The PIVlab result shows a lower flow rate than the actual measured flow rate

55 views
Skip to first unread message

Soccer Travel

unread,
Jul 15, 2024, 3:07:45 AMJul 15
to PIVlab
Hello, I got a good result from your advice last time.
The last problem remains, but there is a difference in the value of the average flow rate obtained by PIVlab than the actual measured flow rate.
The actual measured fluid is approximately 0.43 m/s, but the average flow rate obtained with PIVlab is approximately 0.2 m/s.
The delta T of the image pair used is 1.925 ms.
Please give me proper advice.
thakyou.스크린샷 2024-07-15 155951.png스크린샷 2024-07-15 160711.png

thielickeoptolution

unread,
Jul 15, 2024, 6:47:26 AMJul 15
to PIVlab
Then your delta t or your reference distance are wrong... Please double-check (also with an oscilloscope and a photo-diode or LED inside the laser sheet).

Soccer Travel

unread,
Jul 17, 2024, 3:16:03 AM (12 days ago) Jul 17
to PIVlab
Thankyou for your reply.
The actual length is the length and pixel measured by putting the checkered paper in the duct, so is there a better way?
And the actual delta T is 1.925ms, but the calibration's time step is not to enter delta T?
I'll be waiting for the reply.

Thankyou!

2024년 7월 15일 월요일 오후 7시 47분 26초 UTC+9에 thielickeoptolution님이 작성:

William

unread,
Jul 17, 2024, 9:18:30 AM (12 days ago) Jul 17
to PIVlab
Maybe you have the raw data and the calibration image?
PIV cannot give wrong velocities. Only when the calibration is wrong it will give wrong results.

Soccer Travel

unread,
Jul 17, 2024, 11:13:39 PM (12 days ago) Jul 17
to PIVlab
Thank you for your answer.
Instead of the current calibration image, we put the checked paper in the measurement area to reflect the pixel and the actual length.
Can you recommend another method instead of this one?
I look forward to hearing from you. Have a nice day
test1_00002_B.pngtest1_00002_A.png
thankyou.processed-0AC3F87A-B437-46BE-95BE-6E2FFCDE09C5.jpegprocessed-42D9D0C8-A800-44F6-A399-7E6DC2898F5C.jpeg
2024년 7월 17일 수요일 오후 10시 18분 30초 UTC+9에 William님이 작성:

William

unread,
Jul 18, 2024, 4:55:07 AM (11 days ago) Jul 18
to PIVlab
Your data looks fine, I get the same result. But PIV does not have a bias of 100%, this will only happen if the timing  / scaling is wrong.

Soccer Travel

unread,
Jul 18, 2024, 10:13:43 AM (11 days ago) Jul 18
to PIVlab
How should I modify timing/scaling?
I can't get a clue....

2024년 7월 18일 목요일 오후 5시 55분 7초 UTC+9에 William님이 작성:

thielickeoptolution

unread,
Jul 18, 2024, 10:21:03 AM (11 days ago) Jul 18
to PIVlab
Validate the timing with an oscilloscope and a photo diode / LED.

thielickeoptolution

unread,
Jul 18, 2024, 10:32:17 AM (11 days ago) Jul 18
to PIVlab
Does your calibration image and the PIV image have the same resolution? Maybe you apply pixel-binning somewhere...?

Soccer Travel

unread,
Jul 23, 2024, 12:19:34 AM (7 days ago) Jul 23
to PIVlab
hello!

This is the result of checking with the oscilloscope.
Initially, delta T was set to 1s to verify that it appeared as 1s on the oscilloscope.
Lasers 1 and 2 are connected through channels c and d, respectively.
We proceeded by setting delta T at delay intervals of c, d  in the pulse generator.
In the oscilloscope below, the grid is set at 1s intervals.



Soccer Travel

unread,
Jul 23, 2024, 12:22:31 AM (6 days ago) Jul 23
to PIVlab
KakaoTalk_20240723_125608109_01.jpgKakaoTalk_20240723_125608109_02.jpgKakaoTalk_20240723_125608109.jpg


Soccer Travel

unread,
Jul 23, 2024, 12:24:10 AM (6 days ago) Jul 23
to PIVlab
The image below is when delta T is set to 2ms.
On the oscilloscope, one grid is set to 10ms.
So if we set delta T to 2ms,
In PIVlab, I thought that even if timestep was 2ms, it was not a different value from the actual one.
KakaoTalk_20240723_125608109_03.jpg
KakaoTalk_20240723_125608109_04.jpgKakaoTalk_20240723_125608109_05.jpg

The resolution of the calibration image is no different.
I look forward to your advice.
Thank you for your kind reply.

William

unread,
Jul 23, 2024, 8:26:59 AM (6 days ago) Jul 23
to PIVlab
Hi, thanks for this update. But I can only repeat: " Validate the timing with an oscilloscope and a photo diode / LED."
You are triggering the flashlamps of your laser. Ideally you would trigger the Q-Switches. When you trigger the flashlamps, then there can be additional errors. To be sure, you need to measure the light emission from the laser. This can easily be done by putting a LED or photodiode inside the laser sheet, then connecting it to your oscilloscope. The LED / Photodiode will emit a voltage when it is hit by the laser pulse. Then you can measure the delay between two pulses.

Another question: Has the calibration image be taken under the exact same conditions as the PV images? Have you aligned it with the laser sheet? Have you moved the camera after calibration?