impressed with the Kill-a-Watt units. No tests on waveform
distortion or harmonics were performed to date by me.
Waveform distortion form factor may be where the differences are found.
OTOH the Watts Up may just not be calibrated properly or junk.
------------------
An update on Kill-A-Watt,
I ripped it apart and started probing around. The shunt's output is
rather low.
The shunt is placed across the neutral and looks like a 12 gauge wire
looped into a U-shape, but I'm not sure what its made of.
It gives a 47mV voltage drop with a 12A 1.5kW space heater connected,
which tells me the shunt is 3.917 miliohms. The signal from shunt is
routed on the board for 3" or so to an LM2902N op-amp. With around
560mW of dissipation, the shunt gets hot to touch and I'm not sure how
much the heating affects the resistance of the shunt.
The current resolution on the Kill-A-Watt is 0.01A and this translates
to current signal input resolution of 39ÁV, which might make the device
suspecticle to noise considering the signal path is not shielded at all.
With a one kilowatt resistive load, it jumps around few tens of watts.
==================
(end)