In our lab we also have an ASI CRISP which, like the Definite Focus, projects the image of an IR LED onto the coverslip/water interface. A design schematic is here:
http://www.asiimaging.com/wp-content/uploads/2015/09/CRISP_diag.jpg. (I can verify it's correct because I've taken the CRISP apart multiple times :) The reflected light comes back to the CRISP module and passes through a relay lens. In the focus of the lens is a wedge mirror that is positioned by micrometer so that it blocks half of the focused spot. This essentially makes the direction of the light passing through the focus very sensitive to the displacement of the coverslip/objective distance. I believe the effect is known as Foucalt's knife edge test (
https://en.wikipedia.org/wiki/Foucault_knife-edge_test). To measure this displacement they use a dual photodiode and compute the difference between the two halves, sending a feedback signal to the piezo stage/positioner.
My somewhat subjective opinions on this modality vs. TIR-based autofocus are:
TIRF-based advantages
1) Extremely sensitive to coverslip/objective displacement --> precise locking
2) No problem getting a strong return signal due to use of a laser
TIR-based disadvantages
1) TIR alignment is always challenging and subject to misalignment
2) Any distortion in the return beam shape prevents accurate centroid determination
IR LED-based advantages
1) Not so sensitive to scattering of the IR light by cells on the coverslip --> easier to lock the focus and perform automatic scans
2) Alignment is not as critical because only differences are measured (rather than a fit to a Gaussian)
IR LED-based disadvantages
1) The axial range over which we can of lock appears to be smaller, but I
don't have numbers to confirm this, so I might be wrong. With the
CRISP, we have a small linear range in the response curve; if the
displacement gets driven outside this range, the negative feedback
becomes positive and the system drives itself out of focus.
2) The reflection from a coverslip/water interface is quite weak.
I must also add that we have been struggling with the CRISP firmware and control software for two years. I'm not sure if this is because the Foucalt knife edge method is inherently difficult to control or if the software is just a bit buggy to use.
I hope this helps people think more about their autofocus options and the strengths and weaknesses of both techniques.