Hello Nima,
I have just confirmed that I am still able to acquire at 1kHz with my ORCA Flash 4.0. I used the following parameters, configured from the acq4 manager window:
exposure=1ms
binningX=4, binningY=4
regionX=0, regionY=960
regionH=128, regionW=2048
ScanMode=2.0
If I start the camera via the camera module, the FPS readout varies between about 800 and 1200 fps; I believe the instability in this readout is just due to variance in the arrival time of frames to the software; all frames are still making it in without being
dropped. I can confirm this by using a digital input on my NI board to record the exposure times while flashing an LED in sync:
Are you still having this problem if you use the exact same parameters? Note that the exact Y value 960 is important here, since it makes the ROI centered along the sensor's Y axis.
Cheers,
Luke